Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #1

    Covaris S2 cracking tubes.

    Anyone had issues with a Covaris sonicator cracking tubes? We are using the microTUBE AFA Fiber Pre-Slit Snap-Cap 6x16mm in a Covaris S2 that shipped with our SOLiD years ago.

    No problems with it until a month ago when we noticed that the waterbath degasser was not functioning. Applied Biosystems FSE came and replaced the degasser pump motor. Seemed to be fine, except that about 1 in 3 of the micro tubes are cracking at the neck. Obviously not good. We have tried different lots of tubes -- same thing. So I presume it is an instrument issue.

    Also, the 150 bp Covaris protocol seems to be shearing to ~75 bp.

    Any ideas?

    --
    Phillip
  • ETHANol
    Senior Member
    • Feb 2010
    • 308

    #2
    I've sheared about 100 samples and had probably two tubes crack a little bit at the neck where the cap goes on. Nothing to mess up the experiment. I was shearing down to about 100 bp so I think this is about the limit of what the tubes can take. I imagine if you are shearing down to 75 bp, that is more then the tubes can handle. I don't think the amount of energy to shear smaller fragments is linear, so it seems like the machine is dishing out way too much energy.
    --------------
    Ethan

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Yes, that seems to be what Covaris is thinking as well.

      The only other explanation I can think of is that we do seem to have the most problems when the water bath has been degassing for hours. I always presumed that degassing hit a plateau beyond which there would not be much change in the "transparency" of the water to the acoustic energy. But maybe that is not the case? Maybe the instrument is calibrated to deliver the right amount of energy to the tube after 30 minutes of degassing, but after 4-5 hours it is delivering too much because the dissolved gas in the water has dropped much lower and transmits much more energy to the tube.

      --
      Phillip

      Comment

      • Hamid
        Senior Member
        • Sep 2009
        • 108

        #4
        Hi Phillip,

        There are several possible reasons for tubes breaking during processing on a Covaris instrument:
        1. the Covaris microTUBEs are rated for use with a maximum duty cycle of 10% and intensity of 5. Using settings higher than the maximum will increase the probability of tubes breaking.
        2. Incorrect water level. The proper water level when processing samples on the Covaris instruments using the microTUBEs is such that the water level is 1mm below the cap during processing. if the water level is lower, then the probability of tubes breaking will certainly increase.
        3. The built-in intensifier in the holder is no longer straight. The purpose of the intensifier is to prevent the flow of directly waves from the bottom of the glass, and concentrate the waves from the surrounding area. If the intensifier is no longer straight ( perhaps due do damage caused by dropping the holder), then the probability of breaking the tubes will increase. This will also increase the energy reaching the samples, therefore reducing the shearing size range.
        4. degass pump not working as efficiently. The purpose of the degass pump is to get rid of excess solubilized air from the water in the acrylic tank so that cavitation does not occur in the tank. If the degauss pump is not working efficiently, you will notice a shift to a higher size range for shearing. Since large cavitation bubbles do release very large amounts of energy, then that occurring on the outside of the tube on the edge of the tube and the glass might cause tubes breaking.

        Thank you

        Hamid

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Thanks Ethan and Hamid,

          It now appears that our older degassing pump was not doing a good job. As a result, during factory config of the unit, the "calibration factor" was set to a high value, 85%. This worked fine for years, until our pump died and was recently replaced by Applied Biosystems. This new pump, however, was much more effective. So effective that 85% was too high as a "calibration factor". We were hammering our glass micro tubes with too much energy and they were cracking at a high rate, as well as fragmenting our DNA shorter than expected.

          Covaris gave us a new calibration factor, 75%, and instructions on how to input it into the unit EEPROM. Although we have not had a chance to test it extensively, none of the 4 test sample micro tubes cracked and the fragmentation appears to be back in the expected range for the standard protocols. So, looks like that was what happened.

          My advice to anyone having their degassing pump replaced on one of these instruments is to do a few test samples and see that they are in the expected size range. If not you may have to contact Covaris and ask for instructions on how to adjust the calibration value.

          --
          Phillip

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Yesterday, 11:13 AM
          0 responses
          19 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, Yesterday, 10:48 AM
          0 responses
          18 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Working...