Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • ssing
    Member
    • Jan 2009
    • 21

    #1

    double peak in mRNA truseq kit

    Hi all,

    I have been doing some mRNA TruSeq sample prep, using the protocol *exactly* as advised by the Illumina folk (using PolyT beads to get mRNA, fragmenting for 7 minutes chemically, 15 PCR cycles), and I am getting an odd and consistent pattern where I have my desired peak ~300 bp and a second peak ~900 bp. See an exemplar trace attached.

    Reading through the other posts, seems like most people think this due to overamplification... could that be my issue here? My biggest concern is (because the double peak is similar in size ratio to the rRNA size ratio) that this reflects rRNA contamination. Does anyone have experience with that?

    Thanks!
    Attached Files
  • Prosuite
    Junior Member
    • Feb 2011
    • 3

    #2
    The extra peak is due to SPRI bead carryover from post-PCR clean-up step. Use a
    strong magnet for bead separation and pipette carefully during
    elution to avoid disturbing beads. It should not affect the sequencing run.
    Attached Files

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Hi Prosuite,
      That is very interesting. Where did you find that document from Agilent? I was just looking at the instrument troubleshooting guide this morning and there was no mention of SPRI/AMPure beads.

      --
      Phillip

      Comment

      • Prosuite
        Junior Member
        • Feb 2011
        • 3

        #4
        It is from Agilent eSeminar "BioAnalyzer Applications for Next Gen Sequencing : Updates and Tips-20110301". Here is the link: https://agilenteseminar.webex.com/ag...114f10d5514079

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Never seen this before. Interesting. Anyone buy the "ligase remaining attached to library molecule" issue presented on page 26?

          --
          Phillip

          Comment

          • ETHANol
            Senior Member
            • Feb 2010
            • 308

            #6
            Delete what I wrote. I didn't read yours carefully enough. I doubt it is an over amplification issue because that peak usually is 2x the size of the correct peak.
            Last edited by ETHANol; 11-07-2011, 11:24 AM.
            --------------
            Ethan

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by ETHANol View Post
              Delete what I wrote. I didn't read yours carefully enough. I doubt it is an over amplification issue because that peak usually is 2x the size of the correct peak.
              Did you check out the Prosuite attachment? Some info from Agilent claiming that SPRI beads can cause that result. (I have see it -- a peak much larger than one would expect from multimers/heterodimers/bubble amplicons, etc.)

              If so had you ever seen that info? Be nice if Agilent provided that as part of their normal troubleshooting manual...

              --
              Phillip

              Comment

              • ssing
                Member
                • Jan 2009
                • 21

                #8
                When I contacted Illumina, they gave me three possible explanations:
                1. Incomplete fragmentation (which seems unlikely, given that I was using RNA within the range of what the recommended and a long fragmentation time)
                2. Over-amplification (which I also doubted because I didn't do a ridiculous number of cycles)
                3. Bead carryover (which made the most sense as this was a consistent pattern)

                If it is bead carryover, when I run a normal gel of the cleaned product, I shouldn't see that smear anymore. I'm going to do that tomorrow. I'll report back what I see. Here's hoping that is what it is, because otherwise, they are recommending I do a size selection for all my samples...

                Thanks for all the feedback.

                Comment

                • ECO
                  --Site Admin--
                  • Oct 2007
                  • 1360

                  #9
                  Originally posted by pmiguel View Post
                  Never seen this before. Interesting. Anyone buy the "ligase remaining attached to library molecule" issue presented on page 26?
                  Chiming in with two things from my experience that may be relevant...not commenting on "page 26" because I can't see where you're referring to that....

                  1. Intact ligase has a strong protective effect against any treatments with exonuclease. Denature the ligase or strip out the Mg++ and you're good to go.

                  2. The bioanalyzer (particularly the high sensitivity kit) can be used as a gel-shift assay to detect bound proteins (DNA polymerases in my experience)...it wouldn't surprise me in the slightest that active ligase causes a shift on the bioanalyzer....in fact that's probably why ILMN adds "stop solution" to their ligations.

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Bound ligase shifts Illumina libraries sizes?

                    Originally posted by ECO View Post
                    Chiming in with two things from my experience that may be relevant...not commenting on "page 26" because I can't see where you're referring to that....
                    Sorry, Prosuite posted a link to the presentation in question above. You can skip to page 26 fairly easily. You do have to register so Agilent can spam you -- but I would say the information in the presentation is worth a little spam to anyone that uses the BioAnalyzer much.

                    The slide in question shows two electropherograms of an Illumina library after adapter ligation and PippinPrep size selection. One PippenPrepped after a proteinase K treatment, one without treatment. The peak with treatment is 368 bp. The one without is 291 bp.

                    Actually this addresses one of those perennial Illumina library prep questions that shows up on SeqAnswers. But I have never seen ligase binding suggested as the cause.

                    Originally posted by ECO View Post
                    1. Intact ligase has a strong protective effect against any treatments with exonuclease. Denature the ligase or strip out the Mg++ and you're good to go.

                    2. The bioanalyzer (particularly the high sensitivity kit) can be used as a gel-shift assay to detect bound proteins (DNA polymerases in my experience)...it wouldn't surprise me in the slightest that active ligase causes a shift on the bioanalyzer....in fact that's probably why ILMN adds "stop solution" to their ligations.
                    Good to know.

                    --
                    Phillip

                    Comment

                    • ssing
                      Member
                      • Jan 2009
                      • 21

                      #11
                      In fact, this does not appear to be due to beads. I am not sure what the peak is, but I am going to size select ... trivial exercise and worth ensuring good cluster generation.

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        So you saw the same phenomenon on your agarose gel? Any chance you could post that?

                        You might want to strand denature the sample (94 oC for 2-4 minutes followed by "snap cool" on ice) and run it on an RNA chip.

                        You can get double peak artifacts in TruSeq libraries that end up causing no real issues when clustered. Normally these peaks are closer in size to one another than the image you give above, though.

                        --
                        Phillip

                        Comment

                        • ssing
                          Member
                          • Jan 2009
                          • 21

                          #13
                          Hi Philip,

                          See attached. The ladder is a 100 bp ladder (NEB).
                          Attached Files

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            (The upper loading dye is really distracting. Xylene cyanol?) Is this TBE EtBr? How many cycles of enrichment PCR did you do?

                            Anyway, yes there is a high molecular weight band. No harm in doing a size selection. Actually we nearly always size select (after pooling most or all of the indexes that will go in a lane). I doubt the high MW stuff will cause any issues, though.

                            Still, mysterious. Wonder if it could be the polymerase still bound to some of the molecules. (See ECO's comments above...)

                            --
                            Phillip

                            Comment

                            • ssing
                              Member
                              • Jan 2009
                              • 21

                              #15
                              This is TAE EtBr and I did 15 cycles of enrichment PCR.

                              Not sure what this is, but I went ahead and size selected. I'll let you all know when I get my data...hopefully this is just a weird artifact and doesn't suggest anything went wrong with my library prep.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              40 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              29 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              42 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...