    ribosome depletion causing ribosome degradation?

    basic question concerning use of gram-negative Ribozero ribosome-depletion kit:

    has anyone noticed this system create what appears to be significant degradation of ribosomes?

    (input total RNA was in TE, following exactly provided protocol)

    from attached bioanalyzer trace:
    blue trace: RNA pre-Ribozero treatment, RIN = 9 23s/16s =1.6
    red trace: apparent degraded ribosomes?
    Attached Files
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    Originally posted by valid_name View Post
    basic question concerning use of gram-negative Ribozero ribosome-depletion kit:

    has anyone noticed this system create what appears to be significant degradation of ribosomes?

    (input total RNA was in TE, following exactly provided protocol)

    from attached bioanalyzer trace:
    blue trace: RNA pre-Ribozero treatment, RIN = 9 23s/16s =1.6
    red trace: apparent degraded ribosomes?
    First, your plots are not correctly aligned (the spike-in control peaks do not overlap.)

    Second, yeah I am not so sure that hill is ribosomal. Not what I would expect for degraded rRNA. Could be though.

    --
    Phillip

    Comment

    • valid_name
      Junior Member
      • Dec 2012
      • 2

      #3
      Originally posted by pmiguel View Post
      First, your plots are not correctly aligned (the spike-in control peaks do not overlap.)

      Second, yeah I am not so sure that hill is ribosomal. Not what I would expect for degraded rRNA. Could be though.

      --
      Phillip
      ok... new user of bioanalyzer RNA! let's see if this helps any. with the peaks aligned, it looks like substantial 23s degradation; also spike at 200nt?

      thanks for bearing with such elementary questions!
      Attached Files

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        Originally posted by valid_name View Post
        ok... new user of bioanalyzer RNA! let's see if this helps any. with the peaks aligned, it looks like substantial 23s degradation; also spike at 200nt?

        thanks for bearing with such elementary questions!
        Are you sure that you did not concentrate the depleted sample?

        If 23S degrades you expect degradation products smaller than the original peak. Instead we see a hill on both sides of where the 23S was. Maybe that is the mRNA?

        --
        Phillip

        Comment

        • scooter
          Member
          • Feb 2010
          • 29

          #5
          Not degraded after Ribo-Zero

          Actually, your RNA post-RZ looks exactly like it should. The higher molecular weight hump is like your mRNA and the lower set of peaks is likely to be abundant smaller RNAs like snoRNAs, tRNA, etc.....

          If you call up Epicentre tech support they will likely tell you that they get calls with this issue all the time. Ribo-zero is so effective at removing rRNA that many people assume the worst at first glance. mRNA is only ~1-2% of a total RNA sample, so if you are removing ~99% of the rRNA you would not expect to see much on a BA trace after removal.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
            by SEQadmin2



            CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

            Despite this, “CRISPR helped turn genome editing from a specialized technique into
            ...
            07-31-2026, 11:01 AM
          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 08-06-2026, 07:41 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 08-03-2026, 10:13 AM
          0 responses
          39 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-31-2026, 02:55 AM
          0 responses
          44 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-24-2026, 12:17 PM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Working...