Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • paolo.kunder
    Member
    • Aug 2011
    • 93

    #1

    Adaptor dimer contamination

    Dear All,

    We are facing some problems with adaptor dimers in Illumina RNA seq library prep.

    There is a small amount of adaptor dimer (at ~150 bp) in a few of the samples. We tried a second bead clean up on half of the sample, but lost a significant amount of material. We talked with Illumina tech support, and this is a tiny enough amount of adaptor that they felt if was fine to go forward. There will be some adaptor only reads that the analyst can filter from the data, but it should be a small percentage.

    I attach the profile of the sample most contaminated, do you think this wont influence the amplicfication of my library??

    Many thanks,
    Paolo
    Attached Files
  • Genohub
    Registered Vendor
    • Mar 2013
    • 210

    #2
    Ampure Size Selection to remove adapter dimer

    Hi Paolo,

    What bead to sample volume ratio did you use for your cleanup? If you use the wrong ratio, loss of library is possible. See slide 52: http://www.broadinstitute.org/files/...PrepSlides.pdf

    A ratio of 0.7-0.8x is pretty effective at removing the adapter dimer band without resulting in sample loss.

    I would personally do one more cleanup to completely remove the dimer band. Try it out on a non-precious sample first to test.

    - Genohub

    Comment

    • microgirl123
      Senior Member
      • Jun 2012
      • 199

      #3
      That's a lot of adapter dimer relative to your sample! Adapter dimer tends to hybridize to the flow cell much better than your actual library. I'd be worried about sequencing mainly adapter dimer if I loaded that onto our MiSeq.

      Comment

      • KJoseph
        Junior Member
        • Dec 2013
        • 3

        #4
        In my experience, that amount of adapter dimer would cause some significant problems on a MiSeq, and you would definitely lose quite a number of reads on HiSeq as well.

        Comment

        • hewei
          Junior Member
          • Sep 2013
          • 4

          #5
          dilute your adaptors, up to 10 folds dilution, you can try different ratio to get the best result.

          Comment

          • liuqy
            Junior Member
            • Sep 2018
            • 3

            #6
            Yes, this will definitely fail quality control. Nowadays the beads have improved a lot. You can use AmPure XP or NVIGEN MagVigen. But MagVigen fares better. https://www.nvigen.com/dna-clean-up/

            Disclosure: I work at NVIGEN. In case you find problems using MagVigen. We will be happily work with you to troubleshoot or fine tune it for your needs.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
              by SEQadmin2



              CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

              Despite this, “CRISPR helped turn genome editing from a specialized technique into
              ...
              07-31-2026, 11:01 AM
            • SEQadmin2
              Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
              by SEQadmin2


              Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

              The systematic characterization of the human proteome has
              ...
              07-20-2026, 11:48 AM
            • SEQadmin2
              Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
              by SEQadmin2



              Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
              ...
              07-09-2026, 11:10 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 07:41 AM
            0 responses
            12 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 08-03-2026, 10:13 AM
            0 responses
            27 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-31-2026, 02:55 AM
            0 responses
            39 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-24-2026, 12:17 PM
            0 responses
            25 views
            0 reactions
            Last Post SEQadmin2  
            Working...