  • james hadfield
    Moderator
    Cambridge, UK
    Community Forum
    • Feb 2008
    • 224

    #1

    High phasing on GAIIx and HiSeq

    Has anyone seen high phasing rates recently?

    We have seen phasing of over 0.4% and up to almost 0.8% recently on several runs across both GAIIx and HiSeq.

    This has been evident in paired end runs where one read is fine but the other is poor.

    I believe this could be due to chemistry. I would like to find out if others have experienced similar issues over the past few weeks?

    Feel free to contact me directly if you prefer.
  • hoisinjl
    Member
    • Jun 2009
    • 23

    #2
    When was the last time a maintenance had been done? We've seen these high numbers when either an incorrect control lane was used or it had been too long between maintenance washes.

    Comment

    • james hadfield
      Moderator
      Cambridge, UK
      Community Forum
      • Feb 2008
      • 224

      #3
      Our instruments always get the required maintenance washes. This is unlikely to be the case as we have seen one read in a pair having higher pahsing than the other where a 200cycle SBS kit is used for the first read and a separate on for the second on a pair of flowcells.

      Comment

      • huguesparri
        Member
        • May 2008
        • 97

        #4
        We faced a high phasing/prephasing problem once. Illumina's replaced our kits and gave us the following advices:
        Here are some guidelines for the future:
        Handling properly the reagents is essential for avoiding phasing/prephasing problems.
        Therefore the best practices are:
        - Store SBS kits with the plastic bagging intact. Avoid warming, shaking or aerate the reagents.
        - Keep ICR buffer at 4ºC (or frozen at -20C) and LPR and LFN should be added immediately before loading on the sequencer. ICR should be tightly closed whenever possible and ICR and Incorporation Mix should never be stored at temperatures above 8ºC.
        - LFN should be thawed in room temperature water for 20 min; CMR should be thawed in the fridge overnight and thawing should be finished in room temperature water for not more than one hour and, as you know, in a different container as any other reagent.
        4) Maintenance Washes (water-NaOH-water) are absolutely essential after each run and the waste volumes from each phase should be checked.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          Yesterday, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        20 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        71 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...