  • BA Traces what's going on?

    Quick summary,

    some library prepped using truseq RNA v2 kit slightly modified protocol (mainly volumes)

    Samples 1-8 are library I prepped last week (my first time), some samples at different RNA input conc. (RNA from human blood) PAX gene, and globin cleared (some samples Double globin cleared)
    samples 9-11 library prepped few weeks ago by experienced tech using same protocol.
    BA was done with a DNA high sensitivity chip

    can anyone tell me what they think is going on?

    Why are my peaks so big?
    Why do a few have a double main peak?
    What is the broad trailing peak on some of the samples?
    In samples 9-11 I assume early peak is dimers, that is not good right? Dimers will use up some of my reads right?
    Attached Files
    Last edited by subxero; 09-03-2013, 09:21 AM.

  • #2
    The "double peak" is likely too much DNA. It goes off scale and that is what you see.

    Yes-the first peak in the later traces are likely primer dimer-they will take up sequencing real estate.

    Comment


    • #3
      Those aren't pretty! The double peak may be offscale. The secondary peak to the right may be from overamplification (was there a PCR step after you made your library?). The peak @140 bp is adapter dimer. If you don't get rid of that you'll waste a lot of space on your flow cell as adapter dimer sequences preferentially.

      Comment


      • #4
        Yeah there was a PCR step to amplify the Library, I figured I had over amplification but still don't understand how my samples are over amplified yet the samples prepped a few weeks prior using all the same stuff does not?

        What is common cause of adapter dimers?

        Comment


        • #5
          Adapter dimers can result when your DNA:adapter ratio is off, among other things. If you have them and then do PCR, they amplify rapidly and happily. I generally use an 0.6x concentration of Ampure XP beads to remove them - it's not always perfect, but it usually gets rid of most of them.

          Comment


          • #6
            Ill definitely look into another bead clean up step.

            how much total library should I be aiming for in the end?

            Comment


            • #7
              What instrument are you planning on running your libraries on?

              Comment


              • #8
                cluster with cbot then run on HiSeq2000

                I guess the big question is, what do I have to do to get this library to a point where it can be sequenced efficiently?

                A) get ride of dimers
                B) decrease amplification (simply decrease PCR cycles?)

                But other than that it looks reasonable?
                Last edited by subxero; 09-05-2013, 06:52 AM.

                Comment


                • #9
                  Here is a trace from my latest library prep. I ran these libraries on a DNA 1000 chip instead of the HS chip as I figured had enough library to do so.

                  I still need to do the kappa qPCR QC to get my concentrations but to my untrained eye things look good? reasonable? somewhere in between maybe.

                  here are values to go with trace

                  Size [bp] Conc. [ng/µl] Molarity [nmol/l] Observations
                  15 4.2 424.2 Lower Marker
                  253 2.75 16.5
                  258 0.63 3.7
                  265 0.71 4
                  276 0.62 3.4
                  286 2.25 11.9
                  310 0.82 4
                  320 0.67 3.2
                  331 1.28 5.8
                  356 0.37 1.6
                  365 0.28 1.2
                  372 0.24 1
                  379 0.26 1.1
                  1,500 2.1 2.1 Upper Marker
                  Attached Files

                  Comment

                  Latest Articles

                  Collapse

                  • seqadmin
                    Essential Discoveries and Tools in Epitranscriptomics
                    by seqadmin




                    The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
                    04-22-2024, 07:01 AM
                  • seqadmin
                    Current Approaches to Protein Sequencing
                    by seqadmin


                    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                    04-04-2024, 04:25 PM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by seqadmin, 04-25-2024, 11:49 AM
                  0 responses
                  20 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-24-2024, 08:47 AM
                  0 responses
                  20 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-11-2024, 12:08 PM
                  0 responses
                  62 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 10:19 PM
                  0 responses
                  61 views
                  0 likes
                  Last Post seqadmin  
                  Working...
                  X