  • annaoc
    Junior Member
    • May 2011
    • 1

    #16
    Hi everyone. With the ScriptSeq mRNA-Seq Library Preparation Kit, would you recommend the MinElute or gel electrophoresis for library purification?
    Has anyone had experience of primer dimers after the PCR of 10 cycles? Cutting the band via size selection from a gel would perhaps then be the better choice. However, we do realise that by using the MinElute we can avoid exposing our sample to UV and Ethidium bromide and so we will probably obtain a much purer sample from MinElute as well as it being much a much easier protocol. Any comments on this would be greatly welcomed!
    Also, will our library still include the tRNA and small RNAs? Our bioanalyser run of the rRNA depleted RNA has shown that we have quite a lot of tRNA and we were wondering if they will be in the library then we can avoid them by size selection on the gel?
    Many thanks for your help!

    Comment

    • epibio
      Registered Vendor
      • May 2010
      • 89

      #17
      We recommend using 1.2X Ampure XP beads, if you want to get rid of the small amplicons. MinElute is not as good at doing this; you can expect ~10% adaptor-related reads with MinElute as compared to <3% with Ampure.
      Connect with Epicentre: Facebook | Twitter

      Comment

      • jagruh
        Junior Member
        • Apr 2011
        • 1

        #18
        epibio: Is ScriptSeq mRNA-seq library prep kit (for Illumina) strand specific? From the mechanism, it seems strand specific, but in our sequenced result there are still enormous hits to the other strand. We wonder whether these anti-transcripts are biological (in bacteria) or artifacts.
        Also, The distribution of hits within one gene is very uneven. It seems the PCR artifacts and/or non-random fragmentation inference a lot. Is there any available normalization parameters or tools available for this kit? And is there any published data prepared by this kit available in SRA database?
        Thank you!

        Comment

        • arabidopsis
          Member
          • Oct 2010
          • 13

          #19
          Dear all,

          I am analyzing the result of RNAseq on Hela cells. Libraries were created by Epicentre ScriptSeq kit, and they have around 7.000.000 reads each. I mapped the reads with TopHat to both hg18 and hg19. Most of transcripts look OK, but weird thing happens with RNA genes-snRNA, snoRNAs etc. Almost all reads cover reference RNAs starting from nucleotide +5 to +10, not from the annotated 5' end, even if the total coverage is very high. Out of more than 2000 reads mapped to U1 snRNA only 2 or 3 have their 5' end identical to the gene. Have anyone experienced this? Is there some technical issue with the library construction or sequencing, or a mapping artifact?
          Best,

          Aleks

          Comment

          • epibio
            Registered Vendor
            • May 2010
            • 89

            #20
            @arabidopsis: The ScriptSeq method is designed for mRNA-Seq, and is probably not the best fit for small RNAs. It may be partly a mapping problem as well; you could try allowing for more mismatches.

            @jagruh: Sorry I missed your post--for others who are interested, the ScriptSeq method is directional; however, we don't have data published to SRA as yet.
            Connect with Epicentre: Facebook | Twitter

            Comment

            • upendra_35
              Senior Member
              • Apr 2010
              • 102

              #21
              Originally posted by arabidopsis View Post
              Dear all,

              I am analyzing the result of RNAseq on Hela cells. Libraries were created by Epicentre ScriptSeq kit, and they have around 7.000.000 reads each. I mapped the reads with TopHat to both hg18 and hg19. Most of transcripts look OK, but weird thing happens with RNA genes-snRNA, snoRNAs etc. Almost all reads cover reference RNAs starting from nucleotide +5 to +10, not from the annotated 5' end, even if the total coverage is very high. Out of more than 2000 reads mapped to U1 snRNA only 2 or 3 have their 5' end identical to the gene. Have anyone experienced this? Is there some technical issue with the library construction or sequencing, or a mapping artifact?
              Best,

              Aleks

              Hi could you please provide me your parameters for mapping strand-specific reads using Tophat. I ask this because when i map my reads i can only see 0.5% reads mapped. Thanks in advance

              Comment

              • mmark27
                Junior Member
                • Nov 2011
                • 2

                #22
                Originally posted by arabidopsis View Post
                Dear all,

                I am analyzing the result of RNAseq on Hela cells. Libraries were created by Epicentre ScriptSeq kit, and they have around 7.000.000 reads each. I mapped the reads with TopHat to both hg18 and hg19. Most of transcripts look OK, but weird thing happens with RNA genes-snRNA, snoRNAs etc. Almost all reads cover reference RNAs starting from nucleotide +5 to +10, not from the annotated 5' end, even if the total coverage is very high. Out of more than 2000 reads mapped to U1 snRNA only 2 or 3 have their 5' end identical to the gene. Have anyone experienced this? Is there some technical issue with the library construction or sequencing, or a mapping artifact?
                Best,

                Aleks
                I know this thread is wicked old, but we're using the Scriptseq V2 kit and we see the same thing on our data when loaded into the Genome Browser. Everything is shifted slightly. Epibio didn't have an explanation of why.

                What are people using during TopHat for library type? fr-firststrand, fr-secondstrand????

                Comment

                • MPPFig
                  Junior Member
                  • Mar 2012
                  • 7

                  #23
                  libraries with Chloroplast rRNA contamination-Epicentre Ribo-Zero Kit

                  I used the ScriptSeq(TM) mRNA-Seq Library Preparation Kit v.2 to construct some Brachypodium libraries. mRNA was obtained using the Ribo-Zero Kit (Plant Leaf), it did not work so well, it left a 1300 nt fragment that appears to be chloroplast rRNA. Based on the Agilent High Sensitivity DNA assay my libraries have "spiky" look. The patterns might come from the fragmentation of that abundant 1300 nt seq. I am worried that I will end up wasting a lot of the reads by sequencing chloroplast rRNA. I am attaching the bioanalyzer images. Has someone experience this?
                  Attached Files

                  Comment

                  • diagen
                    Member
                    • Jun 2012
                    • 13

                    #24
                    I get the same spiky pattern on my RNA samples, that also come after RiboZero!
                    I suspect something wrong with the RiboZero step, because on RNAseq prep on total mRNA have no such pattern.
                    Importantly, the yield of Ribozero is only 5% of total RNA, leaving 95% to rRNA. Nevertheless, textbook studies estimate the rRNA cellular content about 50%!

                    I wonder, where is the rest of my precious sample? Stayed on Ribozero beads?
                    What kinds of other biases the RZ kit does???

                    Comment

                    • MPPFig
                      Junior Member
                      • Mar 2012
                      • 7

                      #25
                      Diagen,

                      could you post some images after the RiboZero step, and bioanalyzer traces of the spiky library prep. What organism are you working on? Epicentre thinks that my 1300 nt fragment that you see after RiboZero is some rRNA that is not present on they cocktail. I am working with them trying to figure this out.

                      I am taking the rest of my total RNA (my so... precious sample) and try to clean with the Ambion MicroPoly(A)Purist™ Kit but the amount of total RNA that this kit requires is a concern, what if I lose everything?

                      Comment

                      • diagen
                        Member
                        • Jun 2012
                        • 13

                        #26
                        Here are 2 typical profiles.
                        Attached Files

                        Comment

                        • MPPFig
                          Junior Member
                          • Mar 2012
                          • 7

                          #27
                          diagen,
                          I would talk to the epicentre tech support , one thing they did for me was to BLAST my rRNA sequences against their pool of RiboZero probes, that way they could assess how well their kit would work for your samples.

                          Comment

                          • MPPFig
                            Junior Member
                            • Mar 2012
                            • 7

                            #28
                            diagen,
                            I would talk to the epicentre tech support , one thing they did for me was to BLAST my rRNA sequences against their pool of RiboZero probes, that way they could assess how well their kit would work for your samples.

                            Comment

                            • ruwehann
                              Junior Member
                              • Apr 2012
                              • 1

                              #29
                              chloroplast 23S rRNA

                              Chloroplast 23S rRNA which is properly inserted into the ribosome gets cleaved at two position giving rise to fragments of typically 1.3, 1.0 and 0.5knt. For complete rRNA depletion one would need more than one probe for chloroplast 23S rRNA.

                              Comment

                              • diagen
                                Member
                                • Jun 2012
                                • 13

                                #30
                                My samples are from human cell line. Furthermore, the first step in ScripSeq is non-specific RNA degradation by heat and divalent cations. So in principle no discrete bands are expected, but they are there in samples after RiboZero.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, Yesterday, 10:52 AM
                                0 responses
                                8 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                19 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-13-2026, 09:56 AM
                                0 responses
                                16 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-09-2026, 11:14 AM
                                0 responses
                                27 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...