    Hi Xi,

    Thank you for the explanation. I'll try plotting the p-values. Another question. I have the absolute read counts for 4 male and 4 female drosophila flies. I tried both the MARS and the FET method on them. The results are similar (the FET method catches maybe 50-100 more genes than MARS) Is there a reason why u would use the MARS method over FET or vice-versa?

    Thanks
    Abhijit

    Comment

    • Xi Wang
      Senior Member
      • Oct 2009
      • 317

      #77
      Originally posted by gen2prot View Post
      Hi Xi,

      Thank you for the explanation. I'll try plotting the p-values. Another question. I have the absolute read counts for 4 male and 4 female drosophila flies. I tried both the MARS and the FET method on them. The results are similar (the FET method catches maybe 50-100 more genes than MARS) Is there a reason why u would use the MARS method over FET or vice-versa?

      Thanks
      Abhijit
      Sorry for the late reply.

      Its quite right that those methods give different results. This is because different methods build on different statistical models, and will reflect different aspects of statistical significance. The p-value derived from different methods are not comparable. If the gene ranks are the same, the tradeoff between sensitivity and specificity will be the same between two methods. Then only thing he needs to do it choose the appropriate threshold based on the empirical distribution of p-value.
      Xi Wang

      Comment

      • m!x
        Member
        • Aug 2009
        • 10

        #78
        Hi Xi,

        I got a question regarding the p-value generated by MARS method. Is it corrected for multiple comparisons by any methods, eg. Bonferroni correction?
        I have zero background about this statistical analysis and would appreciate if you could give some insight into this.

        Comment

        • Xi Wang
          Senior Member
          • Oct 2009
          • 317

          #79
          Originally posted by m!x View Post
          Hi Xi,

          I got a question regarding the p-value generated by MARS method. Is it corrected for multiple comparisons by any methods, eg. Bonferroni correction?
          I have zero background about this statistical analysis and would appreciate if you could give some insight into this.
          Hi m!x,

          Actually, in our package, for the methods MARS, FET, LRT, and MATR, both p-values and q-values are calculated. Q-value is the corrected p-value with respect to the multiple testing. Two kinds of q-values are given in DEGseq package, one is given by Benjamini correction, and the other by Storey correction. The two methods take different aspects for correction. The Benjamini correction gives more false negatives, while the Storey correction more false positives.

          More insight of multiple testing correction, please refer to http://en.wikipedia.org/wiki/Multiple_comparisons
          Xi Wang

          Comment

          • RSK
            Junior Member
            • Jun 2009
            • 9

            #80
            Default FDR

            Hi Xi,

            What is the default FDR % used for the Benjamini and Storey corrections for DEGexp? Also where can I find the vignette for the DEGseq package that member Simon Anders refers to in another post?

            Thanks,
            RSK

            Comment

            • Xi Wang
              Senior Member
              • Oct 2009
              • 317

              #81
              Hi RSK,

              The default FDR used for the Benjamini or the Storey correction is 0.001. The documents for DEGseq can be found at:
              DEGseq is an R package to identify differentially expressed genes from RNA-Seq data.


              Thanks.
              Xi Wang

              Comment

              • RSK
                Junior Member
                • Jun 2009
                • 9

                #82
                Thanks Xi!

                Comment

                • Davis McC
                  Member
                  • May 2010
                  • 16

                  #83
                  Gene/transcript length bias

                  Hi Adam

                  Originally posted by adamreid View Post
                  I guess I'm asking whether there is/ought to be a correction for gene length because more reads are expected to map to longer genes.
                  This is definitely something to be taken into consideration. See Oshlack and Wakefield, 'Transcript length bias in RNA-seq data confounds systems biology' (http://www.biology-direct.com/content/4/1/14) for further discussion.

                  The more we look into RNA-seq data, the more it seems that various normalization steps are required to get the most out of the data.

                  Cheers
                  Davis

                  Comment

                  • Xi Wang
                    Senior Member
                    • Oct 2009
                    • 317

                    #84
                    Originally posted by Davis McC View Post
                    Hi Adam



                    This is definitely something to be taken into consideration. See Oshlack and Wakefield, 'Transcript length bias in RNA-seq data confounds systems biology' (http://www.biology-direct.com/content/4/1/14) for further discussion.

                    The more we look into RNA-seq data, the more it seems that various normalization steps are required to get the most out of the data.

                    Cheers
                    Davis
                    Hi Davis,

                    You are quite right that the gene length and also expression levels affect the statistical power of detecting DEGs. Goseq provides a good example on how to solve this problem. Please note that the normalization itself cannot completely eliminate the biases.
                    Xi Wang

                    Comment

                    • apratap
                      Member
                      • Jan 2009
                      • 58

                      #85
                      Just following up on this educating thread on RNA-Seq calculation.

                      Is it fine to normalize by gene length or should we use transcript length( i.e sum of exons) and NOT sum of (exons+introns) for a gene

                      Thanks !
                      -Abhi

                      Comment

                      • Xi Wang
                        Senior Member
                        • Oct 2009
                        • 317

                        #86
                        Originally posted by apratap View Post
                        Just following up on this educating thread on RNA-Seq calculation.

                        Is it fine to normalize by gene length or should we use transcript length( i.e sum of exons) and NOT sum of (exons+introns) for a gene

                        Thanks !
                        -Abhi
                        Hi Abhi,

                        If you just want to identify differentially expressed genes, as is suggested in the DEGseq manuscript, you'd better not normalize the raw read counts. The reason is that the hypothesis tests (including Fisher's exact test, likelihood ratio test, and our MARS) are based on the random sampling model, and this model requires the raw counts as input to correctly calculate p-values. You are right that, according to the known gene annotation, only reads in the exonic regions should be counted.
                        Alternatively, if you want to estimate gene expression levels, you can refer to the concept RPKM. A newly published NBT paper made great efforts to clarify the correct way to estimate gene expression.
                        Hope this helps.
                        Xi Wang

                        Comment

                        • David Lyon
                          Member
                          • May 2010
                          • 11

                          #87
                          Hi Xi or anyone

                          I have the tophat and cufflinks output files from mapping RNA seq against the annotated genome from different tissues. As I am a newbie can someone provide the recommended method in a step by step guide how to use the results for input into DEGseq and process them by DEGseq to get the differential expression from the different tissues.

                          Thank you

                          Comment

                          • Xi Wang
                            Senior Member
                            • Oct 2009
                            • 317

                            #88
                            Hi David,

                            You can first use this script to convert the SAM files output by Tophat to BED files, and then feed the BED files to DEGseq. See the example of function DEGseq in the manual.
                            Any further questions, just let me know.
                            Xi Wang

                            Comment

                            • Davis McC
                              Member
                              • May 2010
                              • 16

                              #89
                              DE Analysis pipeline

                              Hi David

                              I am one of the developers for the Bioconductor package edgeR , which is designed for carrying out differential expression analysis of count data (like RNA-seq). Check out the User's Guide for more details and case studies to provide examples on how to use the package.

                              Colleagues of mine suggest the following sort of steps to go from raw RNA-seq short read data from the raw fasta files, through to GO category testing.

                              It sounds like you would jump in somewhere in the middle, since you have output files already from tophat and cufflinks. NB: similar sort of approach to that suggested by Xi, just a little more fleshed-out.

                              Steps with required tools & files

                              To perform the entire analysis, the following steps and tools will be needed:

                              1. Get some short read RNA-seq data, for at least two different experimental conditions you wish to compare

                              2. Choose a reference to map against, and map your data using a short read mapper that outputs in SAM format. We tend to use bowtie. Other options are bwa, SOAP2, novoalign, shrimp.

                              3. Use SAMtools to convert SAM output into the binary BAM format, which is both smaller on disk and allows for fast indexing.

                              4. Summarize reads on the gene/transcript/exon level. We use the R platform with the Rsamtools and GenomicFeatures packages.

                              5. Calculate DE genes from counts summarized on the gene level. We use the R package edgeR, which we have developed, although there are other tools out there, such as DEGseq. edgeR can account for biological variation in the data (using a negative binomial model), separate biological from technical variation, produce an MDS plot, and conduct exact testing procedures.

                              6. Perform GO category testing on the results of the differential expression analysis, using the R package goseq.

                              Considerations for DE Analysis

                              You should find the visualization tools in DEGseq useful, but you may find that the Poisson/variable sampling rate model does not adequately model the variability in the data. Extra-Poisson variation (or overdispersion) is typical of RNA-seq data, especially if there is biological replication amongst your samples. If you only have technical replicates then this may not be an issue, but I would recommend running your data through edgeR to get some idea of the inter-library variability. If you have overdispersed data, then using a Poisson model will *drastically* overestimate the levels of differential expression in your data. Using a NB model like in edgeR can account for this extra variation in the data and give much better assessment of DE.

                              DEGseq links in quite well with edgeR, so you might look at your data first with DEGseq then use edgeR to deal with overdispersion in the data, investigate inter-library (incl. biological) variability and get exact p-values for DE based on the NB model.

                              Hope that is helpful and good luck with your data analysis. Please ask if you have any more questions I might be able to help with.

                              Best regards
                              Davis

                              Comment

                              • Xi Wang
                                Senior Member
                                • Oct 2009
                                • 317

                                #90
                                Thanks, Davis. You give us a very clear and detailed way to analyze RNA-seq data, especially if one want to compare genes’ differential expression in two experimental conditions. Here, I still have some minor points to complete.

                                1. To process RNA-seq short read data, one may be suggested to use Tophat or SpliceMap for mapping the short reads against the reference genome (e.g., the human genome). These tools have the ability to detect splicing junctions, even novel junctions. As mentioned by Daves, Bowtie is a very fast mapper, but if one wants to apply Bowtie directly to RNA-seq data, one should prepared junction sequences (put two near ends of adjacent exons together to form a new sequence), in order to detect potential junctions.

                                2. SAM format is recommended to save the read mapping results, as it can be used in general cases and reserves lots of information. If one only focuses on DE, we provide a very simple BED-like format for user to choose. We also provide a PERL script (if cannot access, contact me) to convert the general SAM format to the BED-like format. This BED-like format can be directly fed to DEGseq.

                                3. As DEGseq can directly deal with 6-colunm BED-like files, no others additional tools are required. The embedded C++ code is able to process read counting swiftly.

                                4. DEGseq can not only give users a list of DE genes, also provide figures showing the loci of DE genes in the M-A plot, which offers the researchers a direct visual impression on the DE genes identified.

                                5. In the DEGseq Bioconductor package, the function samWrapper, which imports the very popular SAM method in microarray studies, can deal with the RNA-seq data with biological replicates. Rather than the strategy involved in edgeR, samWrapper uses an empirical approach to estimate the FDR of DE gene identification.

                                6. For high level analyses, given the DE gene list by DEGseq, users may have very flexible options to choose tools according to their specific scientific questions. GO term enrichment, pathway enrichment, and gene set enrichment analyses are all under users’ choice. GOseq, specifically designed for RNA-seq (and other techniques may result in selection bias), is recommended.

                                Hope the above information helps. I’d like to help researchers in the community with data analysis or other problems that I am able to.

                                Cheers!
                                Last edited by Xi Wang; 05-27-2010, 07:02 PM.
                                Xi Wang

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                  by SEQadmin2



                                  CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                  Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                  ...
                                  07-31-2026, 11:01 AM
                                • SEQadmin2
                                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                  by SEQadmin2


                                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                  The systematic characterization of the human proteome has
                                  ...
                                  07-20-2026, 11:48 AM
                                • SEQadmin2
                                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                  by SEQadmin2



                                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                  ...
                                  07-09-2026, 11:10 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 08-03-2026, 10:13 AM
                                0 responses
                                15 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-31-2026, 02:55 AM
                                0 responses
                                32 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-24-2026, 12:17 PM
                                0 responses
                                23 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-23-2026, 11:41 AM
                                0 responses
                                21 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...