  • zwang0614
    Member
    • Feb 2014
    • 10

    Problems with ChIP-seq library construction

    Hi all.
    I am trying to prepare several ChIP-seq libraries. I started with 3ng DNA, which was sonicated to 100-500bp length. After the blunt ending, A tail adding, and adapter ligation, I did the qPCR to do the quantitation but found most of the qPCR products were adaptor dimers (about 120bp). Does anybody have similar experiences or any suggestions? Thanks.

    Z.
  • chen
    Member
    • Sep 2009
    • 14

    #2
    Hi zwang,

    we had a similar result starting with ChIP-DNA, using the Truseq nano pre kit. And we find some smear above the dimers when we redo the PCR step, adding the cycles up to 30 comparing with protocol's 8 cycle. we gel purify the PCR product.

    but we are not quite sure about the smear.
    May someone give any advice.
    Cheers,
    Tanya

    Comment

    • NextGenSeq
      Senior Member
      • Apr 2009
      • 482

      #3
      30 cycles will lead to a lot of duplicate reads. We typically do 13 to 17 depending on the starting amount of ChIP-DNA.

      Reduce the amount of adapters in the ligation step when you start with small amounts of ChIP-DNA. This helps reduce adapter artifacts.

      Comment

      • chen
        Member
        • Sep 2009
        • 14

        #4
        Originally posted by NextGenSeq View Post
        30 cycles will lead to a lot of duplicate reads. We typically do 13 to 17 depending on the starting amount of ChIP-DNA.

        Reduce the amount of adapters in the ligation step when you start with small amounts of ChIP-DNA. This helps reduce adapter artifacts.
        yeah. yet this was our first try ChIP-seq library, so we just stick to the protocol(Truseq Nano DNA kit). we may improve the ligation step next time.

        we would take the smear library to do a test run. Hope the data would be not so bad.
        Cheers,
        Tanya

        Comment

        • NextGenSeq
          Senior Member
          • Apr 2009
          • 482

          #5
          I personally do not like the Illumina Nano kit. We get better results with the NuGen Ultralow or Diagenode ChIP kit.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
            by SEQadmin2


            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

            Here are nine questions we think about, in roughly the order they matter, before...
            06-18-2026, 07:11 AM
          • SEQadmin2
            From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
            by SEQadmin2


            Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


            The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
            ...
            06-02-2026, 10:05 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 06-17-2026, 06:09 AM
          0 responses
          41 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-09-2026, 11:58 AM
          0 responses
          102 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-05-2026, 10:09 AM
          0 responses
          123 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 06-04-2026, 08:59 AM
          0 responses
          114 views
          0 reactions
          Last Post SEQadmin2  
          Working...