  • Gina_P
    Member
    • Dec 2011
    • 20

    Fragment size limit

    Hi all, I have some libraries with fragments that look a bit on the large side (650 - 700 bp including adaptors) for re-sequencing. Should I just use these libraries anyway? Or possibly redo gel extraction? I saved the original gels.

    Thanks!
    - Gina
  • DMO
    Member
    • Aug 2010
    • 28

    #2
    What was your ave bp size post sonication?

    Do you have a picture of your gel or BioA QC of the final product?

    Whether or not you should move into sequencing will depend upon whether this was caused by less than suspected efficiency in your shearing method or due to an overamplification in the final enrichment of the library.

    Comment

    • Gina_P
      Member
      • Dec 2011
      • 20

      #3
      Hi DMO,

      Thanks for answering me. I have a BioA QC. There are a few libraries that need to be cleaned up.

      Also, apologies for being vague - this is for Illumina 2x50 sequencing!
      Attached Files
      - Gina

      Comment

      • DMO
        Member
        • Aug 2010
        • 28

        #4
        Do you have a BioA trace post-sonication?

        Comment

        • Gina_P
          Member
          • Dec 2011
          • 20

          #5
          Unfortunately I don't... For now, I have recut the gels below my original extractions. I used the same shearing protocol on Covaris that I have used in the past, with good results, and I have never done a BioA QC check on shearing products. However, now I will probably do so in the future to make sure I the shearing went well... Any other advice? Thanks for answering!
          - Gina

          Comment

          • DMO
            Member
            • Aug 2010
            • 28

            #6
            I see smaller peaks on at least 2 of your traces which may indicate the samples were quite overamplified.

            Can you answer a few more questions?

            What is you shearing protocol?

            What amount of material did you start with?

            How many cycles of amplification did you do for your final enrichment?

            Comment

            • Gina_P
              Member
              • Dec 2011
              • 20

              #7
              I used the whole-genome resequencing protocol for Covaris:
              40 seconds, 10% duty cycle, 5.0 intensity, 200 bursts per second, frequency sweeping mode, 6 degree temperature.

              I started with 1 ug DNA.

              10 cycles of amplification. Should I reduce?
              - Gina

              Comment

              • DMO
                Member
                • Aug 2010
                • 28

                #8
                You certainly could since you are getting plently of material, but based on these parameters I am less likely to jump to the overamplification conclusion. What size did you cut from your gel?

                Comment

                • Gina_P
                  Member
                  • Dec 2011
                  • 20

                  #9
                  I cut from about 500 - 600, which was by accident. I meant to cut from 400 - 500. I have additional gel slices from 400 - 500 range saved in my fridge.
                  - Gina

                  Comment

                  • DMO
                    Member
                    • Aug 2010
                    • 28

                    #10
                    I would extract the DNA from those and amp those. I think you would be fine sequencing what you have, but you would have to scale back your loading concentration on your run a bit. So, if you finished prepping the other cut, you could load your seq run at a more standard concentration and acheive a higher number of reads out doing so.

                    Comment

                    • Gina_P
                      Member
                      • Dec 2011
                      • 20

                      #11
                      Hm, good point about the concentration. Thanks for your help. I'm going to prep the other cuts today.
                      - Gina

                      Comment

                      • Gina_P
                        Member
                        • Dec 2011
                        • 20

                        #12
                        Here's the new BA result with smaller fragment size. There's still some problem with sample 9. For now, I'm not going to put it on the sequencer. It looks like the DNA is dragging along some smaller molecule with it, because a similar bump is in both of my different gel extractions. Also, I used Kapa Biosciences' library amp kit instead of the Illumina one this time...
                        Attached Files
                        - Gina

                        Comment

                        Latest Articles

                        Collapse

                        • SEQadmin2
                          Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                          by SEQadmin2



                          Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                          ...
                          07-09-2026, 11:10 AM
                        • SEQadmin2
                          Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                          by SEQadmin2



                          Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                          There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                          07-08-2026, 05:17 AM
                        • GATTACAT
                          Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                          by GATTACAT
                          Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                          07-01-2026, 11:43 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, 07-13-2026, 10:26 AM
                        0 responses
                        25 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-09-2026, 10:04 AM
                        0 responses
                        35 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-08-2026, 10:08 AM
                        0 responses
                        22 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-07-2026, 11:05 AM
                        0 responses
                        34 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...