    I saw a similar loss of mRNA with the RiboZero Human/Mouse/Rat kit. I wasn't too concerned with the loss of Bioanalyzer peaks but when I looked for hGUS mRNA by qPCR I found a 10-13x fold reduction. Any thoughts on this?

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #17
      Originally posted by whw View Post
      I saw a similar loss of mRNA with the RiboZero Human/Mouse/Rat kit. I wasn't too concerned with the loss of Bioanalyzer peaks but when I looked for hGUS mRNA by qPCR I found a 10-13x fold reduction. Any thoughts on this?
      This really isn't enough information about your experiment to comment intelligently on.

      Do you want to tell us 10-13x fold reduction means? Taken on face value it would mean that the absolute concentration of hGUS in your post-ribozero sample is ~10x lower than in your pre-ribozero sample. That could mean anything, though. It could mean your pre-ribozero RNA is in 200 ul whereas your post-ribozero is in 20 ul. Or they are in the same volume and therefore 90% of your mRNA is lost during ribozero isolation.

      Or you could mean something else entirely. Please, if you are going to post here, give enough information to make it useful to answer you.

      --
      Phillip

      Comment

      • whw
        Member
        • Jun 2011
        • 19

        #18
        Yup, sorry about that was in a bit of a rush yesterday.

        I started with about 800ng of total RNA in 28uL of water and performed RiboZero as written. My final volume after concentration was 40uL which accounts for a small dilution.

        I used nanodrop, bioanalyzer, and qPCR for the housekeeping gene hGUS to assess the quality of my ribo-depleted RNA. Nanodrop was very low between (0.5ng/uL and 1ng/uL) which was fine, but my qPCR showed a 3+ cycle difference between pre- and post-depletion samples. That should correspond to about a 10-fold drop in the amount of hGUS mRNA present. It sounds like some people had a similar experience with the RiboZero plant kit. Bioanalyzer trace post-depletion was flat except for a very small tRNA peak.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #19
          Originally posted by whw View Post
          I started with about 800ng of total RNA in 28uL of water and performed RiboZero as written. My final volume after concentration was 40uL which accounts for a small dilution.
          What method did you use for concentration?

          When we tried using ethanol precipitations it was hit or miss as to whether we recovered any RNA or not. (Which, by the way, is pretty typical for ethanol precipitations of RNA/DNA at low concentrations.)

          When we switched to Zymo columns for RNA concentration, things were much more stable.

          Also, did you do a dilution series on your qPCR? Could be you brought down something that inhibits qPCR, or the qPCR fluor.

          --
          Phillip

          Comment

          • whw
            Member
            • Jun 2011
            • 19

            #20
            I've tried both Zymo columns and ethanol precipitation with similarly disappointing results.

            I did run a dilution series on the qPCR but there was no sign of inhibition.

            Thanks for your help, I just spoke with RiboZero and they sent me a replacement kit. I'll see if that fixes anything.

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #21
              Good luck.

              But I'm not sure that a ~15% yield of ribo-depleted RNA is that bad. The nanodrop is pretty much useless for that purpose. Did you do a pico RNA chip or a nano RNA chip after RiboZero.

              The problem with plants seems to me to be that they just seem to have a higher rRNA/mRNA ratio than animals. So just getting a flat nano RNA chip result is not diagnostic.

              That said, we have not tried the new magnetic bead ribo-zero kit. Hope that is not the issue.

              --
              Phillip

              Comment

              • whw
                Member
                • Jun 2011
                • 19

                #22
                I'd have to agree that the nanodrop is pretty unhelpful with such small amounts of RNA.

                I ran a pico chip. I'm just disappointed to see a loss in mRNA with a protocol that is meant to be rRNA specific. As for the magnetic bead system, I tried essentially the same procedure with my colleague's RiboZero Gold (magnetic) and recovered almost all the mRNA. Hopefully I just got a bad kit.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #23
                  Well, for what it is worth, we have tried lots of kits, both polyA and ribo-capturing ones and only two of them worked at all for us. The one embedded in the TruSeq RNA kit -- that polyA+ magnetic bead methodology works pretty well. And the Ribozero columns work for us. The few times we have needed them.

                  Anyway, hope it works out for you. Please post your results, if you would be so kind.

                  --
                  Phillip

                  Comment

                  • rndouglas
                    Member
                    • Jun 2012
                    • 23

                    #24
                    As an update...

                    I FINALLY received some of my rRNA-depleted samples back from our DNA core this week.

                    They look pretty good, and they have almost no rRNA present. The first two samples came back with 0.26% and 0.05% rRNA. Much much much better than the ~30.0% rRNA contamination I had with Invitrogen's kit.

                    Comment

                    • Francy87
                      Member
                      • May 2013
                      • 28

                      #25
                      Hello!
                      I'm having similar issue with my bacterial RNA samples. I don'tt know how my total RNA looked like before the the RiboZero depletion, since I isn't run it on the BioAnalyzer, but after the rRNA depletion it looked like in the BA traces attached.

                      The sequencing core said I can't go ahead with the library preparation… But I saw on the RiboZero website that after the purification, I should have only a peak at 100-140nt, that is what I have.
                      Any comments about this?
                      Attached Files
                      Last edited by Francy87; 07-02-2014, 10:32 PM.

                      Comment

                      • aubrie
                        Junior Member
                        • Oct 2016
                        • 4

                        #26
                        Assistance please

                        2100 expert_Prokaryote Total RNA Nano_DE72904769_2016-10-07_11-33-43.pdf

                        2100 expert_Prokaryote Total RNA Pico_DE72902741_2016-10-11_14-40-24.pdfHi,

                        I am working with RNA from e.coli k12 strain. I have flash frozen the samples and then stored them in the minus 80, there is a control and then treatments in triplicate with various antibiotics. I have attached my bioanalyzer profiles for the total RNA extraction (of interest are the ct br1, ct br2, and ct br3) and then the profiles for ribosomal depletion (also for controls, control 1, control 2, control 3 in a 1:1 and 1:10 and 1:100). I have included an electropherogram zoom of control br2 at 1:10 dilution.

                        According to a previous post and consulting the ribozero FAQs page, the depletion profile is okay and the peak around 100-150 bps is tRNA, 5S and other small RNAs. I am worried that this is a combination of the residual degraded RNA and I am not sure how to subtract the small RNA peak from my sample in order to move forward with library prep. I am also wondering if I even should move forward. The alternative is to go back to the tried and true trizol extraction which we have not automated.

                        Thank you in advance for your feedback.
                        Attached Files

                        Comment

                        • adam.geber
                          Member
                          • May 2014
                          • 40

                          #27
                          Your 100-150bp peak is pretty sharp, which makes me think that it's not just degraded RNA and is in fact the 5S rRNA and tRNA mentioned in the RiboZero kit. Did you run a negative control for the depletion (i.e. no capture probes) and if so how does it appear?

                          Can I ask what you want to get out of RNA-seq? Are you interested in noncoding RNAs or only mRNA? Also, how were these samples extracted?

                          Finally, what are your sample concentrations after rRNA subtraction? If they're high enough you could easily do a silica column clean-up (e.g. Zymo, Qiagen) to select for those transcripts <200nt. Depending on whether you want to start fresh, you could even incorporate that step into your extraction protocol if you're using a column-based method.

                          Comment

                          • aubrie
                            Junior Member
                            • Oct 2016
                            • 4

                            #28
                            The concentrations for the samples (including the monster peak) are 20-24ng/ul in 16ul.

                            I did not run a negative control. If I did I assume this would be an amount of the tRNA with beads and no probes, simply go through the protocol with no probes?

                            I am interested mostly in the mRNA. From what you are saying about the clean up column, I gather that I could run the clean up on the rRNA depleted sample then re run the pico chip?

                            Sounds like a foot in the right direction. Thank you. I'll try it.

                            Comment

                            • adam.geber
                              Member
                              • May 2014
                              • 40

                              #29
                              Those are quite decent concentrations depending on your input requirements downstream -- have you decided on a prep method?

                              I think the best negative control would be an RNA sample split in half, with one aliquot taken through the RiboZero protocol without the inclusion of probes and then compared with the other aliquot by the Tapestation. The size of your undepleted rRNA peaks will certainly make it difficult to assess whether the abundance of your 100-150bp is affected by the capture protocol (e.g. produced by degradation) but you could at least be more sure that you're removing tRNA/5S rRNA in a clean-up process.

                              Comment

                              • aubrie
                                Junior Member
                                • Oct 2016
                                • 4

                                #30
                                I'm thinking about putting the rRNA depleted sample through the RNeasy clean up to get rid of the small RNAs below 200bp. I have sample to spare and I'm curious to see the concentration of "mRNA" that is being masked.

                                We don't have the Zymo columns on hand. I know those are really good.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                38 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                28 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-15-2026, 10:52 AM
                                0 responses
                                41 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                45 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...