  • LTJensen
    Junior Member
    • Jan 2014
    • 5

    #1

    Sciclone automation and Nextera Rapid Exomes

    We are in the process of implementing a Sciclone robot to prep Nextera Rapid exomes. However, test samples fail the tagmentation step. Using the robot work flow we get tagment sizes well below 200bp (test1 attached).
    The robot workflow differs from the Illumina protocol in the mixing procedure (robot mixes by pipetting, while Illumina protocol recommends shaking for 1min @ 1800rpm) and 58oC incubation (robot incubates in PCR plate, while manual suggests Midi plate).
    We have tried to let the robot add buffer and enzyme, then perform off-robot shaking at 1800rpm in PCR plate, then returning plate to robot for incubation and adding of stop buffer (mixing by pipetting). This increased the tagment size to approximately 220bp (test 5 attached). However, after PCR, the profiles still look bad (test 5 NLS). Same approach with 1800rpm mixing and off robot incubation using Midi plates instead gave the same result.
    Does anyone have any experience using the Sciclone for the Nextera Rapid protocol and/or have any ideas on how to resolve these problems?
    Thanks!
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    From detailed information supplied here, robot does not seem to be the problem. The reason that Illumina recommends MIDI plate not PCR plate is that tagmenation reaction and later bead used for cleaning up the reaction cannot be mixed by shaking if they are in a PCR plate. MIDI plate wells have higher volume and are wider making solution in them suitable for mixing by a shaker. MIDI plate does not fit in PCR machines and they recommend Microheating system because it has insert for MIDI plate to fit. So MIDI plates are recommended for high throughput processing of samples.

    Tagmentation reactions traces indicate that problem could be with DNA quality. If you are following Illumina recommended QC and input DNA amount, I would suspect that only some portions of input DNA are accessible for transposon so they get over-tagmented resulting in smaller fragments and some parts are not so accessible making that large DNA fragment trail stretching beyond upper marker at 10 kb. I would clean input DNA with AMPure beads or Zymo gDNA column and elute it with Tris-Cl pH8-8.5 or TE0.1 followed by QC and tagmentation and see if that makes difference. Close examination of DNA extraction method might be informative as well. We follow Illumina protocol in PCR plate because we do not process many samples at a time and the results are as illustrated in Illumina user guide.

    Comment

    • LTJensen
      Junior Member
      • Jan 2014
      • 5

      #3
      Thank you for your reply, nuracidhunter.

      I am pretty sure that DNA quality is not an issue here, as the used sample are high quality Promega reference DNA, purified by UF filtration. The same Promega sample produces fine tagmentation and post PCR Agilent profiles when prepared manually...

      Comment

      • cfrazar
        Junior Member
        • Oct 2012
        • 1

        #4
        I'm currently running the same protocol on the Sciclone with 96 samples. I took a look at your BA traces and they look just like ours. I ran ~30 samples and they all had the same profile. When overlaid the leading edge of the peak was consistent among samples. It is likely due to ampure cleanup removing smaller fragments.

        Small insert sizes tend to be a problem with Nextera. Initial DNA quantification is pretty key and plays a role in the resulting fragment size. You are using Promega DNA, so then quality should be pretty good. I would continue your test and go through the captures. If you need to, you can play with the final SPRI cleanup to push the average insert size up.

        Good luck.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          For anyone interested in a similar protocol using Tecan instead: http://res.illumina.com/documents/pr...automation.pdf

          Comment

          • LTJensen
            Junior Member
            • Jan 2014
            • 5

            #6
            Thank you very much for the inputs.
            cfrazar, how did the samples perform when sequenced?

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            37 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-15-2026, 10:52 AM
            0 responses
            46 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            46 views
            0 reactions
            Last Post SEQadmin2  
            Working...