    Problem with AMPure purification

    Hi,
    Maybe someone has faced the same problem, as I am:

    I am using Agencourt AMPure XP to clean my PCR-amplicons from PCR-reaction mixture. I am using AMPure at 1x sample volume, to get rid of all the primers.
    In the last few months I started noticing, that the AMPure magnetic beads do not pellet properly on the tube wall after the addition of Elution Buffer (10 mM Tris-HCl, pH 8,5). The DNA concentrations of amplicons also decreased. I tried to elute my amplicons in MQ water and it worked well, I also obtained a bit higher DNA concentrations. But I am still wondering, what is wrong with the elution using EB. I checked my Ethanol and tried to use 80% instead of 70%, it did not change anything. I also tried to use different lab plastic, magnet plate and ordered fresh EB and AMPure. Nothing worked and I still got this „haze“ instead of properly pelleted beads.

    Maybe someone could help me, what am I doing wrong?
  • rnaeye
    Member
    • May 2011
    • 80

    #2
    Are you using the same PCR polymerase/buffer. I know that having proteins like BSA in the sample causes haze during Ampure cleanup. I cannot explain why your problem disappears with MQ water though.

    Comment

    • JanaSEQ
      Junior Member
      • Oct 2013
      • 8

      #3
      I am using Maxima HotStart PCR Master Mix. I tried to leave out ethanol purification steps (just to see, what happenes) and it worked - the haze was gone. But I can´t do that with my real samples..

      Comment

      • bilyl
        Member
        • Aug 2013
        • 52

        #4
        Originally posted by JanaSEQ View Post
        Hi,
        Maybe someone has faced the same problem, as I am:

        I am using Agencourt AMPure XP to clean my PCR-amplicons from PCR-reaction mixture. I am using AMPure at 1x sample volume, to get rid of all the primers.
        In the last few months I started noticing, that the AMPure magnetic beads do not pellet properly on the tube wall after the addition of Elution Buffer (10 mM Tris-HCl, pH 8,5). The DNA concentrations of amplicons also decreased. I tried to elute my amplicons in MQ water and it worked well, I also obtained a bit higher DNA concentrations. But I am still wondering, what is wrong with the elution using EB. I checked my Ethanol and tried to use 80% instead of 70%, it did not change anything. I also tried to use different lab plastic, magnet plate and ordered fresh EB and AMPure. Nothing worked and I still got this „haze“ instead of properly pelleted beads.

        Maybe someone could help me, what am I doing wrong?
        The "haze" sometimes occurs if you incubate the bead mixture for too long. I've seen it a lot but only when I do "in-bead" enzymatic reactions and cleanup afterward with SPRI buffer (PEG/NaCl). I've also seen it if I overdry my beads. Maybe something to think about.

        Comment

        • JanaSEQ
          Junior Member
          • Oct 2013
          • 8

          #5
          Thanks bilyl!
          I have tried different incubation times. I have also tried to elute amplicons right after I removed ethanol and also tried to let the beads dry completely, before I added the elution buffer. I did not notice any changes.

          Comment

          • silviap
            Junior Member
            • Aug 2014
            • 3

            #6
            I' m using AMPure on a Beckman Platform for many years and I used water MQ as elution buffer without any problem. Are you using the original magnet plate?

            Comment

            • JanaSEQ
              Junior Member
              • Oct 2013
              • 8

              #7
              I am using DynaMag magnet plate (Life Technologies). I am not using MQ water, because sequencing company where we send our samples prefers PCR amplicons in Tris-HCl buffer.

              Comment

              • dtm2451
                Junior Member
                • Mar 2013
                • 9

                #8
                I'm not certain what might be going on here, but if MQ seems to work for the elution, could you just elute in water and then dilute in elution buffer?

                Comment

                • JanaSEQ
                  Junior Member
                  • Oct 2013
                  • 8

                  #9
                  Thank you Dan (dtm2451)!

                  I will try this out.

                  But I still would like to know, what could have happned with the purification. My last idea is, that maybe it´s the hot weather, that causes such abnormalities (air conditioner in my lab does not work properly and it is up to 30C/86F) But i refrigirate Ampure (and then bring it to room temperature immediately before using) I also tried to refrigerate ethanol and elution buffer, unfortunately it did not help.

                  Comment

                  • JanaSEQ
                    Junior Member
                    • Oct 2013
                    • 8

                    #10
                    My problem seems to be gone now! During last few Ampure purifications I did not notice haze any more. I did not change anything and so I think the haze was a result of way too hot working conditions. Fortunately is summer over now!

                    Comment

                    • dtm2451
                      Junior Member
                      • Mar 2013
                      • 9

                      #11
                      My lab also gets fairly hot in the Summer due to inefficient AC. Perhaps that was the issue as I'm no longer having SPRI trouble either. I don't like the heat explanation... I really don't get why that would inhibit bead binding, but it might be true. Either way, I'm glad you are no longer having issues either!

                      Comment

                      • docphil
                        Junior Member
                        • Aug 2014
                        • 8

                        #12
                        My Lab uses AMPure beads at 0.7x to remove primer-dimer and any junk below our target after barcoding and final pooling for MiSEQ. We elute into TE and then run the eluted DNA through a QiaQuick spin column to remove any residual beads. You can also adjust the amount of EB you spin through the column if you need a higher DNA concentration. Might be a bit redundant but hopefully it helps!

                        Comment

                        Latest Articles

                        Collapse

                        • SEQadmin2
                          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                          by SEQadmin2



                          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                          ...
                          10-14-2026, 09:55 AM
                        • SEQadmin2
                          New Genomics Technologies Take Aim at Long-Standing Limits
                          by SEQadmin2


                          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                          ...
                          09-28-2026, 10:25 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, Yesterday, 11:13 AM
                        0 responses
                        16 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, Yesterday, 10:48 AM
                        0 responses
                        17 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 10-15-2026, 10:52 AM
                        0 responses
                        35 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 10-14-2026, 05:39 AM
                        0 responses
                        42 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...