    DNA loss on Bioanalyzer

    About two weeks ago I sheared some gDNA samples on the Covaris E210 and purified them with AMPure beads, as instructed by the SureSelect XT protocol. I measured the samples on the Bioanalyzer 2100 immediately following clean-up and received good peaks at the right area. However, I re-measured the samples as a positive control and I see that even though the peaks are still present for my sample, they are much shorter and it appears I have lost some gDNA.

    My first instinct is to consider this nuclease contamination, but if this were the case, wouldn't I not see any evidence of DNA on the Bioanalyzer? I'm also wondering if I lost DNA following the freeze-thaw cycle the samples underwent, for they were stored at -20 degrees Celsius. Does anyone else have any ideas about what could be causing this DNA loss, as well as any suggestions about how I should prevent it? Thanks!
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    What's the original sample source (organism and if applicable, tissue) and purification method? How much shorter is shorter? Did the whole peak move

    Nuclease contamination is a possibility. Freeze-thaw problems are very unlikely in my experience, especially if you're observing a dramatic change in sample quality from a single cycle.

    I always suspect the Bioanalyzer. =] Any quantitation method that relies on pipetting of 1ul is by definition a ballpark method (unless you run several dilutions/reps etc).

    Comment

    • moedog45
      Junior Member
      • Feb 2012
      • 4

      #3
      The original source was tissue, and the DNA was purified with Agencourt AMPure XP magnetic beads. The peak is shorter by about 15 fluorescence units, but the peak itself did not move.

      I talked to tech support about this and they told me if there is only one visible peak and the peak size hasn't moved to a smaller size range, then I shouldn't worry about nuclease contamination. Tech support then validated your opinion, ECO, and said it was likely discrepancy between the two reads. I guess storing nucleic material in water for a long time rather than a buffer makes me a bit queasy.

      Thanks for the feedback!

      Comment

      • vtosha
        Member
        • May 2010
        • 36

        #4
        Do you measure concentration of your samples with another technics? Bioanalyzer not reliable for measuring of concentration.

        Comment

        Latest Articles

        Collapse

        • GATTACAT
          Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by GATTACAT
          Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
          07-01-2026, 11:43 AM
        • SEQadmin2
          Nine Things a Sample Prep Scientist Thinks About Before Sequencing
          by SEQadmin2


          I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

          Here are nine questions we think about, in roughly the order they matter, before...
          06-18-2026, 07:11 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 07-02-2026, 11:08 AM
        0 responses
        25 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-30-2026, 05:37 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-26-2026, 11:10 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 06-17-2026, 06:09 AM
        0 responses
        55 views
        0 reactions
        Last Post SEQadmin2  
        Working...