  • farrel75
    Member
    • May 2012
    • 11

    #1

    Let's Talk About TruSeq

    Hi everyone,
    My lab and I are in the process of taking our maiden voyage on the MiSeq and we are trying to make the process as cost effective as possible. We are considering having me do the library prep rather than have it sent to the core facility. We need more than the 12 barcodes we get from the A kit but we don't need the entire B kit.
    So the question is: What about TruSeq? Can we keep it in the fridge for an extended amount of time? Is it a one time use thing? Any suggestions??
    Thanks!
  • SS00
    PhD Student
    • Jun 2012
    • 33

    #2
    Do you mean storage of the kit? You can store the kit in the freezer (not the fridge!) until the expiry date which is written on the top of the box.

    If you are not going to use the whole kit at once you need to aliquot End Repair Mix, A-tailing Mix, and all the Controls (if you use them) into single use aliquots to avoid freeze thaw.

    Once you have done this, it should be fine to revisit your kit as many times as you need.

    Comment

    • TonyBrooks
      Senior Member
      • Jun 2009
      • 303

      #3
      We used a TruSeq RNA kit that was 9 months passed use-by date and it still produced excellent libraries. We followed SS00's advice and aliquoted the reactions out into smaller volumes when the kit arrived and stored everything at the recommended temperature (written on each individual tube).
      We don't use Ampure beads that are out of date because we've found they don't keep so well.
      There are also alternative prep kits that are supplied in smaller pack sizes which may be more suitable if you don't want the hassle of aliquoting everything out (Bioo, NEB). Personally, I'm not so keen on the TruSeq DNA kits as size selection is a real problem. The TruSeq RNA kits are excellent though.

      Comment

      • Rocketknight
        Member
        • Sep 2011
        • 86

        #4
        @TonyBrooks What size-selection problems have you had with the Truseq DNA kits? Do other kits use different size-selection methods?

        Comment

        • TonyBrooks
          Senior Member
          • Jun 2009
          • 303

          #5
          We've never gotten the post ligation gel purification to run properly. Libraries run too slowly with EthBr and too fast with SYBR. A 400-500 excision either gives us libraries >600bp or 300-400bp. We could probably optimise it, but we don't have the time.
          We didn't see this problem with the older Illumina PE kit or with NEB kits. I also think, the latest NEB kit has a protocol for gel-free selection, giving libraries ~ 350bp. It's not as tight as gel urification, but takes minutes, not hours.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Yesterday, 11:13 AM
          0 responses
          21 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, Yesterday, 10:48 AM
          0 responses
          18 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          36 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          42 views
          0 reactions
          Last Post SEQadmin2  
          Working...