Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • whw
    Member
    • Jun 2011
    • 19

    Low input/adaptor dimers

    Hi all,

    I'm preparing libraries from very small amounts of RNA, around 100-1000x less than the 1ug recommended by illumina. Generating a "library" seems easy but on analysis it turns out to be mostly adaptor. Has anyone experimented with adding less adaptor? Or any other ways of optimizing library preparation for small amounts of nucleic acids?

    I get the impression this is a fairly common problem, I'll be on the phone to illumina in a bit and I'll pass along any gems they give me. In the meantime what do all you seqanswerers think?

    WW
  • arolfe
    Member
    • Jul 2011
    • 29

    #2
    I believe that different Illumina kits used different adapter concentrations. For example, I think the only difference between the ChIP-Seq and genomic (?) kits was a 100X difference in adapter concentration. So yes, it makes sense and you should be able to look around to get an idea of where to start.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      The ratio of adaptor to sample is probably the most critical variable in library construction. You should be able to adapt the ChIP-Seq protocol, which works with 10ng dsDNA. You may need to perform an additional round of size selection to remove adaptor dimers. Alternatively, you can perform linear amplification of your starting material.

      Harold

      Comment

      • whw
        Member
        • Jun 2011
        • 19

        #4
        I have the Ultra-low input mRNA Seq Guide from Illumina which is optimized for 2ng of dsDNA. I'm thinking I'll try their adaptor dilutions and see what happens, hopefully the ligation chemistry won't be too different.

        Comment

        • griffes
          Junior Member
          • Oct 2011
          • 4

          #5
          what adapter dilutions are you using for RNA?

          Comment

          • ETHANol
            Senior Member
            • Feb 2010
            • 308

            #6
            Just in case you were wondering from information I've gathered from other threads the concentrations of the TruSeq adapters are:
            DNA Sample prep Kit adapters = 15 μM
            RNA Sample Prep Kit adapters = 0.25 μM

            And yes, adjusting the amount of adapter to the amount of DNA you want to ligate it to is good practice. The adapter should be in excess but I'd say not more then 10 fold and probably better to be less.
            --------------
            Ethan

            Comment

            • whw
              Member
              • Jun 2011
              • 19

              #7
              Originally posted by griffes View Post
              what adapter dilutions are you using for RNA
              The ultra-low input protocol suggests a dilution of 1:19 for 1-4ng of sample. I thought I'd start in that range.

              Originally posted by ETHANol View Post
              Just in case you were wondering from information I've gathered from other threads the concentrations of the TruSeq adapters are:
              DNA Sample prep Kit adapters = 15 μM
              RNA Sample Prep Kit adapters = 0.25 μM

              And yes, adjusting the amount of adapter to the amount of DNA you want to ligate it to is good practice. The adapter should be in excess but I'd say not more then 10 fold and probably better to be less.
              That's very helpful, thanks for the tip Ethan(ol)

              Comment

              Latest Articles

              Collapse

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 06-09-2026, 11:58 AM
              0 responses
              22 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-05-2026, 10:09 AM
              0 responses
              29 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-04-2026, 08:59 AM
              0 responses
              39 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 06-02-2026, 12:03 PM
              0 responses
              61 views
              0 reactions
              Last Post SEQadmin2  
              Working...