    readlength of size selected transcriptomes to low

    We have done 4 runs of transcriptomes with size selection, and the read length is not like it should be.

    1. Our first few runs were without size selection, and the sequencing looks fine. We had the same sizes on the sequencer as on the Bioanalyzer.
    2. Because we are losing a lot of possible data with the short reads, we did a size selection of 200 bp on the above sample.
    The size selected material is from the tip of the tail of the Gaussian cuve, and not from the middle.
    Sequencing of this gives bad results : very short readlengths, most below 100 bp, size selection was done at 200 bp. We repeated this for an other sample with the same results.
    3. We did this for a new sample, size selection around 250 bp, this time, the tail of the curve went up to 500 bp, so the sample was still from the tail, but not the very end.
    sequencing was disappointing again (but better than the previous one, most reads below 150 bp) :
    4. To prove that sequencing of a 250 bp fragment is possible with the 200 bp kit, we did it also starting from gDNA, and that worked fine, most reads around 250 bp.

    So my question is : why does the sequencing works fine with DNA, and not so good with cDNA from transcriptomics ? Is the size-selected material from the tail problematic cDNA (structure, base sequence, repeats, ...).

    Anyone seen the same thing with size selected cDNA from the middle of the Gaussian cuve ?

    Best regards,
    Andy
  • snetmcom
    Senior Member
    • Oct 2008
    • 159

    #2
    I don't think it's a sequencing issue. There is no difference between cDNA and DNA libraries by the time it reaches the instrument. You probably have something else going on up stream. Do you have any library QC you could share?

    Comment

    • avierstr
      Member
      • Apr 2012
      • 18

      #3
      Hi,

      this is the FastQC report of one of the runs.

      Greetings,
      Andy
      Attached Files

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
        by SEQadmin2



        Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
        ...
        07-09-2026, 11:10 AM
      • SEQadmin2
        Cancer Drug Resistance: The Lingering Barrier to Rising Survival
        by SEQadmin2



        Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

        There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
        07-08-2026, 05:17 AM
      • GATTACAT
        Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
        by GATTACAT
        Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
        07-01-2026, 11:43 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, Today, 10:26 AM
      0 responses
      7 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-09-2026, 10:04 AM
      0 responses
      24 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-08-2026, 10:08 AM
      0 responses
      15 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-07-2026, 11:05 AM
      0 responses
      33 views
      0 reactions
      Last Post SEQadmin2  
      Working...