Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • ChipBrian
    Junior Member
    • Jan 2013
    • 3

    branson S250 shearing crosslinked DNA for ChIP

    I've been trying to shear formaldehyde-crosslinked chromatin using a Branson S250 for ChIP-Seq, and I've been having some problems with attaining the correct size distribution.

    I tried reading all of the posts related to shearing, so if someone could point me in the right direction, that would be great.

    Basically, I keep getting a bimodal size distribution with a cluster at ~2.5 kb that is really recalcitrant to being sheared. I can break it up with longer shearing times, but at this point my fragments are too small (mostly <300 bp). I heard that one should have a tight size distribution (200-800 bp) for going on to the ChIP and library prep.

    I've attached a shearing time course using Bioanalyzer.

    Here are the details:

    I've been following a procedure from a 2006 Nature Methods paper from Richard Young's group for the crosslinking (10 minutes in media with 1% methanol-free formaldehyde at room temp.) and lysing conditions (nuclear isolation).

    The shearing conditions are:

    1. 15 million cells in ~1 mL of sonication buffer (10 mM Tris-HCl, 1 mM EDTA, 0.5 mM EGTA, 100 mM NaCl, 0.5% N-laurylsarkosyl, 0.1% Deoxycholate).

    2. 45% amplitude, 0.7 s pulse on, 1.3 pulse off, then timecourse.

    I tried using SDS (0.1% or 0.2%) instead of N-laurylsarkosyl with not much success. I was thinking about going to 1%.

    Any help would be greatly appreciated!!!
    Attached Files
  • Hamid
    Senior Member
    • Sep 2009
    • 108

    #2
    Hi Brian,

    What you are noticing is due to one of the following reasons:
    1. over-fixed chromatin. when chromatin is over-fixed areas with high protein content tend to over-fix and is quite resistant to shearing. On a bioanalyzer trace it often appears at a peak of 1-7kb. You can resolve this by reducing the fixation time. Since you are using a high power probe sonicator, your choice of reducing the fixation might be limited since the epitopes will be destroyed otherwise. Do you have access to a Covaris instrument? If you let me know where you are located, I can see if there is an instrument at your institution you can use.
    2. Cell culturing. damaged and dead cells tend to fix more easily, and will contribute to the peak that you are noticing.
    3. use of High sensitivity bioanalyzer chip. use regular sensitivity 12k chip only. The high sensitivity chip always seems to show secondary peaks which do not appear on regular sensitivity chips, or agarose gels.

    I have attached a bioanalyzer traces that illustrate effects of over-fixation, and one that show the gradual reduction in size and it the fragment distribution of Covaris fragmented chromatin.

    Thank you

    hamid
    Attached Files
    Last edited by Hamid; 01-30-2013, 12:31 PM.

    Comment

    • kwaraska
      Senior Member
      • Nov 2008
      • 131

      #3
      Brian-
      I don't know where in Boston you are but we at the BioPolymers Facility at Harvard Med School have a Covaris instrument that Hamid talks about.

      Please go to our website if you want more information

      The Biopolymers Facility at Harvard Medical School provides molecular biology services to the research community.

      Comment

      • ChipBrian
        Junior Member
        • Jan 2013
        • 3

        #4
        Thanks for the suggestions and comments; it is much appreciated.

        Is the peak from the high sensitivity DNA chip an artifact? The peak seems to be broken down after awhile with more shearing.

        I saw the earlier posts on over-crosslinking and the effects of methanol and media on the rate of crosslinking.

        I did a little cross linking timecourse study for our protein of interest (it doesn't directly contact DNA) using ChIP-qrtPCR. I did 5 min crosslink in PBS (after 1 PBS wash), 5 min crosslink in media w/serum, 10 min crosslink in media w/serum, and 15 min crosslink in media w/serum, all at RT with 1% formaldehyde. The 10 min crosslink in media was the only condition that gave us good ChIP-qPCR (overall % input and difference between positive and negative promoters).

        I switched over to shearing in serum free media, and it seems to help. As far as I understand, serum helps permeabilize the membrane, so all I'm doing is reducing the crosslinking by getting rid of it.

        I heard great things about the Covaris; the only thing is that our lab has a Branson so it's very convenient and the qrt-PCR studies have already been worked out.

        One last question I have is, how important is it for the size distribution to be 200-700 bp for going onto the library prep? I've seen other people suggest that 100-600 bp is sufficient. I'm going to gel purify and cut out what corresponds to an insert size that is 150-200 bp anyways.

        Thanks again for the help!

        Comment

        • qsonica
          Member
          • Jun 2010
          • 11

          #5
          Probe sonication works well for most people but you must be sure the tip is submerged properly each time. If its too deep or too shallow it will affect the results.

          You also need to ensure you have the correct size tip for your specific sample volume. What is your sample volume? What size tip are you using?

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Cancer Drug Resistance: The Lingering Barrier to Rising Survival
            by SEQadmin2



            Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

            There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
            Today, 05:17 AM
          • GATTACAT
            Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by GATTACAT
            Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
            07-01-2026, 11:43 AM
          • SEQadmin2
            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by SEQadmin2


            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

            Here are nine questions we think about, in roughly the order they matter, before...
            06-18-2026, 07:11 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, Yesterday, 11:05 AM
          0 responses
          7 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-02-2026, 11:08 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-30-2026, 05:37 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-26-2026, 11:10 AM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Working...