    Inconsitencies with qPCR results

    I've begun quantifying all libraries with qPCR (Kapa kit) but I'm getting massive inconsistencies with the number of reads I get from samples in a multiplex pool.
    I dilute each final library 1:1000000 (three 1:100 serial dilution) mixing at least 10 times by pipette after each dilution. This then puts me somewhere in the middle of the Kapa standard curve. Everything is run in triplicate and the R2 of the standard curve is always great >0.99.
    I then normalise, pool and requantify by both Qubit & qPCR and generally get a pooled concentration of 9-11nM (what I might expect). Cluster density for the overall run is fairly good, but after indexing there's massive variation sample-to-sample (5-fold differences for my last set of TruSeq RNA libraries).
    I've also Qubit'd and Bioanalysed each library (I use the average size between 200-700bp to normalise my Kapa qPCR). There are no dimers or evidence of over-amplification. The Qubit concetrations are slightly better than qPCR but still not great (R2 between qubit and cluster# is 0.52, but for qPCR it was 0.41)
    Does anyone have a fool-proof method for qPCR? I was wondering if it's worth running mulitple dilutions of the same sample on the PCR plate, but I would quickly run out of space on my 96 well plate (I often have to do 24 libraries at a time).
    Has anyone tried a homebrew probe assay?
    Attached Files
  • microgirl123
    Senior Member
    • Jun 2012
    • 199

    #2
    Strange, I pretty much use your approach for our gDNA libraries and don't see such huge inconsistencies in number of reads.

    Are your samples PCR-enriched after library prep? I usually make PCR-free libraries and have found that the structure of the adapters can make the samples run strangely in the Bioanalyzer, resulting in incorrect sizing which affects the KAPA math.

    Are all your pooled libraries approximately the same size?

    Pipetting error?

    The other thing that I've been doing (I'm running a MiSeq) is denaturing the samples with NaOH and then denaturing them again at 96 degrees. Somewhere there's an Illumina bulletin that explains how different adapters/indices denature less effectively than others.

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      We use concentration from TapeStation or bioanalyser and normalise to 6 nM. Then dilute it 10000 times in two steps for QPCR using KAPA kit. Our pooled samples gives very close figures specially when they are the same type of library with similar trace.

      Comment

      • TonyBrooks
        Senior Member
        • Jun 2009
        • 303

        #4
        Originally posted by nucacidhunter View Post
        We use concentration from TapeStation or bioanalyser and normalise to 6 nM. Then dilute it 10000 times in two steps for QPCR using KAPA kit. Our pooled samples gives very close figures specially when they are the same type of library with similar trace.
        I was thinking about trying a double normalisation too. First using the Bioanalyser and next by qPCR. Might try that for the next run.
        We'll also work on automating qPCR set up to avoid any pipettting error.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM
        • SEQadmin2
          New Genomics Technologies Take Aim at Long-Standing Limits
          by SEQadmin2


          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
          ...
          09-28-2026, 10:25 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        22 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-15-2026, 10:52 AM
        0 responses
        39 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-14-2026, 05:39 AM
        0 responses
        45 views
        0 reactions
        Last Post SEQadmin2  
        Working...