  • subxero
    Member
    • Jul 2013
    • 16

    #1

    BA Traces what's going on?

    Quick summary,

    some library prepped using truseq RNA v2 kit slightly modified protocol (mainly volumes)

    Samples 1-8 are library I prepped last week (my first time), some samples at different RNA input conc. (RNA from human blood) PAX gene, and globin cleared (some samples Double globin cleared)
    samples 9-11 library prepped few weeks ago by experienced tech using same protocol.
    BA was done with a DNA high sensitivity chip

    can anyone tell me what they think is going on?

    Why are my peaks so big?
    Why do a few have a double main peak?
    What is the broad trailing peak on some of the samples?
    In samples 9-11 I assume early peak is dimers, that is not good right? Dimers will use up some of my reads right?
    Attached Files
    Last edited by subxero; 09-03-2013, 09:21 AM.
  • kwaraska
    Senior Member
    • Nov 2008
    • 131

    #2
    The "double peak" is likely too much DNA. It goes off scale and that is what you see.

    Yes-the first peak in the later traces are likely primer dimer-they will take up sequencing real estate.

    Comment

    • microgirl123
      Senior Member
      • Jun 2012
      • 199

      #3
      Those aren't pretty! The double peak may be offscale. The secondary peak to the right may be from overamplification (was there a PCR step after you made your library?). The peak @140 bp is adapter dimer. If you don't get rid of that you'll waste a lot of space on your flow cell as adapter dimer sequences preferentially.

      Comment

      • subxero
        Member
        • Jul 2013
        • 16

        #4
        Yeah there was a PCR step to amplify the Library, I figured I had over amplification but still don't understand how my samples are over amplified yet the samples prepped a few weeks prior using all the same stuff does not?

        What is common cause of adapter dimers?

        Comment

        • microgirl123
          Senior Member
          • Jun 2012
          • 199

          #5
          Adapter dimers can result when your DNA:adapter ratio is off, among other things. If you have them and then do PCR, they amplify rapidly and happily. I generally use an 0.6x concentration of Ampure XP beads to remove them - it's not always perfect, but it usually gets rid of most of them.

          Comment

          • subxero
            Member
            • Jul 2013
            • 16

            #6
            Ill definitely look into another bead clean up step.

            how much total library should I be aiming for in the end?

            Comment

            • microgirl123
              Senior Member
              • Jun 2012
              • 199

              #7
              What instrument are you planning on running your libraries on?

              Comment

              • subxero
                Member
                • Jul 2013
                • 16

                #8
                cluster with cbot then run on HiSeq2000

                I guess the big question is, what do I have to do to get this library to a point where it can be sequenced efficiently?

                A) get ride of dimers
                B) decrease amplification (simply decrease PCR cycles?)

                But other than that it looks reasonable?
                Last edited by subxero; 09-05-2013, 06:52 AM.

                Comment

                • subxero
                  Member
                  • Jul 2013
                  • 16

                  #9
                  Here is a trace from my latest library prep. I ran these libraries on a DNA 1000 chip instead of the HS chip as I figured had enough library to do so.

                  I still need to do the kappa qPCR QC to get my concentrations but to my untrained eye things look good? reasonable? somewhere in between maybe.

                  here are values to go with trace

                  Size [bp] Conc. [ng/µl] Molarity [nmol/l] Observations
                  15 4.2 424.2 Lower Marker
                  253 2.75 16.5
                  258 0.63 3.7
                  265 0.71 4
                  276 0.62 3.4
                  286 2.25 11.9
                  310 0.82 4
                  320 0.67 3.2
                  331 1.28 5.8
                  356 0.37 1.6
                  365 0.28 1.2
                  372 0.24 1
                  379 0.26 1.1
                  1,500 2.1 2.1 Upper Marker
                  Attached Files

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM
                  • SEQadmin2
                    New Genomics Technologies Take Aim at Long-Standing Limits
                    by SEQadmin2


                    Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                    We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                    ...
                    09-28-2026, 10:25 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, Yesterday, 10:52 AM
                  0 responses
                  12 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-14-2026, 05:39 AM
                  0 responses
                  19 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-13-2026, 09:56 AM
                  0 responses
                  19 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-09-2026, 11:14 AM
                  0 responses
                  27 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...