  • AmitChaurasia
    Member
    • Jan 2013
    • 13

    I had a Metagenomic 16s rRNA run on MiSeq v3 kit 2 x300 cycles. with good PF >89% but

    Hi All,

    I had a metagenomic run on MiSeq using v3 kits of 16s rRNA v3-v4 region. I have loaded 8pM pooled library with 10% PhiX spike-in. MiSeq run had 89% PF with 16 M raw reads but with very low q30 37.5%. I am surprised how come q30 score that too much low even having good PF of 89%.
    I also see a large variation in % reads representation for a few samples (21-samples) out of 257 samples. Could this result in low Q30 of MiSeq run, but how ?
    I am attaching a few SAV screenshots for reference.
    Thanks
    Attached Files
  • jdk787
    josh kinman
    • Apr 2014
    • 72

    #2
    It looks like your Q scores have a big drop at ~80bp which indicates that you have a lot of smaller than expected library fragments. Since you are sequencing v3-v4 libraries I would guess this is due to amplified primer dimer, most likely in the libraries that got the higher read representation which you should see when looking at the data.

    This wouldn't hurt the pass filter rate since that is calculated at read 25.
    Did you visualize these libraries before sequencing?
    Josh Kinman

    Comment

    • AmitChaurasia
      Member
      • Jan 2013
      • 13

      #3
      Hi Kinman,

      Thanks for your reply, I have taken random samples to check with Bioanalyzer and no short peaks were found. Also fastq files were checked to have 301 base length only.

      So this can be ruled out.

      Best,

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Originally posted by AmitChaurasia View Post
        Hi Kinman,

        Thanks for your reply, I have taken random samples to check with Bioanalyzer and no short peaks were found. Also fastq files were checked to have 301 base length only.

        So this can be ruled out.

        Best,
        The length of the reads in your .fastq file likely does not include low-quality base clipping. The MiSeq will happily call bases on background noise. So the length of the reads is not diagnostic for this purpose.

        Your random samples that were run on the bioanalyzer -- did they include any of the ~dozen samples that contributed >2% of reads? Those would be the likely culprits for providing large amounts of primer dimers.

        Even if these did not show primer dimers your samples may have included large amounts of primer-dimers annealed to full length library molecules. To detect these you would need to use some sort of denaturing electrophoretic size analysis. This could be denaturing PAGE, or if you prefer the Agilient Bioanalyzer, try:

        denature your samples for 3 minutes at 95oC and then "snap cool" them in an ice-water bath. Then run them on a denaturing agilent chip -- we use the RNA pico chip for this purpose. Note that during the heat denaturation step you need some method to prevent your sample from just evaporating -- use of a heated lid thermal cycler is good for this purpose.

        --
        Phillip

        Comment

        • AmitChaurasia
          Member
          • Jan 2013
          • 13

          #5
          Originally posted by pmiguel View Post
          The length of the reads in your .fastq file likely does not include low-quality base clipping. The MiSeq will happily call bases on background noise. So the length of the reads is not diagnostic for this purpose.

          Your random samples that were run on the bioanalyzer -- did they include any of the ~dozen samples that contributed >2% of reads? Those would be the likely culprits for providing large amounts of primer dimers.

          Even if these did not show primer dimers your samples may have included large amounts of primer-dimers annealed to full length library molecules. To detect these you would need to use some sort of denaturing electrophoretic size analysis. This could be denaturing PAGE, or if you prefer the Agilient Bioanalyzer, try:

          denature your samples for 3 minutes at 95oC and then "snap cool" them in an ice-water bath. Then run them on a denaturing agilent chip -- we use the RNA pico chip for this purpose. Note that during the heat denaturation step you need some method to prevent your sample from just evaporating -- use of a heated lid thermal cycler is good for this purpose.

          --
          Phillip
          Thanks a lot Phillip for your valuable inputs.

          I have gone through your suggestions and looked into the data in greater details wrt possiblity of primer-dimer or adapter dimers. I have checked the fastq files with fastqc software to look for over-represented sequences but could not find the adapter-dimers or primer dimer sequences in that. SO i believe I can rule out the contamination of these dimers. On the other hand I have also checked out for over-represented samples if any of them have BA traces, I found none of them. I am planning to get BA profile for a few over-represented libraries done to get 100% sure that if there is any contamination in them (is it necessary to do BA on RNA-chip?)

          I am attaching fastqc screenshot for your reference (same sequences are over-represented in all fastq files)
          Attached Files

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            You need to use a scan/trim program (I recommend bbduk.sh from BBMap suite) to scan and trim your data. You can't depend on FastQC to identify dimer contamination. Here is bbduk guide to get you started.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by AmitChaurasia View Post
              Thanks a lot Phillip for your valuable inputs.

              I have gone through your suggestions and looked into the data in greater details wrt possiblity of primer-dimer or adapter dimers. I have checked the fastq files with fastqc software to look for over-represented sequences but could not find the adapter-dimers or primer dimer sequences in that. SO i believe I can rule out the contamination of these dimers. On the other hand I have also checked out for over-represented samples if any of them have BA traces, I found none of them. I am planning to get BA profile for a few over-represented libraries done to get 100% sure that if there is any contamination in them (is it necessary to do BA on RNA-chip?)

              I am attaching fastqc screenshot for your reference (same sequences are over-represented in all fastq files)
              Those are just 16S sequences. Since FastQC only appears to be looking at the first 50 bases, I don't think it is a good assessment of the presence of primer dimers in your libraries. You could follow Genomax's advice. But it would be good to see the FastQC of the reverse read. If it is the reverse complement of that sequence then you either have a very short v-loop in the species you are sequencing, or you do have primer-dimers.

              Do you have a service contract? If so, just ask Illumina for reagents to do a phiX run. This would show if your MiSeq had deteriorated severely from its normal specifications.

              --
              Phillip

              Comment

              • thermophile
                Senior Member
                • Apr 2015
                • 243

                #8
                Phillip-what conditions have you seen primers annealed to full length amplicons? do you routinely heat denature then snap cool before cleaning amplicons?

                OP- I look at %base to see what is happening in my amplicon runs. Does it look like the %base from your previous runs (all of which will be way out of normal Illumina spec)
                Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by thermophile View Post
                  Phillip-what conditions have you seen primers annealed to full length amplicons? do you routinely heat denature then snap cool before cleaning amplicons?
                  Mostly from amplicon libraries submitted by customer labs. I'm speculating that they are primers and primer-dimers. But who knows? We have seen them in some Nextera XT libraries. And to less extents in other standard library types.

                  Heat denaturing prior to Ampure doesn't seem to impact the purification. Ampure for unknown reasons seems to be effective for removing small amounts of primers/primer-dimers annealed to full length library molecules, but not in cases where it looks like >50% of the molar contribution is from low molecular weight products.

                  Would be great to have a single-strand clean-up method that didn't require denaturing PAGE.

                  --
                  Phillip

                  Comment

                  • thermophile
                    Senior Member
                    • Apr 2015
                    • 243

                    #10
                    interesting. I've had a few library pools with a drop off after 50ish bases but couldn't see primer diamers on the tape station (I don't routinely run my amplicon libraries on the tape station, just the problems). I'll keep this in mind for the future.
                    Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                    Comment

                    • pmiguel
                      Senior Member
                      • Aug 2008
                      • 2328

                      #11
                      Originally posted by thermophile View Post
                      interesting. I've had a few library pools with a drop off after 50ish bases but couldn't see primer diamers on the tape station (I don't routinely run my amplicon libraries on the tape station, just the problems). I'll keep this in mind for the future.
                      Just keep in mind you would likely need to run the samples after heat denaturation/snap cooling on a denature chip -- eg, an RNA chip.

                      --
                      Phillip

                      Comment

                      • thermophile
                        Senior Member
                        • Apr 2015
                        • 243

                        #12
                        have you ever tried ampure cleaning after the denaturing step? That should work, right? Carboxyl groups would attract either single or double stranded. But maybe [PEG] would need to be changed?
                        Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          Originally posted by thermophile View Post
                          have you ever tried ampure cleaning after the denaturing step? That should work, right? Carboxyl groups would attract either single or double stranded. But maybe [PEG] would need to be changed?
                          Yes, we tried it, although not extensively. It worked, just no better than without denaturation.

                          Remember there is nothing to stop the denatured molecules from re-annealing--especially at the adapter ends unless you include a denaturation.

                          --
                          Phillip

                          Comment

                          • thermophile
                            Senior Member
                            • Apr 2015
                            • 243

                            #14
                            Originally posted by pmiguel View Post
                            Yes, we tried it, although not extensively. It worked, just no better than without denaturation.

                            Remember there is nothing to stop the denatured molecules from re-annealing--especially at the adapter ends unless you include a denaturation.

                            --
                            Phillip
                            good to know, thanks
                            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                            Comment

                            Latest Articles

                            Collapse

                            • seqadmin
                              Pathogen Surveillance with Advanced Genomic Tools
                              by seqadmin




                              The COVID-19 pandemic highlighted the need for proactive pathogen surveillance systems. As ongoing threats like avian influenza and newly emerging infections continue to pose risks, researchers are working to improve how quickly and accurately pathogens can be identified and tracked. In a recent SEQanswers webinar, two experts discussed how next-generation sequencing (NGS) and machine learning are shaping efforts to monitor viral variation and trace the origins of infectious...
                              03-24-2025, 11:48 AM
                            • seqadmin
                              New Genomics Tools and Methods Shared at AGBT 2025
                              by seqadmin


                              This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

                              The Headliner
                              The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
                              03-03-2025, 01:39 PM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by seqadmin, 03-20-2025, 05:03 AM
                            0 responses
                            49 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-19-2025, 07:27 AM
                            0 responses
                            57 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-18-2025, 12:50 PM
                            0 responses
                            50 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-03-2025, 01:15 PM
                            0 responses
                            200 views
                            0 reactions
                            Last Post seqadmin  
                            Working...