Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • shurjo
    Senior Member
    • Jan 2009
    • 132

    #1

    Pippin Prep vs. gel elution for RNA-Seq

    I had a chance to do a side-by-side comparison of Pippin Prep (PP) and agarose gels (AG) for size selection during RNA-Seq library prep and would like to share these results with the community and get some feedback.

    In the picture below, I show eight RNA-Seq libraries run on a BioAnalyzer HS DNA chip. The four sharp and high peaks are from PP while the four slightly lower and broader peaks are from AG. Equal aliquots of four adapter-ligated cDNAs were used for both platforms (so it's really 4 samples, then two techniques on each sample). Noticeably, the GS libraries have a lot more non-target material both above and below the peak, although all eight have a "bump" close to the upper size marker at 10380bp. A question: which of these would make the better candidate for submission to the sequencing center? I am leaning towards the PP as there is lesser background. Any advice from folks who are routinely using Pippin Prep for RNA-Seq?

    Thanks,

    Shurjo
    Attached Files
  • mbk0asis
    Member
    • Dec 2011
    • 41

    #2
    No way to remove adapter dimers?

    The quote below is an old post from here in Seqanswers.
    This guy compared two library size selection methods, agarose gel elution and pippin prep with his samples.
    I like the sharp and narrow results of pippin prep, but adapter dimers still appeared at ~120bp. I thought most dimers would disappear when used pippin prep.
    Has anyone tried pippin prep? Would somebody tell me how good it is?
    And I'm planing to size select out 200~500bp. Would a broad range elution affect performance of the machine?

    Thank you!!



    Originally posted by shurjo View Post
    I had a chance to do a side-by-side comparison of Pippin Prep (PP) and agarose gels (AG) for size selection during RNA-Seq library prep and would like to share these results with the community and get some feedback.

    In the picture below, I show eight RNA-Seq libraries run on a BioAnalyzer HS DNA chip. The four sharp and high peaks are from PP while the four slightly lower and broader peaks are from AG. Equal aliquots of four adapter-ligated cDNAs were used for both platforms (so it's really 4 samples, then two techniques on each sample). Noticeably, the GS libraries have a lot more non-target material both above and below the peak, although all eight have a "bump" close to the upper size marker at 10380bp. A question: which of these would make the better candidate for submission to the sequencing center? I am leaning towards the PP as there is lesser background. Any advice from folks who are routinely using Pippin Prep for RNA-Seq?

    Thanks,

    Shurjo

    Comment

    • luc
      Senior Member
      • Dec 2010
      • 469

      #3
      The Pippen prep works just fine and reliable in our experience, but so can the gel extraction protocols. I would not think that the GE results from the previous post are representative. GE results can look a lot better - one might dissolve the gel pieces better at RT than at 50 degrees as most GE kit manufacturers recommend. Adapter dimers can also be removed with Ampure or equivalent beads. I don't see any signs of adapter dimer problems in the PDF above however.
      If selecting a wide size range there can be potentially be problems with the collection well overflowing with the PP - I have not tried it, though - perhaps better use a low % agarose cassette then and ask their tech support?


      Originally posted by mbk0asis View Post
      The quote below is an old post from here in Seqanswers.
      This guy compared two library size selection methods, agarose gel elution and pippin prep with his samples.
      I like the sharp and narrow results of pippin prep, but adapter dimers still appeared at ~120bp. I thought most dimers would disappear when used pippin prep.
      Has anyone tried pippin prep? Would somebody tell me how good it is?
      And I'm planing to size select out 200~500bp. Would a broad range elution affect performance of the machine?

      Thank you!!
      Last edited by luc; 07-04-2013, 03:37 PM.

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        The Factors Reshaping Genome Assembly
        by SEQadmin2



        When the Human Genome Project declared the human genome finished in 2003, portions of the sequence were still missing, including many of its most repetitive and structurally complex regions.1 Still, for the next two decades, researchers analyzed their sequencing data against the single linear reference that came out of this effort. Released in 2013 as GRCh38, that reference has powered countless discoveries, but it has gaps and cannot capture the genetic diversity of the world’s...
        11-24-2026, 07:04 AM
      • SEQadmin2
        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
        by SEQadmin2





        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...
        10-30-2026, 06:46 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 11-23-2026, 11:11 AM
      0 responses
      32 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-17-2026, 06:27 AM
      0 responses
      42 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-12-2026, 07:59 AM
      0 responses
      65 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 11-05-2026, 09:25 AM
      0 responses
      67 views
      0 reactions
      Last Post SEQadmin2  
      Working...