Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • adler89
    Junior Member
    • Apr 2012
    • 1

    #1

    Bioanalyzer and a pippen prep necessary?

    Hi guys,
    we are currently thinking about purchasing the Ion PGM and but Ion torrent recommend to use an Bioanalyzer and a pippen prep for library preparation. But the stuff costs an arm and a leg, do you really need this stuff or are there cheap alternatives?

    Thank you in advance
    Chris
    Last edited by adler89; 04-09-2012, 03:08 AM.
  • jonathanjacobs
    Member
    • Apr 2011
    • 23

    #2
    It actually an either/or question - not both.

    The idea behind using the BioAnalyzer is that you can assess your library's quality far far better than using a gel. When we first started running the PGM, using the BioAnalyzer was critical and we saved ourselves from wasting some runs by checking the samples beforehand. However, quite honestly, now that we have run the PGM a few dozen times it rarely is the case that our library preps fail to pass QC.

    So I guess, IMHO, it's not required - but a really good idea to have one-especially at the start. Plus, a BioAnalyzer can. E used for all kinds of other really useful assays in the lab, so it's not something that should be thought of as solely for the PGM.
    @bioinformer
    http://www.linkedin.com/in/jonathanjacobs

    Comment

    • SarahNGS
      Member
      • Nov 2011
      • 17

      #3
      not an either/or

      Actually pippen prep is for what I call "preparative scale" analysis. this machine and others like it (LabChip XT for example) are designed to recover the sample. It is the automatic equivalent of running your whole sample on a gel, cutting out the bands, and recovering the DNA from the gel slice. As such it is not an absolute requirement. It really depends on how many samples you will be doing a day or week and if sample prep becomes a bottleneck.
      The Bioanalyzer and its cousin LabChip GX are analytical scale tools to measure DNA size distribution. you could also run a gel for this but a gel is not as sensitive or informative. the size cutoffs for PGM are pretty strict and you need to be sure your preps are going to give you good results. I'd say it is a must-have.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        I agree with SarahNGS, if you have the means, get the BioAnalyzer. If you don't, then you can, in principle, get by without one. Just go into "blind zen archer" mode until after an amplification. Then you can run an agarose or acrylamide gel and see whether you hit your target or not.

        The Pippinprep is cool but, in practice, we really don't use it much. You can tinker around with double-sided AmPure to do most of what the Pippinprep does.

        --
        Phillip

        Comment

        • JamesH
          Member
          • Aug 2010
          • 19

          #5
          Has anyone tried this:

          The MultiNA II MCE-301 is a microchip electrophoresis system for streamlines genetic analysis workflows by fully automating nucleic acid electrophoresis—a process that typically requires over two hours of manual work.


          instead of the bioanalyzer or labchip?

          Runs are cheaper but does anyone know if it's any good?

          Comment

          • Sandokhan
            Junior Member
            • Oct 2011
            • 5

            #6
            We also did not buy them, since we are already without arms or legs to sell. We are just starting and have managed to get results without them, besides, reagents for them will cost you probably an eye.

            Comment

            • thomasblomquist
              Member
              • Jul 2012
              • 68

              #7
              We're using a bioanalyzer for analysis, and the invitrogen e-gel system for library fractionation (about $1 per sample) with very good success.

              Comment

              • HMorrison
                Senior Member
                • May 2009
                • 121

                #8
                Originally posted by pmiguel View Post
                I agree with SarahNGS, if you have the means, get the BioAnalyzer. If you don't, then you can, in principle, get by without one. Just go into "blind zen archer" mode until after an amplification. Then you can run an agarose or acrylamide gel and see whether you hit your target or not.

                The Pippinprep is cool but, in practice, we really don't use it much. You can tinker around with double-sided AmPure to do most of what the Pippinprep does.

                --
                Phillip
                Phillip, can you give an example of a double-sided Ampure protocol? Thanks,
                Hilary

                Comment

                • snetmcom
                  Senior Member
                  • Oct 2008
                  • 159

                  #9
                  Originally posted by HMorrison View Post
                  Phillip, can you give an example of a double-sided Ampure protocol? Thanks,
                  Hilary
                  quite a few threads on this exist if you search this forum.

                  Comment

                  • HMorrison
                    Senior Member
                    • May 2009
                    • 121

                    #10
                    Originally posted by snetmcom View Post
                    quite a few threads on this exist if you search this forum.
                    I just did. Not finding a protocol, although I did find a posting where selection was for 220 (170-220). I am looking for the equivalent of a gel cut. Never mind. I will return to the search when I have more time to guess key words.

                    Comment

                    • ECO
                      --Site Admin--
                      • Oct 2007
                      • 1360

                      #11
                      Originally posted by HMorrison View Post
                      Phillip, can you give an example of a double-sided Ampure protocol? Thanks,
                      Hilary
                      You will almost certainly have to test this in your own hands and optimize the sizes for your results, but one example would be to first use a very low ratio of AMPure :: Sample, for example 0.5x-0.6x, which should bind everything ~300 and larger to the beads. Then the supernatant is saved, and additional AMPure is added to the supernatant to increase the AMPure::Sample ratio to something that will bind fragments larger than ~150, like 1.3x.

                      In the second step you have your desired fragments on the beads, having effectively performed a gel free size cut. Not as clean as a gel, but infinitely more automatable.

                      Comment

                      • HMorrison
                        Senior Member
                        • May 2009
                        • 121

                        #12
                        Originally posted by ECO View Post
                        You will almost certainly have to test this in your own hands and optimize the sizes for your results, but one example would be to first use a very low ratio of AMPure :: Sample, for example 0.5x-0.6x, which should bind everything ~300 and larger to the beads. Then the supernatant is saved, and additional AMPure is added to the supernatant to increase the AMPure::Sample ratio to something that will bind fragments larger than ~150, like 1.3x.

                        In the second step you have your desired fragments on the beads, having effectively performed a gel free size cut. Not as clean as a gel, but infinitely more automatable.
                        Thanks so much, Eric. We'll experiment with MW ladder and try to add it to the repertoire. In fact, we'd like to eliminate very large fragments as well as the ones under 300. Standard Ampure clean up tends to overenrich the undesirable large products!

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          Hi Hilary,

                          Some detail in:

                          Rodrigue S, Materna AC, Timberlake SC, Blackburn MC, Malmstrom RR, et al. (2010) Unlocking Short Read Sequencing for Metagenomics. PLoS ONE 5(7): e11840. doi:10.1371/journal.pone.0011840

                          and

                          Lennon N, Lintner R, Anderson S, Alvarez P, Barry A, Brockman W et al. (2010). A scalable, fully automated process for construction of sequence-ready barcoded libraries for 454. Genome Biol 11: R15.
                          doi:10.1186/gb-2010-11-2-r15


                          But I guess ECO is right -- we just recently got around to trying it ourselves. At this juncture it looks pretty easy to get a not very tight size distribution.

                          Not sure what the effect of Ampure on bubble products is, though.

                          --
                          Phillip

                          Comment

                          • snetmcom
                            Senior Member
                            • Oct 2008
                            • 159

                            #14
                            Originally posted by HMorrison View Post
                            I just did. Not finding a protocol, although I did find a posting where selection was for 220 (170-220). I am looking for the equivalent of a gel cut. Never mind. I will return to the search when I have more time to guess key words.
                            Ampure will not be equivalent to a gel cut. It's quite broad for size selection, but it is easy to automate.

                            Comment

                            • lzembek
                              Junior Member
                              • Dec 2009
                              • 5

                              #15
                              I have also tested Invitrogen's E-Gel® SizeSelect™ 2% Agarose and got similar results to PippipPrep.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Today, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              22 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              15 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              68 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...