  • Problematic bioanalyzer results (QC after mRNA capture)

    Did mRNA capture (Dynabead) for my RNA extract samples (mushroom). When run mRNA pico assay for QC, these kind of result appeared. Sometime it work for some of the samples. But repeated assay using same sample seem to produce different result. Contacted field application specialist and made the listed changes but problem persists.

    Changes made:

    1. washed electrode with RNAzap and consequently with RNase free water and air dried for more than 10 minutes. Even diagnosed for short circuit before run.

    2. check priming station for failure. It seem to be working because it bounce normally when the latch was released

    3. my sample were treated with DNase from MoBio.

    4. diluted my samples.

    Attached here are pdf. containing results for last few run.

    2100 expert_mRNA Pico_DE13805702_2016-06-27_18-03-51.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-14_17-28-00.pdf

    2100 expert_mRNA Pico_DE13805702_2016-06-27_16-40-47.pdf

  • #2
    Hi sbchua.1990,
    We have seen traces such as these when the chip priming/loading was not performed correctly. Did you ensure that the chip priming station plunger was set to the highest notch and the plunger was released after exactly 30 seconds? Was the chip loaded with fresh reagents?

    A couple of other possibilities include overloading of the chip and salt or detergent carryover. You noted that you already tried diluting the samples; how much material was used? We typically load 1-5 ng/well onto the Pico chip.

    Finally, have you tried an additional clean-up of your samples?

    Hope that helps!
    Denise Stephens
    Technical Support Scientist at NuGEN Technologies

    Comment


    • #3
      Thanks for reply, I am pretty sure that the chip priming/loading was performed correctly. Reagents are still within expiration date.

      My samples concentration should be around 300pg - 1500pg. I just loaded 1 ul without dilution.

      No I did not try additional clean-up for my samples.

      Comment


      • #4
        The samples most likely are contaminated with residual removal resin. You would need to contact MoBio for cleanup instructions.

        Comment


        • #5
          Hi nucacidhunter, did you encounter such problem in the past? Or what make you think so? Thanks for your reply.

          Comment


          • #6
            Profile indicates presence of insoluble contaminants which could be protein (residual DNase) or removal resin. I suspect the resin because it can be carried over easily if spinning was not long enough or one tried to take the supernatant without leaving any liquid behind in the tube. There is also possibility of protein protectant carry over which is used to stabilise DNase at RT.

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Latest Developments in Precision Medicine
              by seqadmin



              Technological advances have led to drastic improvements in the field of precision medicine, enabling more personalized approaches to treatment. This article explores four leading groups that are overcoming many of the challenges of genomic profiling and precision medicine through their innovative platforms and technologies.

              Somatic Genomics
              “We have such a tremendous amount of genetic diversity that exists within each of us, and not just between us as individuals,”...
              05-24-2024, 01:16 PM
            • seqadmin
              Recent Advances in Sequencing Analysis Tools
              by seqadmin


              The sequencing world is rapidly changing due to declining costs, enhanced accuracies, and the advent of newer, cutting-edge instruments. Equally important to these developments are improvements in sequencing analysis, a process that converts vast amounts of raw data into a comprehensible and meaningful form. This complex task requires expertise and the right analysis tools. In this article, we highlight the progress and innovation in sequencing analysis by reviewing several of the...
              05-06-2024, 07:48 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 06-03-2024, 06:55 AM
            0 responses
            12 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 05-30-2024, 03:16 PM
            0 responses
            26 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 05-29-2024, 01:32 PM
            0 responses
            29 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 05-24-2024, 07:15 AM
            0 responses
            215 views
            0 likes
            Last Post seqadmin  
            Working...
            X