    #16
    Are you trying to go lower then 100 pM?

    Comment

    • rpm27
      Junior Member
      • Sep 2014
      • 3

      #17
      Hi
      The concentration is 150pM.
      Thanks.

      Comment

      • acockburn
        Member
        • Aug 2013
        • 10

        #18
        Assuming you want to load 10 pM, 1% spike in of PhiX and using the above subnanomolar protocol.

        Add the following --- Volume (uL)
        0.15 nM sample DNA --- 52
        0.2 N NaOH --- 52

        Denature 5 min.

        Add 20 uL of Tris-HCl pH 7.3 to Hyb buffer.

        Add the following --- Volume (uL)
        Denatured DNA --- 104
        Neutralization buffer --- 208

        Add a volume of Tris-Hyb buffer to the desired final loading concentration which is assumed to be 10 pM, so factoring in the PhiX.

        Add the following: --- Volume (uL)
        Denatured/diluted library --- 311.11
        Chilled Hyb buffer --- 458.89

        Check pH with indicator paper. It should be between 7 - 8.5.

        FINAL LOAD LIBRARY
        Add the following: --- Volume (uL) Final Concentrations
        10.1010101 pM Denatured Lib --- 693 -- 10 pM
        10 pM PhiX Library --- 7 -- 0.1 pM


        Load 600 uL into cartridge.
        Last edited by acockburn; 02-24-2015, 08:27 AM.

        Comment

        • rpm27
          Junior Member
          • Sep 2014
          • 3

          #19
          Thank you. What I ended up doing is to combine my low concentration libraries with phix to generate a 2nm library. I do not need lots of reads and merely for verifying sequence. It still failed because of over clustering. I feel as my SybrG assay was not that accurate. I am going to run digital pcr to get more accuracy.

          Comment

          • sriman
            Junior Member
            • Apr 2017
            • 2

            #20
            Neutralization Buffer

            Hello,

            I came across acockburn's suggestion on using the protocol published in the Supplementary Protocol 12: Modified hybridization buffers of the following paper: A large genome center's improvements to the Illumina sequencing system for sequencing low concentrated libraries.

            My question is on preparing the neutralization buffer, it is not clear to me whether I should add the 1 M Tris-HCl pH7.3 in the neutralization buffer or is this only added to the HT1?

            Please let me know if anyone has prepared this buffer before.

            Thanks a lot!!

            Originally posted by acockburn View Post
            Assuming you want to load 10 pM, 1% spike in of PhiX and using the above subnanomolar protocol.

            Add the following --- Volume (uL)
            0.15 nM sample DNA --- 52
            0.2 N NaOH --- 52

            Denature 5 min.

            Add 20 uL of Tris-HCl pH 7.3 to Hyb buffer.

            Add the following --- Volume (uL)
            Denatured DNA --- 104
            Neutralization buffer --- 208

            Add a volume of Tris-Hyb buffer to the desired final loading concentration which is assumed to be 10 pM, so factoring in the PhiX.

            Add the following: --- Volume (uL)
            Denatured/diluted library --- 311.11
            Chilled Hyb buffer --- 458.89

            Check pH with indicator paper. It should be between 7 - 8.5.

            FINAL LOAD LIBRARY
            Add the following: --- Volume (uL) Final Concentrations
            10.1010101 pM Denatured Lib --- 693 -- 10 pM
            10 pM PhiX Library --- 7 -- 0.1 pM


            Load 600 uL into cartridge.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM
            • SEQadmin2
              New Genomics Technologies Take Aim at Long-Standing Limits
              by SEQadmin2


              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
              ...
              09-28-2026, 10:25 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Yesterday, 10:52 AM
            0 responses
            11 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-14-2026, 05:39 AM
            0 responses
            19 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-13-2026, 09:56 AM
            0 responses
            18 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-09-2026, 11:14 AM
            0 responses
            27 views
            0 reactions
            Last Post SEQadmin2  
            Working...