    Innaccurate Kapa Quant Standards

    On my campus two labs, both with very extensive experience in NGS, have reported that recently purchased Kapa qPCR standards have under-estimated library concentrations. Both sets of libraries were quantified following two similar but lab dependent protocols (one uses the full kit the other uses just the standard and an off the shelf taq-mix) and then diluted to 2 nM and the dilutions were then quantified. Typical amounts of library (10 and 12 pmol/lane; the latter exceeds Illumina's rec. but the group swears by it) were run on HiSeq 8-lane flowcells but produced ~50% of the expected reads (~4x10^8/lane). Typically these labs see 6 - 8 x10^8 reads/lane. One lane contained a sample which was being re-run and was quantified months before with a different lot of Kapa. This recent run of this sample produced the expected number of reads. There was no issue with the quality scores any samples and the data are fine (just lower coverage). The library quantifications were double checked using the suspect Kapa and it doesn't appear to be anything to do with the PCRs.

    The dates of purchase and shipping/receiving of the suspect Kapa kits are very similar (Nov 2013) but the lot numbers are different. Has anyone had any trouble with Kapa standards recently?

    UPDATE: We spoke to Kapa and they are having problems with lots. I would call ahead before running anything.
    Last edited by Daytwa; 02-07-2014, 12:07 PM. Reason: New News
  • AdrianaGeldart
    Junior Member
    • Feb 2014
    • 6

    #2
    I am a Kapa Biosystems Application Scientist and just wanted to clarify the above post. When a customer calls into Kapa Biosystems Support with a complaint regarding a product's performance, we ask for supporting data and the product lot number(s). This is standard procedure for all complaint inquiries, and does not imply that we have an issue with any lots. I would like to reassure everyone that there is no systemic issues with our KAPA Library Quantification DNA Standards. If a customer is concerned about their product, please feel free to contact Kapa Biosystems directly at [email protected] or 855-527-2246. Thank you.

    Comment

    Latest Articles

    Collapse

    • SEQadmin2
      Nine Things a Sample Prep Scientist Thinks About Before Sequencing
      by SEQadmin2


      I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.


      Here are nine questions we think about, in roughly the order they matter, before...
      06-18-2026, 07:11 AM
    • SEQadmin2
      From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
      by SEQadmin2


      Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


      The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
      ...
      06-02-2026, 10:05 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, 06-17-2026, 06:09 AM
    0 responses
    25 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-09-2026, 11:58 AM
    0 responses
    42 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-05-2026, 10:09 AM
    0 responses
    48 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-04-2026, 08:59 AM
    0 responses
    49 views
    0 reactions
    Last Post SEQadmin2  
    Working...