    Shorter fragments than expected

    My problem is having shorter fragments after sequencing than expected. The size of my library should be around 300 bp and I checked the size with Tape station which was ~300 bp. However, my sequencing results show that my actual transcripts are too short that that even can`t be mapped to anywhere. Then, I checked my libraries with Sanger sequencing and they were also around 150 bp.
    What can it be the reason that I see at Tape station results that 300-350 bp but our fragments are less than 150??
    The only mistake I am sure is my libraries (CEL-Seq libraries, which you can pool many samples in one tube) is not compatible with Hi-Seq v4 chemistry (don`t know the reason but the method published to be so, now the authors says it works with v4 only %20 Phix is used) but works fine with v3 chemistry or Rapid mode. Previously the same protocol worked fine with Rapid mode and now we tried v4 with %1 Phix and the results are very strange with short fragments.
    Any idea??
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Two possible reasons:
    1- Tape Station indicates library fragments length which includes insert plus adapter. Average adapter length for standard libraries are around 120b but Cell-seq adapters are longer due to inline cell specific barcodes so the transcript insert length are shorter
    2- Shorter fragments in libraries are sequenced preferentially so average sequence read length will be shorter than average library insert

    You have not mentioned sequencing configuration. Cell-seq design is poor as in standard sequencing R1 has to be read at least 25 cycles. This will include reading through polyT and without high PhiX spike-in the %PF will be low.

    Posting library Tape profile and FastQC results will give better ideas for troubleshooting.

    Comment

    • peripoe
      Junior Member
      • May 2017
      • 2

      #3
      Hi,

      Please see the files that I upload. I also need to correct that Sanger result is 180 bp including adapters which leaves 50-60 bp and sometimes shorter fragments. With Tape station, I see a peak ~300 bp which makes us think that our product is ~150 bp.
      Attached Files

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        Tape profile shows some adapter/primer dimers that they also will be sequenced preferentially along with shorter insert fragments.

        Plots confirm sequencing short inserts as after cycle 80 polyA is clearly distinguishable and overall high A residue also indicates sequencing short fragments containing only polyA. If you need longer reads for mapping then fragmentation step has to be modified to get larger RNA if you are following original publication protocol. This can be done by one way bead size-selection on the current library as well but I am not sure what will be the effect on results.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Yesterday, 11:13 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, Yesterday, 10:48 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        36 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        42 views
        0 reactions
        Last Post SEQadmin2  
        Working...