Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • gringer
    David Eccles (gringer)
    • May 2011
    • 845

    #16
    What is the current consensus accuracy for PacBio?

    Comment

    • AllSeq
      Registered Vendor
      • Oct 2013
      • 138

      #17
      PacBio claims >99.999% (which I think is based on 50X coverage) due to their random errors. If ONT has random errors (or can get 'differently biased' errors from more than one pore type), then they should be able to have a very high consensus error rate as well. But I still haven't seen a clear claim of a random error model from ONT.
      AllSeq - The Sequencing Marketplace
      [email protected]
      www.AllSeq.com

      Comment

      • gringer
        David Eccles (gringer)
        • May 2011
        • 845

        #18
        Consensus is not that good with ONT at the moment, mostly because the base calling model has consistency issues with particular pentamers (with AAAAA being a particularly good/poor example) -- we were only able to manage 99% with our Influenza genome.

        I still maintain that this is a software issue, and that by throwing enough excited programmers at a proper base caller, we'll be able to re-call old reads and demonstrate that the accuracy was there all along.

        Comment

        • Brian Bushnell
          Super Moderator
          • Jan 2014
          • 2709

          #19
          I think the most important advantage of PacBio over Nanopore is the ability to do consensus for individual reads. This is fundamental and will not change with future iterations of either platform.

          There is no single "consensus accuracy" for PacBio. For all-to-all mapping, the more data, the more accurate... and for shorter inserts, the more passes, the more accurate. Considering reads of insert, which are only corrected via self-consensus, the longer the raw read length, the more accurate it will be. And the raw length is increasing rapidly with new chemistry advances.

          Both Nanopore and PacBio can do all-to-all mapping for error-correction. In this situation PacBio has a huge advantage in that the errors are more randomly distributed. But Nanopore can only do self-correction with 2 passes. PacBio can do self-correction with an arbitrary number of passes, proportional to the movie length versus insert size. Therefore, things like accurate full-length 16S sequencing are already possible for PacBio, but will never be possible for Nanopore.

          Comment

          • gringer
            David Eccles (gringer)
            • May 2011
            • 845

            #20
            Isn't that just a chemistry thing though? PacBio is a sequencing-by-synthesis method after all, and I don't see any reason why Circular Consensus Synthesis couldn't be done on nanopore as well by generating the sequence prior to putting it through the sequencer.

            Comment

            • Brian Bushnell
              Super Moderator
              • Jan 2014
              • 2709

              #21
              Originally posted by gringer View Post
              Isn't that just a chemistry thing though? PacBio is a sequencing-by-synthesis method after all, and I don't see any reason why Circular Consensus Synthesis couldn't be done on nanopore as well by generating the sequence prior to putting it through the sequencer.
              No... nanopore sucks the reads through like spaghetti. They are read once. 2D reads are achieved by reading a sequence and its reverse-complement. There is no provision for reading any sequence more than twice, because they are not circular.

              If there was some way to make replicates of a read linked together in a giant molecule, then yes, Nanopore would be able to make CCS reads. But such a thing is not possible today, or in any claimed future developments.

              Comment

              • gringer
                David Eccles (gringer)
                • May 2011
                • 845

                #22
                Originally posted by Brian Bushnell View Post
                If there was some way to make replicates of a read linked together in a giant molecule, then yes, Nanopore would be able to make CCS reads. But such a thing is not possible today, or in any claimed future developments.
                How about attaching hairpins to each end of a double-stranded sequence, then using a strand-displacement polymerase to bind to a primer sequence and run around in a loop, generating a single long sequence that repeats the initial template multiple times:

                I am a fan of the Pacific Biosciences PacBio RS sequencing instrument, as some of the readers of this blog maybe already are aware of. We use data from this instrument in our work, and the Norwegia…

                Comment

                • GenoMax
                  Senior Member
                  • Feb 2008
                  • 7142

                  #23
                  According to GenomeWeb, Sequel has a list price of $350,000, half that of a RSII (and a physical footprint a third the size of a RSII).

                  Comment

                  • Chipper
                    Senior Member
                    • Mar 2008
                    • 323

                    #24
                    Originally posted by Brian Bushnell View Post
                    If there was some way to make replicates of a read linked together in a giant molecule, then yes, Nanopore would be able to make CCS reads. But such a thing is not possible today, or in any claimed future developments.
                    RCA and LAMP is certainly possible today. Maybe not practical, or needed if accuracy goes up.

                    Comment

                    • ymc
                      Senior Member
                      • Mar 2010
                      • 496

                      #25
                      Originally posted by gringer View Post
                      This is very close to the current cost of MinION sequencing. It wouldn't surprise me if that weren't a coincidence.

                      A great MinION run will currently put out about 1Gbp of sequence, but that will change substantially after fast mode kicks in (about 20x sequencing speed). I wonder how flexible PacBio are with their pricing for the chip and reagents.
                      But for de novo assembly, only ONT 2D reads can be used. That means usable throughput for de novo assembly is only about 20%. Therefore, even if it gets the 20GB upgrade, the usable throughput is only 4GB. If the pricing remains $1000 per box, it is still not competitive not to mention the lower accuracy.

                      Of course, ONT is the only game in town for the portable market. I think that will be where its niche is until it can improve accuracy significantly.

                      Comment

                      • gringer
                        David Eccles (gringer)
                        • May 2011
                        • 845

                        #26
                        Originally posted by ymc View Post
                        But for de novo assembly, only ONT 2D reads can be used. That means usable throughput for de novo assembly is only about 20%.
                        Your information is a little old. The new reagent kits get about 80% 2D reads by using a sample preparation process that binds the hairpin to streptavidin beads.

                        While 1D reads are not particularly useful now, I have hope that a bright mathematician or computer scientist from outside ONT will substantially improve the base calling model. That will allow all those "useless" 1D reads to be retroactively recalled with much improved accuracy in the future.

                        Comment

                        • gringer
                          David Eccles (gringer)
                          • May 2011
                          • 845

                          #27
                          Originally posted by ymc View Post
                          If the pricing remains $1000 per box, it is still not competitive
                          It's currently $500 USD per flow cell if ordering in bulk, with reagent cost ~$200 per run, so basically the same marginal cost as the Sequel (ignoring initial equipment cost).

                          There's a new flow cell design coming out in the next 6-12 months with six times the number of pores and a substantially lower marginal cost, currently projected to be $20 USD for a short 3-hour run producing 2Gbp. I expect that would mean about $100 to run the flow cell for as long as it can go.
                          Last edited by gringer; 10-09-2015, 05:41 AM.

                          Comment

                          • nucacidhunter
                            Jafar Jabbari
                            • Jan 2013
                            • 1250

                            #28
                            Originally posted by gringer View Post
                            It's currently $500 USD per flow cell if ordering in bulk, with reagent cost ~$200 per run, so basically the same marginal cost as the Sequel (ignoring initial equipment cost).

                            There's a new flow cell design coming out in the next 6-12 months with six times the number of pores and a substantially lower marginal cost, currently projected to be $20 USD for a short 3-hour run producing 2Gbp. I expect that would mean about $100 to run the flow cell for as long as it can go.
                            In most places to get funding for a project one must estimate costs based on available technologies with predictable outputs. Cost estimation based on promised outcomes from technology providers with unknown real costs would result in rejection by fund providers. ONT currently has no definitive output and quality specs making it interesting to some bioinformaticians who see a fresh ground for software developments. It is also notable that non-optical based sequencing technologies has been less successful so far.

                            Comment

                            • Geneus
                              Member
                              • Dec 2010
                              • 60

                              #29
                              I was at ASHG on Wednesday and Thursday and the PacBio booth sure was VERY busy.

                              Comment

                              • gringer
                                David Eccles (gringer)
                                • May 2011
                                • 845

                                #30
                                Originally posted by Brian Bushnell View Post
                                If there was some way to make replicates of a read linked together in a giant molecule, then yes, Nanopore would be able to make CCS reads. But such a thing is not possible today, or in any claimed future developments.
                                Following on from this, there has been a recent preprint demonstrating a very similar approach for ONT sequencing:

                                Nanopore sequencing provides a rapid, cheap and portable real-time sequencing platform with the potential to revolutionize genomics. Several applications, including RNA-seq, haplotype sequencing and 16S sequencing, are however limited by its relatively high single read error rate (>10%). We present INC-Seq (Intramolecular-ligated Nanopore Consensus Sequencing) as a strategy for obtaining long and accurate nanopore reads starting with low input DNA. Applying INC-Seq for 16S rRNA based bacterial profiling generated full-length amplicon sequences with median accuracy >97%. INC-Seq reads enable accurate species-level classification, identification of species at 0.1% abundance and robust quantification of relative abundances, providing a cheap and effective approach for pathogen detection and microbiome profiling on the MinION system.


                                [I hope the authors don't mind sharing links to this]

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                  by SEQadmin2




                                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                  10-30-2026, 06:46 AM
                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-28-2026, 09:57 AM
                                0 responses
                                45 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-28-2026, 09:51 AM
                                0 responses
                                28 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                76 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                49 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...