Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • ymc
    Senior Member
    • Mar 2010
    • 496

    #1

    P6C4 has higher error rate than P5C3?

    Dear all,

    I tried to compare the P6C4 and P5C3 chemistries.

    P6C4 E coli K12 raw data from:
    The DevNet project on github stores the PacBio DevNet website. - PacificBiosciences/DevNet

    P5C3 E coli K12 raw data from: (referred by the PBcR site)


    P6C4 P5C3
    Total 780M 374M
    Coverage 168x 80x
    avgReadLen 11074 8682
    stdevReadLen 7107 5394
    longestRead 41390 28647

    I ran them through the PBcR-quiver pipeline and then use QUAST to evaluate the results:

    PBcR-quiver(P6C4) PBcR-quiver(P5C3)
    #misassemblies 5 3
    #local_misassemblies 2 2
    duplication_ratio 1.003 1.0
    bases_not_covered 93 5756 (reference has 4641652bp)
    #mismatches 4 2
    #indels 27 3 (indels is the number of indels not number of indels in basepairs)
    #genes 4513+3p 4508+2p (total 4516 genes, p means partial)

    We can see that the coverage is a lot better most likely due to longer read length and higher throughput. But the errors also increased significantly even though the throughput doubled. What do you think was the cause? Different labs perhaps? What's your experience with the two chemistries?
  • rhall
    Senior Member
    • Aug 2012
    • 324

    #2
    Are the assemblies complete? i.e. circular genomes? The Quast metrics are not comparable unless both assemblies are complete, and the genomes have been correctly circularized.
    P6C4 is generally higher accuracy, and higher consensus accuracy at lower coverage, but it is possible to overload a P6C4 cell degrading the data quality. P6C4 has been shown to perform better in all metrics than P5C3, witch is no longer available.

    Comment

    • ymc
      Senior Member
      • Mar 2010
      • 496

      #3
      Thank you for your reply. What do you mean by a complete assembly?

      I see that my assembly has only one contig and with no N's. I checked it with dotplot and didn't see duplication near the ends, so I didn't circularize it using the method suggested by PacBio. Is this a complete assembly?

      Comment

      • rhall
        Senior Member
        • Aug 2012
        • 324

        #4
        If you see no duplication at the ends the assemblies did not result in circular genomes. Did both P5 and P6 generate single contigs? The discrepancy in the number of bases not covered makes me think that the P5 assembly is not complete.

        Comment

        • ymc
          Senior Member
          • Mar 2010
          • 496

          #5
          Originally posted by rhall View Post
          If you see no duplication at the ends the assemblies did not result in circular genomes. Did both P5 and P6 generate single contigs? The discrepancy in the number of bases not covered makes me think that the P5 assembly is not complete.
          The P5C3 assembly results in one single contig of 4636871bp and I didn't see duplications in dotplot.

          The P6C4 assembly results in two contigs. One is 4656910bp and one is 36612bp. The smaller one has duplication at the ends but the long one is normal. Can the small one be a plasmid? I generated the above stat by taking out the smaller contig.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
            by SEQadmin2




            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
            10-30-2026, 06:46 AM
          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-28-2026, 09:57 AM
          0 responses
          38 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-28-2026, 09:51 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          72 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          48 views
          0 reactions
          Last Post SEQadmin2  
          Working...