    Okay, I think you are getting ahead of yourself here. The simplest explanation for the result you posted initially is that your RNA is degraded. But it looks strange enough that it would be worth rerunning it. But for TruSeq DNAse treatment is recommended, so you might as well do that first -- if you clean everything up with a column. Then you can take a look at it on a chip again.

    But, the most likely outcome is: degraded RNA, start again.

    One step off the path as designed by Illumina, I'm willing to chip in suggestions. Two steps -- okay I'm still listening, not sure anything I say is going to be of much help. But you are 3 steps out now and on your own.

    Let us know if it works, but you (1) have circumvented the normal QC step by declaring its failure was the result of genomic DNA in your sample -- which would not be there is you had not (2)also skipped a step asked for by the protocol by not DNAsing and now you (3)think that the fragmentation step should be sufficient to inactivate any DNAse still around after the polyA+ isolation. So, I can't say it will fail, but I don't like your chances.

    The big problem is that the most likely failure mode, high 3' bias caused by low intactness of your total RNA, will not be obvious until you complete mapping back to a reference genome.
    --
    Phillip

    Comment

    • eab
      Member
      • May 2011
      • 66

      #17
      One more thing to factor in: these are irreplaceable human samples. There is no starting again.

      So...I need to squeeze the best possible data out of them. If these data don't live up to my wildest dreams, I can accept that. 3' bias? Not the end of the world. I gain nothing by throwing the samples away.

      Another consideration I have not made totally clear: there is very little RNA here. I do not like the idea of column purifying for that reason.

      Definitely a mistake not to do on-column DNAse digestion, no argument from me there. I use RNAzol for my extractions, and this kind of contamination has not been a problem for me before. In this case, I have been trying to help a colleague make libraries from cells sorted into RLT buffer. Should it come up again, I shall know that the DNAse treatment is mandatory.

      But, back to these samples. To me, there is just no way that a hump of material between two intact-looking rRNA peaks can be degraded RNA. If degraded material is going to pile up in a single hump, that better be a low-MW hump, not several Kb in length. I am sufficiently convinced by 1) the absence of a solid alternative explanation and 2) the similarity to the pics I sent out previously (see http://www.chem-agilent.com/pdf/5989-0991EN.pdf, especially page 3, item B) that I am going to proceed as if it's genomic DNA. And yes, I do think that the 95-degree, 6-min fragmentation step should be sufficient to kill any DNAse that manages to persist through two rounds of dynabeads purification (which to my mind isn't so different from a column purification, anyway).

      But of course I could be wrong. The plan is to pool aliquots of residual material I had frozen separately for the Bio-A, and make one single library according to my made-up method. I will post results either way.

      Thanks all for your advice,
      Eli

      Comment

      • kerplunk412
        Senior Member
        • Jun 2012
        • 119

        #18
        Dynabeads and DNAse?

        I don't think it is a good idea to have active DNAse during the dynabeads procedure- aren't the oligos made of DNA? If so this will ensure that you don't recover any RNA, as the DNAse will degrade the oligos so there will be nothing for your RNA to hybridize to. Also, you run the risk of seeing lots of oligo dT in your final library. Maybe I missed something?

        Comment

        • eab
          Member
          • May 2011
          • 66

          #19
          nope, don't think you missed anything. yeah, the "d" in "oligo-dT Dynabeads" is, like, the same d in DNA. embarrassing.

          thank you!

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by SEQadmin2


            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.


            Here are nine questions we think about, in roughly the order they matter, before...
            Yesterday, 07:11 AM
          • SEQadmin2
            From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
            by SEQadmin2


            Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


            The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
            ...
            06-02-2026, 10:05 AM
          • SEQadmin2
            Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
            by SEQadmin2


            With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


            Introduction

            Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
            05-22-2026, 06:42 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 06-17-2026, 06:09 AM
          0 responses
          16 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-09-2026, 11:58 AM
          0 responses
          37 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-05-2026, 10:09 AM
          0 responses
          43 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-04-2026, 08:59 AM
          0 responses
          49 views
          0 reactions
          Last Post SEQadmin2  
          Working...