Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • captainentropy
    Member
    • Mar 2009
    • 89

    #1

    TruSeq index mixing strategies

    I can't find this answer anywhere. I was told that there are some combinations of barcodes/indexes that should be avoided in a lane.

    Does anyone have a list of the least problematic combinations?
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Hi captainentropy,

    Well, there should be! Maybe there was at one time? We have only been doing Illumina runs for a few months. For the most part Illumina seems to have their act together in ways Roche and Applied Biosystems do not. However this is one case where Roche and ABI seem to have worked out issues that Illumina is letting their users stumble over.

    The only guidance in this matter I have been able to find came in another thread from ETHANol. He recommended this link:

    http://www.chem.agilent.com/Library/...exed_v.1.2.pdf

    I quote the relevant passage from "SureSelect XT Target Enrichment System Kit for Illumina Multiplexed Sequencing" pp 54-55 below:
    Some sequencing experiments require the use of fewer than 12 index sequences in a lane with a high cluster density. In such cases, select indexes carefully to ensure optimum base calling and demultiplexing by having different bases at each cycle of the index read. Illumina recommends the following sets of indexes for low-level pooling experiments.
    Pool of 2 samples:
    • Index #6 GCCAAT
    • Index #12 CTTGTA
    Pool of 3 samples:
    • Index #4 TGACCA
    • Index #6 GCCAAT
    • Index #12 CTTGTA
    Pool of 6 samples:
    • Index #2 CGATGT
    • Index #4 TGACCA
    • Index #5 ACAGTG
    • Index #6 GCCAAT
    • Index #7 CAGATC
    • Index #12 CTTGTA
    I think the main idea is that there be clusters visible during each scanning pass so the instrument doesn't decide it has fallen out of focus and attempt to focus on a blank lane. If I understand correctly our sequencer (HiScanSQ) does A/C and G/T in separate passes. So if at least, say, 10% of the clusters are lit during both scanning passes, the instrument won't get confused. That is, any mixture of indexes that includes both an "M" (A or C) and a "K" (G or T) at each of its six positions should be okay. For the HiScanSQ. For a HiSeq, you would want to check to make sure that it scans A/C and G/T in those pairs. Probably, I guess. Or do mixtures that have all 4 bases at each position of the index.

    --
    Phillip
    Last edited by pmiguel; 08-31-2011, 03:48 AM.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      As far as avoiding particular combinations, the TruSeq adapters are designed so that a single error in the index sequence read does not convert it to another index sequence (thereby preventing assignment of the read to the wrong sample). CASAVA allows the option of tolerating zero or one mismatches in the index during demultiplexing.

      -Harold

      Comment

      • epistatic
        Senior Member
        • Mar 2009
        • 129

        #4
        The information that Agilent posts comes from the Illumina Multiplexing sample preparation guide. The latest version I have (rev. D, 2/2011) has on page 39 the same "low-level pooling" recommendations.

        Indexes:
        The twelve supplied sequences are unique and can be distinguished from one another by considering only the first three nucleotides. In addition, no sequence has more than three positions of agreement with any of the others. In this way they have been engineered to be “self‐correcting” in that an index with a single error will still match better to the correct index than to any one of the other indexes, and can therefore still be used. The twelve indexes are as follows:

        Low-Level Pooling:
        Some sequencing experiments require the use of fewer than 12 index sequences in a lane with a high cluster density. In such cases, a careful selection of indexes is required to ensure optimum cluster discrimination by having different bases at each cycle of the index read. Illumina recommends the following sets of indexes for low‐level pooling experiments.

        2 samples: 6, 12; 3 samples: 4,6,12; 6 samples: 2,4,5,6,7,12.

        The phix control lane contains single index (#3).

        Comment

        • ashish
          Junior Member
          • Nov 2010
          • 5

          #5
          Originally posted by epistatic View Post
          The phix control lane contains single index (#3).
          Only if you are using PhiX v2. The new v3 is unindexed.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by ashish View Post
            Only if you are using PhiX v2. The new v3 is unindexed.
            Yes, isn't that odd? The idea is that the unindexed PhiX just ends up in the index-unassigned "bilge" along with low-quality clusters that failed to demultiplex. But we sometimes include a non-indexed library mixed in with indexed samples. So I don't really like that idea.

            You can still get indexed PhiX (v2), just buy that weird multiplex kit:

            PE-400-2002 | Multiplexing Sequencing Primer and PhiX Control Kit v2.0 (10 flow cells)

            --
            Phillip

            Comment

            • gringer
              David Eccles (gringer)
              • May 2011
              • 845

              #7
              Also see this thread about indexes influencing spot identification:

              Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              27 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              20 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-15-2026, 10:52 AM
              0 responses
              36 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              43 views
              0 reactions
              Last Post SEQadmin2  
              Working...