Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Liam_Gallagher
    Member
    • Oct 2011
    • 18

    #1

    UPS (Uninterruptible Power Supply) with Ion Torrent

    Hi all!
    Does anyone of you use an UPS for preventing power failures ? If yes...which one do you use ?
    I'm preparing the lab for the installation of the Ion Torrent PGM and Torrent Server and I have a lot of doubts about the required equipment....thanks!!
    Cristian
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    If possible contact the sales rep and ask for a pre-installation guide. Applied Biosystems always has these available for their instruments. They will probably have a recommended UPS.

    You might want to use a different one, but at least you have some specifications to start with.
    --
    Phillip

    Comment

    • Liam_Gallagher
      Member
      • Oct 2011
      • 18

      #3
      Originally posted by pmiguel View Post
      If possible contact the sales rep and ask for a pre-installation guide. Applied Biosystems always has these available for their instruments. They will probably have a recommended UPS.

      You might want to use a different one, but at least you have some specifications to start with.
      --
      Phillip
      Thank you Phillip!
      Yes, we alredy have the "Site Preparation Guide" form Life Technologies...but in this user manual it is only specified that an UPS (from Multiple Suppliers) is required and should be compatible with 1500W output or greater for 1 Torrent Server and 1 Torrent PGM Sequencer.
      So...I am just wondering wich one (brand or model) you are using with Ion Torrent.
      Thanks,

      Cristian

      Comment

      • Apollo704
        Junior Member
        • Jul 2009
        • 6

        #4
        We are not using a UPS with our sequencers. The run time being as short as it is, we haven't been as concerned with losing power during a run as we would be with a Hi-seq.

        Comment

        • rbridenbecker
          Junior Member
          • Nov 2011
          • 5

          #5
          A UPS also provides protection from power surges and brownouts that can kill electronics so there is value beyond just battery backup.

          The Torrent Server has an optional SW module to shutdown the server if the power goes out, so I'd make sure the Torrent Server UPS is compatible. There is some info in the User Manual. --rob

          Comment

          • ptran
            Junior Member
            • Jun 2011
            • 3

            #6
            We are using an APC SMT2200

            Torrent Server (Optiplex) 115V*10A =~1150VA
            Ion Torrent PGM 110V *9A = ~990 VA

            total ~ 2140 VA

            Comment

            • Liam_Gallagher
              Member
              • Oct 2011
              • 18

              #7
              Thank you all!
              We just bought the same one suggested by ptran.
              It also should be compatible with the SW module that shutdown the server if the power goes out.

              Cristian

              Comment

              • lek2k
                Member
                • Aug 2011
                • 32

                #8
                We have ours connected to the UPS/Backup power point supplied by the Hospital we are based in. Unfortunately there is a split second delay before it kicks in which causes the Torrent Server to reboot or something. To compound matters they test this backup once a month! Anyways at least it's not a dirty shut down which takes fdsk (or whatever it's called) several hours to check the drives.

                Comment

                • HMorrison
                  Senior Member
                  • May 2009
                  • 121

                  #9
                  Ours is from Franek Technologies--we have an emergency generator, so this is basically to protect it during the transition.

                  FT1-IONPGM-IT Certified Laboratory Power Protection System for an Ion Torrent PGM
                  Sequencer and WorkStation (computer & monitor)

                  Pluggable Unit - Input & Output: Rated and configured for the above listed
                  instrument.
                  Input, 120 VAC: Provided with a NEMA 5-15P plug on a 6' power cord.
                  Output, Correct voltage, power and receptacles to match the Ion Torrent
                  PGM Sequencer and WorkStation (computer & monitor)
                  Maximum Envelope dimensions in: 14.7"H x 7.6" W x 18.9" D
                  Maximum installed weight 69 lb.
                  Backup time has been optimized for laboratories that are equipped with an
                  emergency generator system with an Automatic Transfer Switch (ATS).
                  Typical Backup time: Approximately 6 minutes minimum at full
                  instrumentation load.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                    by SEQadmin2




                    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                    10-30-2026, 06:46 AM
                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-28-2026, 09:57 AM
                  0 responses
                  45 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-28-2026, 09:51 AM
                  0 responses
                  26 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 11:13 AM
                  0 responses
                  74 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 10:48 AM
                  0 responses
                  49 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...