Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • kylini
    Member
    • Jul 2010
    • 13

    #1

    Does Glycine actually quench Formaldehyde?

    EDIT: See post #15. Glycine does not quench, but Tris does!
    EDIT 2: Most figures removed in anticipation of publication.

    Summary: 125 mM glycine, 333 mM glycine, and 750 mM glycine do not quench 333 mM (1%) formaldehyde. 750 mM Tris does.
    Last edited by kylini; 11-16-2012, 07:55 AM.
    PhD Student | Molecular & Cellular Biology | The University of Iowa
  • kylini
    Member
    • Jul 2010
    • 13

    #2
    ..........
    Last edited by kylini; 11-16-2012, 07:50 AM.
    PhD Student | Molecular & Cellular Biology | The University of Iowa

    Comment

    • FiReaNG3L
      Member
      • Apr 2012
      • 17

      #3
      Are you using methanol-free formaldehyde from single use ampoules?

      Comment

      • kylini
        Member
        • Jul 2010
        • 13

        #4
        Originally posted by FiReaNG3L View Post
        Are you using methanol-free formaldehyde from single use ampoules?
        Yes. I made the switch about a month ago. That said, I get the exact same "no difference between 20 sec and 10 min" with methanol-containing CH2O. My thought is that while the methanol increases permeability, the glycine is meant to quench the unused formaldehyde still in the media. If it's not quenching, it doesn't matter as CH2O will still get in just fine.

        There has to be a condition of the quenching reaction that is obvious that I'm missing. It would make no sense for every lab in the world to use a method that doesn't work.
        PhD Student | Molecular & Cellular Biology | The University of Iowa

        Comment

        • FiReaNG3L
          Member
          • Apr 2012
          • 17

          #5
          Try quenching with 250 mM Tris just to see? Maybe your glycine is bad or something.. although thats improbable.

          Comment

          • kylini
            Member
            • Jul 2010
            • 13

            #6
            I'll definitely give Tris a try. I also thought about a very real possibility:

            Many protocols concentrate their cells prior to crosslinking (pellet and resuspend in PBS). They then use a small volume of CH2O to achieve the desired 1%. In short, large number of cells, small number of CH2O molecules.

            I don't concentrate my cells and crosslink in media as I have data that suggests that transcription changes after such a perturbation. Therefore, I need a larger volume of CH2O to reach 1%, a much larger volume actually. Since the cell number is the same, I'm therefore drastically increasing the CH2O:cell ratio even if the concentration is technically the same. Concentrations don't crosslink; molecules do.

            In any case, a crosslinking concentration assay never hurt anyone!
            PhD Student | Molecular & Cellular Biology | The University of Iowa

            Comment

            • FiReaNG3L
              Member
              • Apr 2012
              • 17

              #7
              Thats a pretty good idea actually, I think the rationale behind using less glycine in molar than formaldehyde was that most of it would be spent on actually fixing proteins / dna in the cells. Now if you have less cells but much more volume of Formaldehyde, I can see how its not sufficient to quench everything. How much volume for how many cells if I can ask?

              Comment

              • kylini
                Member
                • Jul 2010
                • 13

                #8
                ...........
                Last edited by kylini; 11-16-2012, 07:51 AM.
                PhD Student | Molecular & Cellular Biology | The University of Iowa

                Comment

                • theduke
                  Member
                  • Aug 2010
                  • 14

                  #9
                  This is not the first time I have seen shearing like this using the Covaris. I think the glycine thing is a "red herring" here and I would suspect your shearing methodology is the problem. Have you tried any other sonicators? I would hope so after two years.
                  Last edited by theduke; 10-31-2012, 12:25 PM.

                  Comment

                  • kylini
                    Member
                    • Jul 2010
                    • 13

                    #10
                    Originally posted by theduke View Post
                    This is not the first time I have seen shearing like this using the Covaris. I think the glycine thing is a "red herring" here and I would suspect your shearing methodology is the problem. Have you tried any other sonicators? I would hope so after two years.
                    I use a Fisher Sonic Dismembrator 550 and, recently, have confirmed that I am able to use the Covaris. Both sonicators result in the same profile and both were used for quenching experiments.

                    My sonication certainly isn't yet optimal, but it still doesn't change the result that adding 125 mM or 333 mM glycine BEFORE adding formaldehyde yields the SAME crosslinking efficiency as CH2O exposure without it. I have no reason to trust that glycine is actually quenching formaldehyde and this is why I need input.
                    PhD Student | Molecular & Cellular Biology | The University of Iowa

                    Comment

                    • theduke
                      Member
                      • Aug 2010
                      • 14

                      #11
                      I don't think there is much doubt that glycine quenches formaldehyde. I would drop that line of thinking quickly if I was you. There is too much literature and too much evidence against you and it can't all be wrong. There is also no doubt that fixing time per se influences shearing efficiency. Something else is going on with your experiments. Either your glycine is off/concentration is wrong or something is up with your formaldehyde or your technique. It looks to me that you are massively over-crosslinking.

                      Generally, I have 10^8 cells per 200 ml, up to 6*10^8 cells per 1.2 L. I grow HeLa cells at a concentration of 5*10^5 cells/ml until I'm ready and then dump them into a vessel already containing the requisite 16% formaldehyde (final conc. 1%).

                      This is where I think your problems are. Something doesn't smell right. That is a ****-ton of cells and a lot of volume. Are you doing the quenching etc in suspension?

                      Comment

                      • kylini
                        Member
                        • Jul 2010
                        • 13

                        #12
                        That is correct. [omitted]

                        The reason I'm so concerned about the quenching is that 1% = 10 g per liter. Since CH2O has a molecular weight of about 30, it follows that there are 30 g per mol. 10g/L times mol/30g = 333 mM. How can something that reacts 1:1 to stop formaldehyde from crosslinking further possibly work at a 333 mM to 125 mM ratio?

                        Again, I added 125 mM glycine FIRST before formaldehyde and the cells went right on to over-crosslink (all of the material was crosslinked). I then added 333 mM glycine FIRST before formaldehyde and the cells STILL went right on to over-crosslink (again, all material was crosslinked). If I'm crosslinking in media, there should be a bit of extra quenching power from its components. Despite this, the amount of glycine I'm adding doesn't actually quench.

                        If quenching doesn't actually happen, that means my cells are crosslinking while I aliquot them into 50 ml tubes (another 5-10 minutes) and while they pellet (another 5 min). Instead of crosslinking for 30 seconds, I'm really crosslinking for 30 minutes, hence the reason my sonication sucks.

                        I can't drop the line of thinking that glycine isn't quenching formaldehyde in these conditions because I have reproducible, empirical evidence showing it doesn't and it explains my sonication failures despite two years of optimization using protocols directly from the labs of Richard Young and the ENCODE Consortium. That's why I think that there are other factors at play, such as the number of molecules (not the concentration) per cell.
                        Last edited by kylini; 11-16-2012, 07:52 AM.
                        PhD Student | Molecular & Cellular Biology | The University of Iowa

                        Comment

                        • theduke
                          Member
                          • Aug 2010
                          • 14

                          #13
                          Originally posted by kylini View Post
                          I can't drop the line of thinking that glycine isn't quenching formaldehyde in these conditions because I have reproducible, empirical evidence showing it doesn't and it explains my sonication failures despite two years of optimization using protocols directly from the labs of Richard Young and the ENCODE Consortium. That's why I think that there are other factors at play, such as the number of molecules (not the concentration) per cell.
                          Well, I don't know what career-stage you are at, but spending two years figuring out sonication and ChIP-Seq is far, far, far too long in my opinion. At the very least, I would have moved on to MNase a long time ago in your situation. Actually a compelling argument can be made for the preferential use of MNase over sonication these days, but that is a different discussion.

                          I would start thinking about optimization experiments in small dishes and then work up from there. Have you done this? I don't like the large number of cells and excessive volumes you are using. I typically do my x-linking in 10 cm dishes with 5 ml media at RT. I usually do multiple dishes at the same time but do not combine them into one massive container.

                          ENCODE ChIP-Seq data is far from perfect!!

                          Comment

                          • kylini
                            Member
                            • Jul 2010
                            • 13

                            #14
                            Originally posted by theduke View Post
                            Well, I don't know what career-stage you are at, but spending two years figuring out sonication and ChIP-Seq is far, far, far too long in my opinion.
                            Don't I know it! I've definitely been pursuing other experiments on the side and this is definitely not the only thing my lab does. It'd just be nice to actually do these experiments in house and not rely on collaborators. Notably, I already have samples sequencing despite the sonication (horray for size selection after library prep) so this isn't holding me up. Other fun deep-sequencing projects are in the works too!

                            It's just that this glycine first experimental result goes completely against established protocols. I'm sure the number of molecules of formaldehyde vs. the number of cells (ignoring volume) is part of the problem. I was curious if anyone else knew what could be going on.
                            PhD Student | Molecular & Cellular Biology | The University of Iowa

                            Comment

                            • kylini
                              Member
                              • Jul 2010
                              • 13

                              #15
                              Fun update!

                              Here's the edited experimental setup:

                              17.6 million HeLa cells were treated for 30 seconds with 1% methanol-free formaldehyde premixed for 3 minutes with a 3-fold molar excess of either glycine or Tris.

                              Here's the result:



                              In conclusion, glycine does not quench formaldehyde but Tris does.

                              [omitted]

                              I think this settles my problem. If you have quenching issues, give Tris a try! It's certainly been used in the literature; this just confirms that it works.
                              Last edited by kylini; 11-16-2012, 07:54 AM.
                              PhD Student | Molecular & Cellular Biology | The University of Iowa

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              42 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              32 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-15-2026, 10:52 AM
                              0 responses
                              44 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...