  • chipmonk
    Member
    • Jul 2010
    • 18

    BioAnalyzer inverted peaks!!

    I am running my transcription factor chip seq samples in Bioanalyzer HS and getting weird inverted peaks(pic attached)in input samples!! has anybody come across such 'peaks' before?
    These are only seen in input samples whereas the pulldown and controls show no bumps at all.Even the tei meth histone L27 pulldown used as positive control fails to show peaks. where as the PCR verification of targets for controls works fine.
    I am using HS DNA kit for BA. Magnify kit for pulldown. The kit elutes the DNA finally in 150 ul. which is too dilute to work with. has anyone else worked with this kit and what is the next step to say doing a NGS chip seq.

    Thanks
    Attached Files
  • NextGenSeq
    Senior Member
    • Apr 2009
    • 482

    #2
    We saw this too in a ChIP-Seq sample and I'm not sure what causes it. It may be due to the proteins crosslinked to the DNA?

    Comment

    • chipmonk
      Member
      • Jul 2010
      • 18

      #3
      @next gen seq

      were you able to rectify this and if yes by what means!. could purification again help or is it a problem with incomplete de-crosslinking??

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        My guess would be a dye or fluor of some sort. The fluor used by the bioanalyzer presumably has some background emission level when not bound to DNA or RNA. The emission increases in the presence of RNA or DNA. In the presence of some other compounds one might see a decrease in emission.

        You could check with Agilent tech support. They might have an answer for you.
        --
        Phillip

        Comment

        • chipmonk
          Member
          • Jul 2010
          • 18

          #5
          thanks

          I have written to tech support. will post if i get a reply!

          Comment

          • julianaect
            Junior Member
            • May 2012
            • 1

            #6
            Originally posted by chipmonk View Post
            I have written to tech support. will post if i get a reply!
            I got inverted peaks too. Did you get reply?

            Thanks!

            Comment

            • chipmonk
              Member
              • Jul 2010
              • 18

              #7
              not much help

              hi julian
              That was a one off thing i guess. some tech people told it may be because of the remnant magnetic beads. so if you are using dynabead like reagent keep the tubes in rack while pipetting for the BA sample. others speculated there might be excessive protein as well. i think the former may be close to something that may work. all the best.

              Comment

              • FredTam1
                Junior Member
                • Aug 2013
                • 8

                #8
                No way, that is a common occurrence for sure. And it's random, which leads me to think it has something to do with an error in the voltage - a machine error. If Agilent won't cop to it, or they keep requesting 'the data' even after a blatant overt description of the graph then it's a common problem inherent to the machine imo

                Comment

                • Olivia16
                  Junior Member
                  • Feb 2012
                  • 7

                  #9
                  We had similar issues in some of our bioanalyzer runs and after working with tech support for about a month they said that it was due to either: 1- dirty electrode pins 2- mistake when making the gel dye mix 3- need a new gasket in the chip priming station....any how we did all the things that Agilent had suggested and the problem got better after cleaning the electrode pins very well and getting a new kit. Therefore it is hard to say if it was the pins or the kit that inverted peaks were coming from. Hopefully that is a bit helpful?

                  Did you look at the bioanalyzer troubleshooting manual (link below)?

                  Comment

                  • chipmonk
                    Member
                    • Jul 2010
                    • 18

                    #10
                    not machine issue

                    thanks for the replies, but when i run these samples that give inverted peaks I also run samples which i know have worked well so random chances or voltage issue or any other machine based issues are ruled out. I think it is the magnetic remnants of the prot-a mag beads. that may cause. i have pipetted using the magrack while loading on gel. hope someone figures out what exactly is happening. but nevertheless my chipseq works!

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                      by SEQadmin2


                      I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                      Here are nine questions we think about, in roughly the order they matter, before...
                      06-18-2026, 07:11 AM
                    • SEQadmin2
                      From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                      by SEQadmin2


                      Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                      The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                      ...
                      06-02-2026, 10:05 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, Yesterday, 11:10 AM
                    0 responses
                    7 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 06-17-2026, 06:09 AM
                    0 responses
                    42 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 06-09-2026, 11:58 AM
                    0 responses
                    103 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 06-05-2026, 10:09 AM
                    0 responses
                    125 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...