    Problematic bioanalyzer results (QC after mRNA capture)

    Did mRNA capture (Dynabead) for my RNA extract samples (mushroom). When run mRNA pico assay for QC, these kind of result appeared. Sometime it work for some of the samples. But repeated assay using same sample seem to produce different result. Contacted field application specialist and made the listed changes but problem persists.

    Changes made:

    1. washed electrode with RNAzap and consequently with RNase free water and air dried for more than 10 minutes. Even diagnosed for short circuit before run.

    2. check priming station for failure. It seem to be working because it bounce normally when the latch was released

    3. my sample were treated with DNase from MoBio.

    4. diluted my samples.

    Attached here are pdf. containing results for last few run.

    2100 expert_mRNA Pico_DE13805702_2016-06-27_18-03-51.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-14_17-28-00.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-27_16-40-47.pdf
  • DStephens@NuGEN
    Registered Vendor
    • Jun 2016
    • 3

    #2
    Hi sbchua.1990,
    We have seen traces such as these when the chip priming/loading was not performed correctly. Did you ensure that the chip priming station plunger was set to the highest notch and the plunger was released after exactly 30 seconds? Was the chip loaded with fresh reagents?

    A couple of other possibilities include overloading of the chip and salt or detergent carryover. You noted that you already tried diluting the samples; how much material was used? We typically load 1-5 ng/well onto the Pico chip.

    Finally, have you tried an additional clean-up of your samples?

    Hope that helps!
    Denise Stephens
    Technical Support Scientist at NuGEN Technologies

    Comment

    • sbchua.1990
      Junior Member
      • Sep 2015
      • 3

      #3
      Thanks for reply, I am pretty sure that the chip priming/loading was performed correctly. Reagents are still within expiration date.

      My samples concentration should be around 300pg - 1500pg. I just loaded 1 ul without dilution.

      No I did not try additional clean-up for my samples.

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        The samples most likely are contaminated with residual removal resin. You would need to contact MoBio for cleanup instructions.

        Comment

        • sbchua.1990
          Junior Member
          • Sep 2015
          • 3

          #5
          Hi nucacidhunter, did you encounter such problem in the past? Or what make you think so? Thanks for your reply.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            Profile indicates presence of insoluble contaminants which could be protein (residual DNase) or removal resin. I suspect the resin because it can be carried over easily if spinning was not long enough or one tried to take the supernatant without leaving any liquid behind in the tube. There is also possibility of protein protectant carry over which is used to stabilise DNase at RT.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by SEQadmin2


              I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

              Here are nine questions we think about, in roughly the order they matter, before...
              06-18-2026, 07:11 AM
            • SEQadmin2
              From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
              by SEQadmin2


              Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


              The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
              ...
              06-02-2026, 10:05 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 06-26-2026, 11:10 AM
            0 responses
            8 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-17-2026, 06:09 AM
            0 responses
            44 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-09-2026, 11:58 AM
            0 responses
            104 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 06-05-2026, 10:09 AM
            0 responses
            125 views
            0 reactions
            Last Post SEQadmin2  
            Working...