Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • moerae
    Junior Member
    • Feb 2014
    • 9

    #1

    Covaris M220 and AmPure beads

    Hi,

    I'm to the NGS world and I just started my experiments the other day. I'm using a custom designed SureSelect XT capture kit and sheared my DNA using a Covaris M220. The protocol uses an S220/E220 but I don't have access to those machine and thought that the M220 would be all right. The settings that was in the SureSelect XT doesn't apply as one of the settings can't go that high on the M220. So I just sheared my DNA using the M220 protocol for fragmenting DNA to 150 bp as the SureSelect protocol wants fragments that are between 150 - 200 bp. I analysed the fragments on a Bioanalyzer high sensitivity DNA chip and found that the fragments are around 200 bp in size (the peak is between 220 - 300 bp on the Bioanalyzer). How important is the sheared DNA size and would this affect my downstream experiments?

    Also, for the first AmPure bead extraction, part of the protocol suggest drying the pellet on a 37C heat block for 5 mins and not to let the pellet dry excessively as it affects elution. Someone changed my setting to 99C and I didn't realise and I over-dried my pellets. My elution is quite low for my samples... how much of an affect will this have on my downstream experiments?

    Thank you
  • moerae
    Junior Member
    • Feb 2014
    • 9

    #2
    Can someone please please help? I'm really really stuck. If I carry on how likely am I to screw up the whole experiment?

    Comment

    • Skamp
      Junior Member
      • May 2012
      • 4

      #3
      I have done a sequence capture experiment using another kit and I think the SureSelect works in a similar way. You create a whole genome shotgun library, incubate it with your probes, pull out the library fragments that bind to the probes and do a final clean up.

      1. Does the size rang of the sheared DNA matter?

      Probably, yes. If the SureSelect protocol uses AMPure XP beads to do the size selection, then yes the fragment size does matter because Agilent has adjusted the ratios of beads to keep a narrow size range during the library preparation step.

      2. Is high incubation of AMPure beads bad?

      Yes, if you over-dry your AMPure XP beads it is hard to elute the DNA off of them. This will result in less DNA for whatever you plan to do with it. Depending where you are in the protocol, this may have a big effect. Why don't you just try continuing with one or two samples and see how it goes?

      Comment

      • moerae
        Junior Member
        • Feb 2014
        • 9

        #4
        Thanks for the reply. I'm about 50 bp out from what the protocol requires, and I've tried to ask Agilent but no one is replying to my email. I will probably just go on with the experiment and see what happens.

        Comment

        • mzahir89
          Junior Member
          • Aug 2015
          • 6

          #5
          Hi moerae,

          I have a similar question to yours. Can you kindly update what happened when you went on with the experiment? Did the excess 50-100bp have any effect?

          Comment

          • luc
            Senior Member
            • Dec 2010
            • 469

            #6
            I assume longer inserts would havean increased tendency towards cross hybridizations; but I have no data to back this up.

            Comment

            • jlei_face
              Member
              • Oct 2014
              • 10

              #7
              I know this reply is super late but since it seems there are other posters with similar issues I'll put in my two cents:
              When doing a library prep for first few times, it is recommended that you adhere to the supplied protocol as much as possible. Deviating even slightly (especially at the beginning) can have a huge impact on the end result and you won't know really how to troubleshoot.
              Moerae: before beginning the actual prep, you should have done a few experiments with the covaris to obtain the proper fragment size. By slightly adjusting treatment time and duty factor you may have been able to find the correct settings for a 150-200bp fragment size.
              As far as over-drying the beads, yes, this is a huge impact on your yield (as you saw). Decreased yield will also lead to lower sequencing diversity, maybe higher duplicates, etc. Given you knew what went wrong and how expensive sequencing is, I would not continue with this prep.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2




                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                10-30-2026, 06:46 AM
              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-28-2026, 09:57 AM
              0 responses
              45 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-28-2026, 09:51 AM
              0 responses
              28 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              76 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              49 views
              0 reactions
              Last Post SEQadmin2  
              Working...