  • liuqy
    replied
    Originally posted by XIAOXIAO View Post
    Hey if you used AmPure beads for purification, what is the ratio of library versus beads for purificaiton?
    There is a charter picture here, which demonstrates beads volume vs sample volume (note: not sample to beads). NVIGEN MagVigen uses the same protocols as AmPure XP.
    MagVigen magnetic nanobeads enables simple and high throughput automated workflow, with increased capability and capacity for effective DNA Size Selection.

    Leave a comment:


  • liuqy
    replied
    Originally posted by yuehuang View Post
    Thanks much for the tips. I was always bothered by the primer dimer when used NEB kits for library prep. The AMPure XP worked fabulously for size selection, however didn't work for getting rid of the primer dimer. I had to use Pippin Prep to clean-up the products, which caused more than half loss. After switched to Nextera kit, no more primer dimer problem, but primer contamination presented sometimes. The manufacturer's suggestion is only 25uL beads for 50uL products. I would try your 1.8:1 ratio next time.
    You may try NVIGEN MagVigen. It's observed that AmPure XP lags MagVigen in primer dimer/adapter dimer cleanup. You can check these different contamination levels (>200%, >100%, >50%, 20%) (https://www.nvigen.com/dna-clean-up/)

    Disclosure: I work at NVIGEN. We cordially invite you to try MagVigen. In case you find problems using it, we will be happily work with you to troubleshoot or fine tune it to meet your needs.

    Leave a comment:


  • XIAOXIAO
    replied
    thanks! i will try 1:1

    Leave a comment:


  • Simone78
    replied
    Originally posted by XIAOXIAO View Post
    Hey if you used AmPure beads for purification, what is the ratio of library versus beads for purificaiton?
    DEpending on which fragments you want to retain and which ones you would like to get rid of. To remove short fragments (dimers/primers) use a 0.8-1:1 ratio beadsNA. If you want to pull down everything use a 1.8:1. If you want a specific size range, do a double (two-sided) purification where in the first your DNA is in the supernatant and in the second is on the beads, or vice versa.
    Best,
    Simone

    Leave a comment:


  • XIAOXIAO
    replied
    Hey if you used AmPure beads for purification, what is the ratio of library versus beads for purificaiton?

    Leave a comment:


  • yuehuang
    replied
    Originally posted by pmiguel View Post
    I have a partial answer. We tried a few modifications to the TruSeq protocol, one of which was to do the PCR step on non-size selected library, followed by size selection on the Pippin Prep. Our yield of final library was much lower than what we got following the normal protocol. But it was still enough to sequence.

    It worked.

    But there are a couple of additional issues for a non-amplified library:

    (1) The Y-adapters apparently cause library fragments to migrate aberrantly on some, but not all, electrophoretic conditions. Details are in another thread. I think they may run at their correct size on e-gels, but appear larger than they actually are under most commonly used electrophoretic conditions. After enrichment PCR the adapters are no longer Y-ed, so the effect disappears and you see the amplicon's true sizes.

    (2) Just to state the obvious: when you work with very limiting amounts of DNA, you have to watch losses to binding against plastic-ware, etc. A typical microfuge tube may be able to bind 1 ng of DNA (complete guess.) So, if you have 200 ng, who cares. But if you have 1 ng total, you might lose the majority of your library if you don't use low-bind plastic-ware.

    --
    Phillip
    thank you so much for the low-bind tube tip. I tried the same procedure, got the perfect size selected and clean library prep products, but much lower yield. Lucky to me, it's still sequencable, but very scary.

    Leave a comment:


  • yuehuang
    replied
    Originally posted by monad View Post
    2nd gel purification is the best option here, but you lose lots of library DNA. In this case, you can set additional round of PCR will take care of it. Stick with not more than 5 round of PCR. If your bioanalyer peak is good, you will be fine even with 10-15 cycles of PCR.
    Did you mean, the already purified DNA library to go another PCR at about 5 cycles, checked by bioanalyzer then go amplification again in 10-15 cycles to accumulate enough yield for sequencing?

    Leave a comment:


  • yuehuang
    replied
    Originally posted by pbluescript View Post
    I have attached an image showing the effect of altering the ratio of AMPure beads/sample. You can see that using a lower ratio will remove increasing sizes of DNA fragments. You can even use this for size selection by using two different concentrations. For example, if you wanted to get a band centered ~400 bp, you could use 0.7X beads and then take the supernatant. This would include all the DNA fragments that were smaller than ~500bp and didn't bind to the beads. Then mix the supernatant with 0.9X beads and everything larger than ~300bp would bind. Elute the bound DNA and you'll have a fairly broad, but good enough to sequence, library.
    If you try the beads, there can be a bit of lot-to-lot variability, so I would test them with a 100 bp library as I did in this gel.

    I haven't used the Truseq kit, but I have used the Truseq adapters with outsourced enzymes/buffers and the same problem can occur.

    Great job! I played with this too. But my concern focused on the size adjusted concentration, your gel is very convincible.

    Leave a comment:


  • yuehuang
    replied
    Originally posted by Simone78 View Post
    I had the same problem(s). To get rid of the adaptor-adaptor dimers I now do size selection using 2% E-gel (great because you can also select more than one single fraction, so you have a backup in case something goes wrong during the pre-amplification step). And you don't need an additional purification step, just pipette out the water (containing your DNA) from the well. To remove the primer dimers I use AMPure XP (1.8:1 ratio) which removes (almost) 100% of the fragments <100 bp.
    Thanks much for the tips. I was always bothered by the primer dimer when used NEB kits for library prep. The AMPure XP worked fabulously for size selection, however didn't work for getting rid of the primer dimer. I had to use Pippin Prep to clean-up the products, which caused more than half loss. After switched to Nextera kit, no more primer dimer problem, but primer contamination presented sometimes. The manufacturer's suggestion is only 25uL beads for 50uL products. I would try your 1.8:1 ratio next time.

    Leave a comment:


  • captainentropy
    replied
    Originally posted by rnaeye View Post
    pbluescript,
    what is the DNA ladder you used on this gel. thank you!
    He's using both. The 100 bp ladder was what he was size-selecting (to illustrate the use of the SPRI bead ratios), the 1kb+ ladder is the size marker for the gel.

    Take a look again at the text above each lane.

    Leave a comment:


  • rnaeye
    replied
    pbluescript,
    what is the DNA ladder you used on this gel. thank you!

    Leave a comment:


  • DavidAntonio
    replied
    Originally posted by pbluescript View Post
    The 100 bp ladder in the image is this one:

    The 1kb+ ladder in the image is this one:
    http://products.invitrogen.com/ivgn/product/10787026
    Thank you so much. Excuse, but I have another question; Have you used Agilent Haloplex?

    Leave a comment:


  • pbluescript
    replied
    Originally posted by DavidAntonio View Post
    I am sorry about the Necro Post, but I have a problem in this moment about this. Could you tell me the weigth (in bp) of your 1 kb ladder? It would be very useful for my thesis. Thanks a lot.
    The 100 bp ladder in the image is this one:

    The 1kb+ ladder in the image is this one:

    Leave a comment:


  • DavidAntonio
    replied
    Originally posted by pbluescript View Post
    I have attached an image showing the effect of altering the ratio of AMPure beads/sample. You can see that using a lower ratio will remove increasing sizes of DNA fragments. You can even use this for size selection by using two different concentrations. For example, if you wanted to get a band centered ~400 bp, you could use 0.7X beads and then take the supernatant. This would include all the DNA fragments that were smaller than ~500bp and didn't bind to the beads. Then mix the supernatant with 0.9X beads and everything larger than ~300bp would bind. Elute the bound DNA and you'll have a fairly broad, but good enough to sequence, library.
    If you try the beads, there can be a bit of lot-to-lot variability, so I would test them with a 100 bp library as I did in this gel.

    I haven't used the Truseq kit, but I have used the Truseq adapters with outsourced enzymes/buffers and the same problem can occur.
    I am sorry about the Necro Post, but I have a problem in this moment about this. Could you tell me the weigth (in bp) of your 1 kb ladder? It would be very useful for my thesis. Thanks a lot.

    Leave a comment:


  • Heisman
    replied
    Originally posted by odile View Post
    I need to remove fragments > 500bp and already removed everything <100bp (I basically have a smear that starts around 100-150bp). I would like to try to use 0.7x (from bluescript gel) and keep the supernatant.
    What would you do next? Clean-up the supernatant with Qiaquick or MinElute or add 1.8x beads and start all over again?
    You can add 1.8x beads and it will work. The idea is that you need to get the PEG concentration above whatever the threshold is for the size you care about. You could actually get away with less than 1.8x beads since you already have 0.7x bead solution, but it won't hurt to add more.

    Leave a comment:

Latest Articles

Collapse

  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM
  • SEQadmin2
    New Genomics Technologies Take Aim at Long-Standing Limits
    by SEQadmin2


    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
    ...
    09-28-2026, 10:25 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, Yesterday, 11:13 AM
0 responses
18 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, Yesterday, 10:48 AM
0 responses
18 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-15-2026, 10:52 AM
0 responses
36 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-14-2026, 05:39 AM
0 responses
42 views
0 reactions
Last Post SEQadmin2  
Working...