    Are Illumina missing a trick over cluster density?

    Don't need any help, just curious. As I understand it Illumina make their flowcells with an excess of primers bound to the slide. It's the up to whoever loads the cBot/clusterstation to work out how much library to add. This is quite a delicate business as everyone wants the highest read numbers possible, but if you overload then the clusters overlap and you loose a lot of the data. Would it be easier for Illumina to make the flowcells with the primers already attached at the optimum density. Then the operator can add an excess of library happy in the knowledge that optimal cluster density can be achieved. If they were feeling really flash then they could somehow print the primers on the slide in an ordered pattern so that the computer knows where to look for clusters, and thus pack them in even tighter.
    I'm probably missing something really obvious. Maybe someone can enlighten me.
    Cheers.
  • james hadfield
    Moderator
    Cambridge, UK
    Community Forum
    • Feb 2008
    • 224

    #2
    Hi Henry,

    It is not the oligos that dtermine cluster density but the concentration of library loaded by the user. There is little Illumina can do to help individuals quantitate and dilute accordingly. Although there are protocols for Q-PCR to help with this.

    James.

    Comment

    • henry.wood
      Member
      • Apr 2010
      • 63

      #3
      Originally posted by james hadfield View Post
      It is not the oligos that dtermine cluster density but the concentration of library loaded by the user
      I know this is the case at the moment, but I was just idly wondering if it might be easier the other way round. At the moment the flowcell is set up so that if the user loads the library properly then the clusters will be dotted around at the correct density with unused oligos all over the place. As you say it's the library concentration that's the limiting step. If the flowcells were made differently with the oligos already at the correct density, with the goal of getting a library fragment onto every oligo then it would be easier for the user since they could simply wash over an excess of library. I don't mind being wrong about this, we were just having a conversation over coffee and it seemed like a good idea.

      Comment

      • krobison
        Senior Member
        • Nov 2007
        • 734

        #4
        The way I picture this (and please do correct me if I'm off base) is that your DNA fragment falls to the flowcell surface & then amplifies out a certain radius from the initial site. That amplification relies on the surface-bound primers. So the library concentration ensures scatter around the flowcell, but you need a certain density of primers to ensure amplification.

        But, this assumes all primers are the same. If you had two flavors of primers ("docking" and "amplification"), then the docking primers could potentially be less dense (and perhaps patterned to avoid cluster collisions) but retaining a forest of amplification primers.

        Comment

        • henry.wood
          Member
          • Apr 2010
          • 63

          #5
          That's pretty much what I imagined. I had kind of envisaged the 'docking' primers being there at the outset, and then washing the flowcell with 'amplification' primers during the various cycles. I could see a possible problem if some people wanted 200bp fragments and others wanted 500bp, so the bridges would be a different size.
          It's all a bit of coffee time musing really. Probably the good folks at Illumina have brighter sparks than me working to increase output. I'll kick myself if someone runs with it and makes millions.

          Comment

          • SeqTruth
            Member
            • Apr 2009
            • 12

            #6
            I had a similar idea:

            If one printed the capture/amplification oligos onto the flowcell, then it might be possible to perform a sort of sequence capture/enrichment in situ on the flowcell before cluster amplification -- each address (feature) would simply contain sequence-specific oligos targeting a region of interest. Of course, one would require a lot of redundancy amongst features, and very high loading concentrations of fragmented sample DNA to ensure that each feature had a high probability of capturing its cognate target. But kind of intriguing none-the-less.

            One might also imagine a printed flowcell with capture/amplification oligos for genomic targets of interest, in a device that allows one to flow a constant stream of sample across the flowcell. One could then carry out an extended capture/hybridisation process by flowing a large volume of (indexed and pooled?) sample material through the flowcell, possibly circulating the sample repeatedly through the flowcell.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Just guessing, but I presume the oligos are either synthesized on the surface of the flowcell or bound later. So to implement the docking/amplification scheme, you would need most of the oligos to be blocked for binding. Likely double stranded. Plus you would need to be able to place the docking oligos at known intervals.

              Seems like a fairly simple idea, so presumably Illumina has looked at it. Must be too expensive to fabricate such a flowcell.

              Also you would need to look at what the incentive would be for Illumina to implement this method. Currently if you are high or low the effect would be your needing to run another lane (or set of lanes.) That doesn't hurt Illumina. Especially since skilled operators are said to be capable of accurately deliver the right density of library molecules to get good results. (I think this has reach near "Urban Legend" status.)

              Generally Illumina will report results of good runs for their specs, so they remain competitive against their bead-based competitors. (Who have a version of exactly the same problem--during emulsion PCR.)

              Were the docking/amplification method deployed, there would still likely be manufacturing issues that would create bad batches. But Illumina would be on the hook for these, whereas their customers currently accept the risk of over or under-estimating the concentration of their sequencing amplicons.

              So my prediction would be that some competitor would need to pull closer to them in market share before this methodology would be seriously considered by Illumina.

              --
              Phillip

              Comment

              • HESmith
                Senior Member
                • Oct 2009
                • 512

                #8
                In addition to the fabrication challenge, the kinetics of reassociation might prove problematic. Annealing rates are dependent upon both library and primer concentrations. Currently, 10e8 - 10e9 library molecules (100 ul/lane @ pM concentration) are annealed to a vast excess (several orders of magnitude higher) of tethered primers with ~10% yield (10e7 - 10e8 clusters). If the docking density for clusters were increased 10X (to 10e9 docking primers), the library concentration would have to approach the original concentration of tethered primers (micromolar range?) to anneal with the same kinetics. And, at higher library concentrations, there's the confounding effect of library molecules reannealing to each other.

                Comment

                • konrad98
                  Member
                  • Jan 2009
                  • 17

                  #9
                  A year or two ago Illumina were talking about producing an ordered flowcell so that you could pack more clusters in using technology borrowed from some of their other products.

                  I imagine Illumina would like to get to the stage where one or two lanes could generate enough to cover a whole human genome. I can't imagine they'd give up on research which would keep up the pressure on Life Tech.

                  But as pmiguel says - who knows when they might chose to release it if they do perfect it?

                  Comment

                  • csc
                    Member
                    • Aug 2009
                    • 22

                    #10
                    As konrad98 has said "Semi-ordered arrays" did show up in some of Illumina's future higher throughput roadmaps but back then (perhaps two-ish years ago) the timeline indicated that it would be potentially available by now. They have done a good job scaling the throughput the last couple of years so maybe they decided to hold onto it as a future improvement when they have exhausted the image processing avenues...

                    Comment

                    • simonandrews
                      Simon Andrews
                      • May 2009
                      • 870

                      #11
                      My impression from the semi-ordered arrays was that the advances Illumina have made in image analysis greatly reduced the benefits from ordering the clusters.

                      I can't remember who I was talking to (I think it was an Illumina person), but I heard that the clusters on a flowcell never actually overlap - they naturally stop expanding when they hit an adjacent cluster, so theoretically you could end up with a solid lawn of clusters which you later let your image analysis sort out. I can't imagine that ordered arrays would offer much of a capacity increase over this.

                      Comment

                      • GW_OK
                        Senior Member
                        • Sep 2009
                        • 411

                        #12
                        Yeah, they had data on this at last years AGBT. It was a really interesting demonstration, the clusters they showed were superimposed but not commingled. This lead them to the improved cluster calling algorithm that was being released to the GAIIx at that time.

                        If you could get a really well defined pixel map of the flowcell i think you could probably deal with a solid lawn of clusters.

                        Comment

                        • james hadfield
                          Moderator
                          Cambridge, UK
                          Community Forum
                          • Feb 2008
                          • 224

                          #13
                          Originally posted by konrad98 View Post
                          I imagine Illumina would like to get to the stage where one or two lanes could generate enough to cover a whole human genome. I can't imagine they'd give up on research which would keep up the pressure on Life Tech.
                          A recent report on GenomeWeb from Illumina CEO Jay Flatley decribes an example in which company researchers generated 1.13 terabases of data in a 14-day run on the HiSeq 2000 with 2 x 150 base reads, about 80 gigabases per day.

                          Thas't a human genome in less than a lane!

                          I am sure I posted something working out how it might be possibel to get to 2TB on HiSeq with some simple improvments in read length and cluster density. Seems they have cracked it and we are bound to hear a lot more at AGBT.

                          I wonder what Life Tech et al are going to show?

                          Comment

                          • simonandrews
                            Simon Andrews
                            • May 2009
                            • 870

                            #14
                            Originally posted by james hadfield View Post
                            A recent report on GenomeWeb from Illumina CEO Jay Flatley decribes an example in which company researchers generated 1.13 terabases of data in a 14-day run on the HiSeq 2000 with 2 x 150 base reads, about 80 gigabases per day.
                            Oh great - even more data.

                            Won't someone please think of the bioinformaticians?

                            Comment

                            Latest Articles

                            Collapse

                            • SEQadmin2
                              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                              by SEQadmin2




                              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                              Today, 06:46 AM
                            • SEQadmin2
                              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                              by SEQadmin2



                              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                              ...
                              10-14-2026, 09:55 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by SEQadmin2, 10-28-2026, 09:57 AM
                            0 responses
                            22 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-28-2026, 09:51 AM
                            0 responses
                            14 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-22-2026, 11:13 AM
                            0 responses
                            67 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-22-2026, 10:48 AM
                            0 responses
                            47 views
                            0 reactions
                            Last Post SEQadmin2  
                            Working...