We are a core facility that receives samples from multiple labs. We have one lab in particular that submits which are ran on a 3730XL. We are trying to determine what is happening with these particular results. The first time that we run their plate the results don't look particularly good but if we put that same plate back into the ABI we see a much better signal. I've spoken with another facility who thought it was due to some contamination. I don't really know where this would be coming from as we run the bigdye reaction and cleanseq on all the plates and it's only seen with this particular lab's plates. I would really like to try to help troubleshoot this issue for them. I have attached images of the first and second run. Subsequent plates look pretty good on the first run but could also be improved on the second run. These are also being ran back to back.
Seqanswers Leaderboard Ad
Collapse
Announcement
Collapse
No announcement yet.
X
-
If the second run from the same plate is better, also the signal is declining with reads length, than it is salts being injected instead of labelled DNA fragments on the first run.
I would suggest to use 200ul of 70% ethanol during the ethanol wash stage (if using ethanol precipitation in plates for sequencing reaction products purification). If doing precipitation in 1.5 ml tubes - than use 1ml of 70% ethanol on the wash step.
- Likes 1
Latest Articles
Collapse
-
by seqadmin
The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...-
Channel: Articles
04-22-2024, 07:01 AM -
-
by seqadmin
Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...-
Channel: Articles
04-04-2024, 04:25 PM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, Today, 11:49 AM
|
0 responses
11 views
0 likes
|
Last Post
by seqadmin
Today, 11:49 AM
|
||
Started by seqadmin, Yesterday, 08:47 AM
|
0 responses
16 views
0 likes
|
Last Post
by seqadmin
Yesterday, 08:47 AM
|
||
Started by seqadmin, 04-11-2024, 12:08 PM
|
0 responses
61 views
0 likes
|
Last Post
by seqadmin
04-11-2024, 12:08 PM
|
||
Started by seqadmin, 04-10-2024, 10:19 PM
|
0 responses
60 views
0 likes
|
Last Post
by seqadmin
04-10-2024, 10:19 PM
|
Comment