    Wrong library size reads

    Hi all,

    I've prepared a library using the protocol for 200 bp reads and checked by Bioanalyzer library size was 238 bp.
    Then, OT runs and PGM runs, but the readings obtained were too short (53 bp mean, 42 bp median and 28 bp mode).
    What happened?
    Maybe the emulsion was broken? Why the emulsion could be broken?

    I hope your answers!!!

    Thank you and have a nice day

    Soledad
  • Sharp_GeCKO
    Junior Member
    • Apr 2014
    • 1

    #2
    How long is your target sequence?
    I guess that your total length of A adaptor+barcode+target fragment is nearly 53bp. So the rest 238-53=185bp might be the same sequence. When PGM runs, it can not recognize and correct the bases if all the cores on the chip are the same bases. It is just like that all the lights shining at the same time, then you can not see anything.

    To solve this problem, you should redesign your primer by adding 1-9bp variable length sequence between your A adaptor and barcode.

    Comment

    • snetmcom
      Senior Member
      • Oct 2008
      • 159

      #3
      Originally posted by Sharp_GeCKO View Post
      How long is your target sequence?
      I guess that your total length of A adaptor+barcode+target fragment is nearly 53bp. So the rest 238-53=185bp might be the same sequence. When PGM runs, it can not recognize and correct the bases if all the cores on the chip are the same bases. It is just like that all the lights shining at the same time, then you can not see anything.

      To solve this problem, you should redesign your primer by adding 1-9bp variable length sequence between your A adaptor and barcode.
      This is incorrect. I sequence fragments with homologous bases for the first 30+ bases all the time.

      It sounds more like poor quality sequencing or an issue with the library.

      Comment

      • MrGuy
        Member
        • Mar 2009
        • 68

        #4
        How did you do your size selection?
        How did you quantify you library prior to OT2?
        Are all your sequencing reagents fresh and from the same package?

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          Today, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        24 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        15 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        68 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...