  • Bioanalyzer for ChIPed DNA

    Dear All,
    I have a question regarding bioanalyzer for ChIPed DNA.
    When I do bioanalyzer for the input DNA, I get peaks at the proper range in bioanalyzer.
    But, for the immunoprecipitated samples, I got no peaks at all or strange peaks. I even do not have the marker peaks!
    I should mention that the concentration of ChIPed DNA was as low as 0.3 ng/µl (measured by Qubit HS kit) and I used 1 µl for bioanalyzer. So, I expected very low signals in bioanalyzer, but literally got nothing.
    These samples were loaded in a same bioanalyzer chip used for some other library prepped samples and they all worked well, which shows that there was no problem in bioanalyzer chip or experiment.
    Can anyone help me how to explain this?
    I am totally confused and cannot figure out what I should do!
    In the attached you can find a gel photo of the sonicated chromatin as well as two bioanalyzer graphs of two different ChIPed samples.

    waiting for your replies...
    Click image for larger version

Name:	Ladder.png
Views:	1
Size:	27.0 KB
ID:	309065



    Click image for larger version

Name:	ChIPed Sample12.png
Views:	1
Size:	29.6 KB
ID:	309066

    Click image for larger version

Name:	gel photo after sonication.jpg
Views:	1
Size:	10.3 KB
ID:	309067
    Attached Files
    Last edited by zgene; 06-13-2016, 12:21 PM.

  • #2
    It seems that one of the wells was not loaded with marker and the strange peak in other well could be due to contaminants from immunoprecipitation step. I could not figure which Bioanalyser trace is from input DNA.

    Comment


    • #3
      Originally posted by nucacidhunter View Post
      It seems that one of the wells was not loaded with marker and the strange peak in other well could be due to contaminants from immunoprecipitation step. I could not figure which Bioanalyser trace is from input DNA.

      Dear nucacidhunter,

      Thanks for the reply.
      Marker was loaded in all the wells.
      I did not attach the Input traces there, there was only a gel photo for that.
      The Input traces are now attached here.
      Click image for larger version

Name:	input.png
Views:	2
Size:	36.8 KB
ID:	305104
      For the contaimnation, I should say that I used the columns available in the kit (Abcam high sensitivity ChIP kit) which includes 3 washing steps with ethanol after DNA binding to the column.
      Do you think that its better to replace this step with qiagen kit to get less contaminants?

      Comment


      • #4
        I would contact kit supplier before trying another column for clean up. It could be result of some inefficiency at steps prior to clean up. Sample showing no trace in BA should be rerun to make sure that it is reproducible.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Best Practices for Single-Cell Sequencing Analysis
          by seqadmin



          While isolating and preparing single cells for sequencing was historically the bottleneck, recent technological advancements have shifted the challenge to data analysis. This highlights the rapidly evolving nature of single-cell sequencing. The inherent complexity of single-cell analysis has intensified with the surge in data volume and the incorporation of diverse and more complex datasets. This article explores the challenges in analysis, examines common pitfalls, offers...
          06-06-2024, 07:15 AM
        • seqadmin
          Latest Developments in Precision Medicine
          by seqadmin



          Technological advances have led to drastic improvements in the field of precision medicine, enabling more personalized approaches to treatment. This article explores four leading groups that are overcoming many of the challenges of genomic profiling and precision medicine through their innovative platforms and technologies.

          Somatic Genomics
          “We have such a tremendous amount of genetic diversity that exists within each of us, and not just between us as individuals,”...
          05-24-2024, 01:16 PM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, Yesterday, 06:58 AM
        0 responses
        13 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 06-06-2024, 08:18 AM
        0 responses
        20 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 06-06-2024, 08:04 AM
        0 responses
        18 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 06-03-2024, 06:55 AM
        0 responses
        13 views
        0 likes
        Last Post seqadmin  
        Working...
        X