    RNASeq: Read length different from expected

    Hello all,

    I have received paired-end reads for 40 samples.
    The reads are supposed to be 100bp per end.
    Instead, 20 of my samples are 101bp per end, while the other 20 are 100bp as expected.

    Because of this, we assume that these 20 samples were all in the same lane; and that by accident there was an extra iteration in the illumina sequencing.
    However, we also see a strong negative correlation with read length and quality; the samples in which we had 101bp per end lose about 30% of the reads in the trimmomatic quality step.

    My question is: Has this happened to anyone else? Does it occur often, and if so, does it often affect quality? We are really quite puzzled by this.

    Any ideas / clues appreciated.

    Thank you!
    -Thies Gehrmann
  • mastal
    Senior Member
    • Mar 2009
    • 666

    #2
    RNASeq: Read length different from expected

    The read length of 101 bp instead of 100 bp would probably be because the operator thought there were enough reagents left to sequence the extra base.


    Presumably the run that gave the 101 bp reads had some problems that resulted in poorer base qualities than the run or lane that gave the 100 bp reads.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      That is odd. I do not think one can set up lanes on a single flowcell to run for different number of cycles.

      Have you checked with your provider to confirm if all the samples ran together (were they multiplexed) or if there was a batch difference (2 separate runs?) that can account for your observation. I think it would probably be the latter case.

      Comment

      • mastal
        Senior Member
        • Mar 2009
        • 666

        #4
        The read identifiers for each sample should tell you if they were on the same flow cell or lane.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Originally posted by mastal View Post
          The read identifiers for each sample should tell you if they were on the same flow cell or lane.
          Great point. This should be easy to check.

          Comment

          • kmcarr
            Senior Member
            • May 2008
            • 1181

            #6
            It is somewhat standard practice for Illumina sequencing that if you want read lengths of N bases you run N+1 cycles. This has to do with the way base calling works on Illumina; to properly call the base at position n in a read you need data from cycle n+1. The last base in a read will always have a lower Q-score reflecting the added uncertainty in the base call. To mitigate this you run N+1 cycles but just report N bases per read, dropping the last, low quality base. This practice is ingrained in the Illumina run recipes; the standard PE100 recipe (with indexing) on the HiSeq runs a 209 cycles (101 + 7 + 101) adding an extra cycle each to read 1, the index read and read 2. (Interestingly the MiSeq recipes still add the extra cycle to reads 1 & 2 but not to the index read.)

            Some core labs simply call and report all 101 cycles, some stick to the original practice of clipping the last base. It may be that your samples were all run together on the same flow cell for 2x101 cycles but for one set of 20 they reported all 101 cycles and the other they clipped the last base. You need to check the IDs of your samples to identify the flow cell and lanes used for each.

            Comment

            • gogodidi
              Junior Member
              • Aug 2012
              • 4

              #7
              Originally posted by mastal View Post
              The read identifiers for each sample should tell you if they were on the same flow cell or lane.
              Indeed, they were!
              Thank you (Don't know why it didn't occur to me).

              Comment

              • gogodidi
                Junior Member
                • Aug 2012
                • 4

                #8
                Originally posted by kmcarr View Post
                It is somewhat standard practice for Illumina sequencing that if you want read lengths of N bases you run N+1 cycles. This has to do with the way base calling works on Illumina; to properly call the base at position n in a read you need data from cycle n+1. The last base in a read will always have a lower Q-score reflecting the added uncertainty in the base call. To mitigate this you run N+1 cycles but just report N bases per read, dropping the last, low quality base. This practice is ingrained in the Illumina run recipes; the standard PE100 recipe (with indexing) on the HiSeq runs a 209 cycles (101 + 7 + 101) adding an extra cycle each to read 1, the index read and read 2. (Interestingly the MiSeq recipes still add the extra cycle to reads 1 & 2 but not to the index read.)

                Some core labs simply call and report all 101 cycles, some stick to the original practice of clipping the last base. It may be that your samples were all run together on the same flow cell for 2x101 cycles but for one set of 20 they reported all 101 cycles and the other they clipped the last base. You need to check the IDs of your samples to identify the flow cell and lanes used for each.
                Thank you for this information.
                It is strange then, that they cut it for one lane, and not for the other.
                I'll have to call them to ask about that.

                Thank you so much, all of you!

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                  by SEQadmin2



                  CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                  Despite this, “CRISPR helped turn genome editing from a specialized technique into
                  ...
                  07-31-2026, 11:01 AM
                • SEQadmin2
                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                  by SEQadmin2


                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                  The systematic characterization of the human proteome has
                  ...
                  07-20-2026, 11:48 AM
                • SEQadmin2
                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                  by SEQadmin2



                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                  ...
                  07-09-2026, 11:10 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 10:13 AM
                0 responses
                14 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-31-2026, 02:55 AM