    Blue Pippin problem with TruSeq DNA libraries

    Hi. Our genomics core recently received a new Blue Pippin instrument. Another user and myself each prepared TruSeq DNA whole genome libraries (TruSeq DNA sample Prep kit, v2) and attempted to do a tight size selection with 1.5% cassettes for 450 bp targets. Yet the Bioanalyzer showed our final libraries were only 300 bp. I've since contacted Sage and they noted that the TruSeq Y-shaped adapters alter migration, so that users would select a range of 500-600 bp if they want 450 bp (but also noted this range as "approximate"). Has anyone else done this, esp with the new Blue Pippin that uses internal standards (ie no separately loaded EtBr ladder). I am nervous about losing more libraries testing this out, and given our other alternative (traditional agarose gels) will not be as precise, I really hope we can make the Blue Pippin work. Thank you in advance for any insight.
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Originally posted by Hilary April Smith View Post
    Hi. Our genomics core recently received a new Blue Pippin instrument. Another user and myself each prepared TruSeq DNA whole genome libraries (TruSeq DNA sample Prep kit, v2) and attempted to do a tight size selection with 1.5% cassettes for 450 bp targets. Yet the Bioanalyzer showed our final libraries were only 300 bp.
    The Bioanalyzer chip is post amplification?

    BTW, I have seen a slide from the Broad that claimed a similar shift in apparent MW caused by ligase continuing to bind to the ligation site. Further they were able to overcome the effect by removing the ligase. (Using proteinase K, IIRC.)

    I guess you could just do Pippinprep post amplification. That way you have plenty of library and any Y-adapter/ligase-mediated shifts in apparent MW would be behind you.

    --
    Phillip

    Comment

    • Hilary April Smith
      Member
      • Jul 2011
      • 22

      #3
      Yes; the Bioanalyzer chips were post-amplification.

      Interesting that continued ligation could cause a problem. We do use a mix (Illumina's Proprietary "Stop Ligation Buffer") to terminate the ligation, so I don't think that is the problem. Great idea on doing the Blue Pippin prep after the PCR. Do you know if that typically causes any other problems? There are 2 AMPure bead cleanups after the ligation, so I assume that should remove most of the adapters before the PCR ... but I was not sure if the potential for any carry-over in un-ligated adapters could cause a problem for the PCR. (I've meanwhile emailed Sage Science and Illumina to see if they have any recommendations, but I find that sometimes the responses are helpful and others not ... so I'm very grateful for help from an experienced user!).

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Depends on how tightly bound the Ligase is to an amplicon. During ligation, my understanding is that the ligase is covalently linked with both ends that it is joining. If it remains complexed (with or without stop solution) then it might drag the DNA-ligase complex away from the electrode towards which the DNA is migrating.

        Of course this is all just an interpretation of various outcomes the researcher at the Broad saw. Could be something else going on. Actually not even helpful unless you feel like firing up the proteinase K, post-ligation. Even then probably a little risky.

        --
        Phillip

        Comment

        • Hilary April Smith
          Member
          • Jul 2011
          • 22

          #5
          Sage Science noted that using the Pippin after PCR would avoid needing to optimize (which I gather is a bit of trial $ error) the size you tell the program to select. Yet Illumina, when contacted, is saying to not deviate from their protocol. I like your idea of trying the Pippin gel selection after the PCR and we may try that. Thank you again.

          Comment

          • yaximik
            Senior Member
            • Apr 2011
            • 199

            #6
            I cannot cite exactly the reference document now, but Illumina's recommendation on use of the gel sizing method (it is on their web site) mentions that selection results can be skewed by the way staining is made (pre-, in gel, or post-stain).

            Comment

            • Hilary April Smith
              Member
              • Jul 2011
              • 22

              #7
              Hi. Yes. We have low input DNA (we've been running this with 150 ng; we may be able to increase to ~300 ng this time). Thus we're trying to find an alternative that will give higher yield, and I think the Pippin preps can increase both quantity of yield and precision of the cut. Yet the best I can get from Sage is that an approximate range of 500-600 bp, which sounds like it may need optimization, is often used with the TruSeq DNA kits to get the 450 bp band (ultimate aim: 100 bp PE HiSeq reads). Since we're limited by starting material and already destroyed several samples due to not realizing that we needed some sort of an offset (the ~100 bp offset Sage seems to suggest or ???), I'm trying to see if there's a better route than trial & error to determine the optimal offset (esp. since I also need to produce these quickly; the need for obtaining replacement samples after the first run has already delayed the project more than anticipated...). Illumina seems to keep saying to follow their protocol Exactly ... but given they have both a gel-method and a gel-free method (the latter with Exome enrichment), it would seem some modifications are allowed.

              Comment

              • HESmith
                Senior Member
                • Oct 2009
                • 512

                #8
                Hi Hilary,

                A previous poster resolved the issue of aberrant migration with four cycles of PCR prior to size selection. (Phillip, do you recall who it may have been? Ethanol, perhaps?) It's enough cycles to eliminate the Y ends, but not so many that your library becomes biased for smaller products (like adapter dimers). The extra step is a hassle, but it works, and, since you'll recover more material, you can reduce the final amplification by the same number of cycles.

                Hope that helps,
                Harold

                Comment

                • Hilary April Smith
                  Member
                  • Jul 2011
                  • 22

                  #9
                  Dear Harold,
                  That is great! I'll see if I can backtrack and find the protocol or earlier SeqAnswers thread as I've found this all very informative. I was debating whether to do the PCR before or after (following PMiguel's suggestion) and wouldn't have thought of doing it both times. Thank you so much for your help.
                  Best,
                  Hilary

                  Comment

                  • HESmith
                    Senior Member
                    • Oct 2009
                    • 512

                    #10
                    Hi Hilary,

                    Found it in ETHANol's ChIP-Seq protocol; the link is http://seqanswers.com/forums/showthread.php?t=13093.

                    Good luck,
                    Harold

                    Comment

                    • Hilary April Smith
                      Member
                      • Jul 2011
                      • 22

                      #11
                      Excellent; thank you very much!

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #12
                        Harold and Hilary,
                        Actually we normally only do about 4 cycles of PCR for a typical DNA TruSeq library. Depends on how things go, but generally 10 cycles is vast overkill and will give you those irritating "bubble product" peaks.

                        I happened to talk to someone from Sage Science 2 days ago. They said the Y-adapter migration artefact did not happen in the absence of the ethidium bromide in the gel. So, if you have the blue pippen, you should be able to use the ethidium bromide-free cassettes.

                        But the main reason not to do your amplification prior to size selection is that it will allow some amount of any adapter dimers to anneal to full length library molecules and, hence, avoid size selection. (My observation.) If you don't see an adapter dimer peak prior to size selection, it probably is not an issue for you.

                        --
                        Phillip

                        Comment

                        • HESmith
                          Senior Member
                          • Oct 2009
                          • 512

                          #13
                          Originally posted by pmiguel View Post
                          But the main reason not to do your amplification prior to size selection is that it will allow some amount of any adapter dimers to anneal to full length library molecules and, hence, avoid size selection.
                          We haven't observed this phenomenon, even with ChIP-Seq libraries (where adapter dimers are more frequently a problem). Not sure why it's a problem for you; perhaps it's platform-specific? I'd be interested to hear Hilary's results.

                          -Harold

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            Originally posted by HESmith View Post
                            We haven't observed this phenomenon, even with ChIP-Seq libraries (where adapter dimers are more frequently a problem). Not sure why it's a problem for you; perhaps it's platform-specific? I'd be interested to hear Hilary's results.

                            -Harold
                            Well, you may not agree with my interpretation of what is happening. But my assay for this phenomenon is seeing adapter dimer sequence in a library that shows no 120 bp peak on a high sensitivity chip after enrichment PCR + Ampure clean up. Since I don't see that peak, I attribute it to ssDNA adapter dimer annealed to library molecules. I posted about it in another thread.

                            I don't expect it to be a big issue unless during ligation you ended up with a substantial fraction (>5%) of your products being adapter dimers. Given the safe-guards present in the TruSeq kit this is only likely to occur in cases where insert is limiting or an phosphorothioate-bond-competent exonuclease gets carried into the ligation step. (I think T4 DNA polymerase is such an enzyme.)

                            To tell you the truth, I have mainly seen this phenomenon in RADseq libraries made in another lab -- not using TruSeq kits at all. So its prevalence may be rare with TruSeq libraries. In point of fact, I don't hesitate to do size selection after enrichment PCR. But everything else being equal, doing size selection prior to enrichment PCR seems preferable to me.

                            (Of course everything else is not equal. The Y-adapter and/or persisting ligase-binding may make doing a pre-enrichment size selection untenable.)

                            --
                            Phillip

                            Comment

                            • HESmith
                              Senior Member
                              • Oct 2009
                              • 512

                              #15
                              Hi Phillip,

                              I don't dispute your interpretation: if it's in the sequence data, then it must be in the library .

                              As I said, we have not experienced this problem, but our sample set is limited to TruSeq libraries (albeit a large number of them). Your post suggests that the problem IS platform-specific (i.e., limited to RADseq and/or non-TruSeq libraries), which is good to know.

                              -Harold

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                How Immunogenomics Decodes Immunity’s Genetic Blueprint
                                by SEQadmin2




                                The immune system’s power comes from its genetic diversity, allowing myriad threats to be neutralized through first recognizing foreign antigens. That diversity is also what makes the immune system so difficult to study. Recent advances in sequencing technology and computational biology, however, are giving researchers new tools to understand immune responses and immune-related diseases in greater detail.

                                This convergence of genetics, immunology, and computation...
                                Yesterday, 05:41 AM
                              • SEQadmin2
                                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                by SEQadmin2



                                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                ...
                                07-31-2026, 11:01 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 08-24-2026, 10:32 AM
                              0 responses
                              44 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 08-20-2026, 11:17 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 08-18-2026, 10:05 AM
                              0 responses
                              55 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 08-13-2026, 12:22 PM
                              0 responses
                              50 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...