  • d00b9p
    Junior Member
    • May 2013
    • 2

    Genomic DNA QC using Bioanalyzer?

    This might sound like a daft question, but if I was to run 1 ul of genomic DNA on a Bioanalyzer HS chip would I be able to detect sample degradation? I realise that high molecular weight genomic DNA will be far too large to be detected, but would any smaller weight 'smear' show up on the Bioanalyzer trace?

    We have some insect extractions with very little DNA (~100 ng) from a whole insect. We have found running out the DNA on a gel to be quite a poor way of seeing the DNA quality - needing at least 30 ng and using Ethidium staining. While we can sometimes see the faint high molecular weight band, it is near impossible to see if there is much 'smear'. And the main problem is that we have so little DNA in the first place.

    I wondered if using the Bioanalyzer may help, since it uses so little of the precious sample.
  • chevrm
    Member
    • Apr 2012
    • 14

    #2
    I'd consider a 12000 chip if you're willing to burn 1uL...it should be able to visualize anything over 0.5ng/uL

    Comment

    • Susanne
      Member
      • Aug 2009
      • 33

      #3
      The HS chip is quite sensitive with regards to genomic DNA and overloading. So I would recommend not to load more than 50 ng/µl, better around 10-15 ng/µl if you want to quantify. There are several problematic aspects to it though - if your sample runs into the upper marker, your quantification will be biased, and if you load too much, the quantification will also be wrong.
      Genomic DNA cannot enter the channels and it's possible that the chip doesn't run properly (if the gDNA is not degraded).
      The 0.5 ng/µl quoted for the 12000 chip is referring to this concentration in a single peak. You will not see this concentration in a smear sample (more like 5 ng/µl required).

      Comment

      Latest Articles

      Collapse

      • GATTACAT
        Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
        by GATTACAT
        Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
        07-01-2026, 11:43 AM
      • SEQadmin2
        Nine Things a Sample Prep Scientist Thinks About Before Sequencing
        by SEQadmin2


        I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

        Here are nine questions we think about, in roughly the order they matter, before...
        06-18-2026, 07:11 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 07-02-2026, 11:08 AM
      0 responses
      18 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-30-2026, 05:37 AM
      0 responses
      19 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-26-2026, 11:10 AM
      0 responses
      21 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 06-17-2026, 06:09 AM
      0 responses
      54 views
      0 reactions
      Last Post SEQadmin2  
      Working...