    Post PCR Enrichment Gel Selection for ChIP-Seq

    Hello,

    After I PCR enrich ChIP-seq libraries following the Illumina protocol I notice bioanalyzer peaks corresponding to what I think is primer or primer (<100bp). Has anyone else seen this? Should we consider a gel size selection post-PCR enrichment?

    Thanks
  • HESmith
    Senior Member
    • Oct 2009
    • 512

    #2
    Adapter dimers are a common contaminant of Illumina libraries. If it constitutes a significant fraction of the total, I would recommend size selection after PCR enrichment. If it's only a minor contaminant, it's typically not worth the effort.

    Comment

    • indu87
      Junior Member
      • May 2012
      • 4

      #3
      Originally posted by Daytwa View Post
      Hello,

      After I PCR enrich ChIP-seq libraries following the Illumina protocol I notice bioanalyzer peaks corresponding to what I think is primer or primer (<100bp). Has anyone else seen this? Should we consider a gel size selection post-PCR enrichment?

      Thanks
      Hi Daytwa- I notice the same with my library prep. How did you get rid of the peaks (adaptors/primers) ? Did you reduce the concentration of adaptors? Or use beads for size selection? Or gel purify again after PCR? Or run gel longer (5hrs) & slower before PCR enrichment?

      Comment

      • Daytwa
        Member
        • Dec 2010
        • 19

        #4
        @indu

        what i've found works best (especially when constructing low input libraries (~10 ng or less) is two use the ampure xp beads following ligation (i do this 3 times). i then run a ~3% gel. Make sure you do not add EtBr to the gel. Do a post stain. It takes more time but it helps with keeping adapter aggregates from contaminating your library. Also run the gel slow. Once stained I do an old fashioned clean blade gel slice followed by a Qiagen gel purification column. After PCR enrichment I will typically do another Ampure purification. This isnt always needed but i don't notice much loss so what the hey. If you have an extra chip you can always run a Bioanalyzer high sensitivity chip to inform whether or not you need the 2nd ampure purification. hope this helps.

        Comment

        • indu87
          Junior Member
          • May 2012
          • 4

          #5
          thank you for your response, i think this helps. i will try beads instead of direct size selection on gel.

          just one more question- i notice that the concentration of my samples are really poor in comparison to input after library prep, although the starting material used was 10ng throughout. i doubt if my initial QUBIT measurement was wrong considering that input I measured along with samples have worked fine. i re-checked my original calculations to make sure i had 10ng to begin with and this also seems ok.

          what puzzles me is that i seem to have lost most of the DNA for samples and not input, whilst the original yield from the IPs were pretty good (range 100-250ng total)

          any suggestions would help ! many thanks

          Comment

          • Daytwa
            Member
            • Dec 2010
            • 19

            #6
            @indu,

            I am somewhat confused. You are worried about the concentration of you library vs the input? If that is the case don't be. It is normal for the prep to be somewhat lossy. Yields are never 100% with columns and I bet it's even worse at the gel step. That said you need so very little library to get a run that you'll likely be fine.

            Also I want to stress that I do Ampure selection -> Gel -> Column Purification -> PCR ->Ampure (may or may not be needed). I would strongly recommend quantifying your library by PCR too. Also check the size distribution on a bioanalyzer if possible. These last two steps are very important to determining the quality of the library.

            Best.

            Comment

            • indu87
              Junior Member
              • May 2012
              • 4

              #7
              hi, maybe i was not clear. pl find attached. the DNA from the CHIPs have poor yield post library prep in comparison to Input.

              starting material for the library prep was 10ng for all my samples & input.
              Attached Files

              Comment

              • indu87
                Junior Member
                • May 2012
                • 4

                #8
                I have got samples (after library prep) which almost look like H20 control with no peaks or very negligible sized peaks around the size range 200-300bp.
                this ofcourse is after blind size selection of DNA on gel before PCR amplification.

                while the input has worked very well (processed alongside samples), albeit contamination around 100bp (presumably adaptors/primers), i seem to have lost or got little DNA left from the actual IPs after library prep. the only dominating peak for my samples is the actual contaminant you can see around 100bp.

                thanks for your reply in advance !

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                  by SEQadmin2


                  I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                  Here are nine questions we think about, in roughly the order they matter, before...
                  06-18-2026, 07:11 AM
                • SEQadmin2
                  From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                  by SEQadmin2


                  Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                  The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                  ...
                  06-02-2026, 10:05 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 11:10 AM
                0 responses
                7 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-17-2026, 06:09 AM
                0 responses
                43 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-09-2026, 11:58 AM
                0 responses
                104 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 06-05-2026, 10:09 AM
                0 responses
                125 views
                0 reactions
                Last Post SEQadmin2  
                Working...