    Bad bioanalyzer results from ChIP DNA

    We've done several ChIP experiments that we want to sequence. We sent our samples to the sequencing facility and were told that they had bad Bioanalyzer results. We then used our own Bioanalyzer and retested some of the leftover samples and also prepared new samples and tested them. After ChIP the DNA is purified using a Qiagen column and eluted in the Qiagen elution buffer (10 mM Tris·Cl, pH 8.5). The below analysis is on the purified ChIP DNA not prepared libraries.

    Here are some examples: http://i.imgur.com/rDXjTJi.png

    1. Old input used as a control and I think it's the only one that looks good

    2+3. New inputs, they have 2 peaks and one has a dip at ~300bp.

    4+5. IP samples, both have low concentration and one has a dip at ~300bp.

    Here are some other samples: http://i.imgur.com/BbC8Fkr.png

    These are two samples tested at two different facilities at two different times. They all seem to show low concentration but only one facility shows the dip at ~300bp. I've asked the people who run both Bioanalyzers and neither know what this dip is, though one said that he had seen it a few times before with other peoples' samples.

    All samples tested by realtime PCR at sites of interest amplify as expected. We have sequenced other ChIP samples in the past but at different facilities and were never told of any problems with our DNA, but I don't know if that's because we really had no problems or because they didn't run the samples on a Bioanalyzer. Is it standard practice to test the ChIP'd DNA on a Bioanalyzer before library prep or anything else is done with it? The sequencing results we got in the past were generally ok though a few times we had high duplication rates.

    So what could be going on here? Why do some of the inputs have two peaks, do some of the samples just have really low concentrations, what is the dip at 300bp and why would it only show up when a sample is tested at one place but not another? Is it worth it to try make libraries and sequence these?
    Last edited by biocomputer; 04-03-2014, 10:13 AM.

Latest Articles

Collapse

  • GATTACAT
    Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by GATTACAT
    Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
    07-01-2026, 11:43 AM
  • SEQadmin2
    Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by SEQadmin2


    I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

    Here are nine questions we think about, in roughly the order they matter, before...
    06-18-2026, 07:11 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 07-02-2026, 11:08 AM
0 responses
21 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-30-2026, 05:37 AM
0 responses
22 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-26-2026, 11:10 AM
0 responses
21 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-17-2026, 06:09 AM
0 responses
54 views
0 reactions
Last Post SEQadmin2  
Working...