  • kmcarr
    Senior Member
    • May 2008
    • 1181

    What fragment size do you use for your Illumina libraries?

    Illumina recommends insert sizes of ~200bp for SE libraries and ~300bp for PE libraries but clearly indicates in their literature that you can go larger (500-600bp). I'd like to ask the NextGen community a couple of questions:

    1. What insert size do you routinely use for your SE & PE libraries?

    2. Have you noticed any relationship between overall sequence quality and library fragment size? (Actual data correlating these would be terrific.)

    Thanks.
  • kainsteven
    Member
    • Jul 2009
    • 16

    #2
    For RNA-Seq applications, we have seen better results by excising material corresponding to the peak of the bioanalyzer trace for library preparation. In most cases, this corresponds to ~ 250 – 350 bp in length. We have used this range for both SR and PE libraries, but only with short read lengths of 32-36 bp. For longer read lengths and PE sequencing, I would imagine you would want to lean towards the longer end of this range.

    Best,
    Steve

    Comment

    • greigite
      Senior Member
      • Mar 2009
      • 145

      #3
      I don't have any data on this myself, but this paper reports the successful use of a 600 bp library, which reduced overall coverage variability: Harismendy & Frazer 2009

      Comment

      • maureen
        Junior Member
        • Feb 2010
        • 3

        #4
        Solexa recommends 200-250 bp and a tight range, but I definitely know of successful routine sequencing of libraries 300-700 bp. In fact my own library is 300-600 bp.

        Comment

        • der_eiskern
          Member
          • Jul 2009
          • 46

          #5
          illumina may recommend smaller library sizes but i'm pretty sure there are people using 1kb or 2kb insert sizes in parallel with the 200-700 bp sizes for structural variant discovery.

          Comment

          • edge
            Senior Member
            • Sep 2009
            • 199

            #6
            Hi kainsteven,

            I have an Illumina RNA-seq, pair-end read, 2x90bp....
            However, I not really sure about the fragment size/insert size
            It might affect my assembly result.

            Based on your experience, do you have any idea regarding the fragment size of Illumina pair-end RNA-seq data if my read length is 90nt.

            Many thanks for advice.

            Originally posted by kainsteven View Post
            For RNA-Seq applications, we have seen better results by excising material corresponding to the peak of the bioanalyzer trace for library preparation. In most cases, this corresponds to ~ 250 – 350 bp in length. We have used this range for both SR and PE libraries, but only with short read lengths of 32-36 bp. For longer read lengths and PE sequencing, I would imagine you would want to lean towards the longer end of this range.

            Best,
            Steve

            Comment

            • RPiddock
              Lab Monkey
              • Jun 2011
              • 16

              #7
              Our libraries team generally shear for about 500bp with Illumina TruSeq libraries.

              Comment

              • edge
                Senior Member
                • Sep 2009
                • 199

                #8
                Thanks for replying
                Is there got any other way to inferred insert size by undergo some statistic analysis?
                Thanks.

                Originally posted by RPiddock View Post
                Our libraries team generally shear for about 500bp with Illumina TruSeq libraries.

                Comment

                • tony19871929
                  Junior Member
                  • Feb 2012
                  • 1

                  #9
                  question about illumina seq.

                  hello every one , i really want to know the different Illumina standard (300bp insert) from overlapping 180bp.? can you help me ,thank you

                  Comment

                  • kainsteven
                    Member
                    • Jul 2009
                    • 16

                    #10
                    Thank you for the question. As mechanical or enzymatic fragmentation methods will produce a distribution of fragment sizes we typically choose a median size of 350 to 400 bp for 2 x 90 bp reads. There is no harm is going to this length, and it reduces the likelihood of any forward and reverse sequencing overlaps.

                    Steve

                    Comment

                    • apeterson
                      Junior Member
                      • Jan 2011
                      • 3

                      #11
                      Originally posted by der_eiskern View Post
                      illumina may recommend smaller library sizes but i'm pretty sure there are people using 1kb or 2kb insert sizes in parallel with the 200-700 bp sizes for structural variant discovery.
                      Does anyone have any more information on making 1kb insert size libraries?

                      Comment

                      • cement_head
                        Senior Member
                        • Mar 2012
                        • 262

                        #12
                        Originally posted by edge View Post
                        Hi kainsteven,

                        I have an Illumina RNA-seq, pair-end read, 2x90bp....
                        However, I not really sure about the fragment size/insert size
                        It might affect my assembly result.

                        Based on your experience, do you have any idea regarding the fragment size of Illumina pair-end RNA-seq data if my read length is 90nt.

                        Many thanks for advice.
                        It should not affect your assembly, IF you have enough (ideal) coverage. What will affect your assembly is if you have overlapp in your SBS. For example, if your fragment length was 300 bp and your paired-end reads were 200 bp you will have a problem because the reads actually have 100 bp overlap and the assembler will have difficulty discerning whether or not it's an overlap or two different reads.

                        What you need to calculate, or consider, is how much of the end of the fragment that you need to sequence in order to identify it as unique. In a haploid organism that answer is not very much. In a diploid organism that answer is more complicated. You have to consider variants (like alleles) and how you will be able to assign reads to a given gene. Then you might consider SBS to within a few bases of each other (e.g. paired-end sequencing of 140 bases of SBS on a 300 bp fragment) to cover most of the fragment.

                        How did you prepare your library? If you used Nextera, you have sizes that range from 300 to 1400 bp. If you used disruption, you should have a much narrower distribution of sizes.

                        Hope this helps,
                        Andor

                        Comment

                        • chiaraf
                          Member
                          • Oct 2012
                          • 10

                          #13
                          Hi!
                          I routinely sequence libreries with a mean fragment size of 450 bp for a 100PE for DNA and libreries witha mean fragment size of 350-400 for a 100PE for RNA seq.

                          Comment

                          • pc2009open
                            Junior Member
                            • Jun 2009
                            • 3

                            #14
                            Hi everyone,

                            I am also interested in 1Kb-1.5Kb library. Do you know any lab that has the expertise?

                            Thanks a lot.



                            Originally posted by apeterson View Post
                            Does anyone have any more information on making 1kb insert size libraries?

                            Comment

                            Latest Articles

                            Collapse

                            • seqadmin
                              Pathogen Surveillance with Advanced Genomic Tools
                              by seqadmin




                              The COVID-19 pandemic highlighted the need for proactive pathogen surveillance systems. As ongoing threats like avian influenza and newly emerging infections continue to pose risks, researchers are working to improve how quickly and accurately pathogens can be identified and tracked. In a recent SEQanswers webinar, two experts discussed how next-generation sequencing (NGS) and machine learning are shaping efforts to monitor viral variation and trace the origins of infectious...
                              Yesterday, 11:48 AM
                            • seqadmin
                              New Genomics Tools and Methods Shared at AGBT 2025
                              by seqadmin


                              This year’s Advances in Genome Biology and Technology (AGBT) General Meeting commemorated the 25th anniversary of the event at its original venue on Marco Island, Florida. While this year’s event didn’t include high-profile musical performances, the industry announcements and cutting-edge research still drew the attention of leading scientists.

                              The Headliner
                              The biggest announcement was Roche stepping back into the sequencing platform market. In the years since...
                              03-03-2025, 01:39 PM
                            • seqadmin
                              Investigating the Gut Microbiome Through Diet and Spatial Biology
                              by seqadmin




                              The human gut contains trillions of microorganisms that impact digestion, immune functions, and overall health1. Despite major breakthroughs, we’re only beginning to understand the full extent of the microbiome’s influence on health and disease. Advances in next-generation sequencing and spatial biology have opened new windows into this complex environment, yet many questions remain. This article highlights two recent studies exploring how diet influences microbial...
                              02-24-2025, 06:31 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by seqadmin, 03-20-2025, 05:03 AM
                            0 responses
                            34 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-19-2025, 07:27 AM
                            0 responses
                            42 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-18-2025, 12:50 PM
                            0 responses
                            35 views
                            0 reactions
                            Last Post seqadmin  
                            Started by seqadmin, 03-03-2025, 01:15 PM
                            0 responses
                            190 views
                            0 reactions
                            Last Post seqadmin  
                            Working...