Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • mcnaircm
    Junior Member
    • May 2017
    • 2

    #61
    Hi all,
    I'm just piggybacking on this thread, but I think it is probably the best place to post my question - and something I can't be the only person having an issue with!
    I'm prepping my samples for ATAC-Seq (planning to do single end sequencing 75bp reads, just looking at regions of open chromatin), and I know that most of the protocols have custom primers (both Ad1 and index primers). I was wondering if you HAD to use these, or if there were capable primers in the Nextera library prep kit and index kit to prep the samples. Again, I'm not sure if people use the custom primers because it is cheaper or if it is because Illumina is so vague with their primer mixes and cocktails. Any input would be appreciated!

    Cheers

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #62
      Not sure what protocol you are using, but the only one I have seen seemed to use somewhat modified Nextera adapters. This means that to sequence these libraries on a HiSeq 2500 you need to either do a Rapid run (which includes nextera sequencing primers) or add the appropriate Nextera primers to your high output reagents. We would be loathe to do the latter unless the investigator was buying all 8 lanes of the flow cell.

      --
      Phillip

      Comment

      • mcnaircm
        Junior Member
        • May 2017
        • 2

        #63
        Hi Phillip,
        So you are correct in that the majority of the protocols that I have seen use modified Nextera adapters. My question was whether or not these modified adapters had to be used (or were used mainly because they were cost effective), or if there were already reagents in the library and index kits for the nextera prep that would work.

        Additionally, adding the primers to the high output reagents wouldn't be problem, since we have indeed purchased the entire flow cell.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #64
          ATAC-seq primers differ from Nextera by:
          1- Lack of i5 index in Ad1 (equivalent to Nextera index2 read primer) so libraries are single indexed
          2- Addition of 6-7 bases to 3’ end of primers that extends primer binding into transposome recognition sequence common between both adapters

          I do not see any difference beside possible different concentrations of adapters used in PCR noting that Nextera adds PPM to supplement indexed primers as their concentration is lower than needed for Nextera. If you are using Nextera primers you might need to adjust annealing temperature and primer volume to optimise reaction.

          Comment

          • jamesdinn
            Junior Member
            • Dec 2016
            • 2

            #65
            Hi all,
            I would like to ask you about atac-seq sequencing,
            I sequence a library for 100M reads, and after filtering i get much less reads (TotalReadPairs:9 thousands !!) and i don't know whay this huge difference after filtering, Filtering file:
            xxx.trim.filt.srt.nodup.pbc.qc

            best

            Comment

            • finixtree
              Junior Member
              • Aug 2015
              • 3

              #66
              What are those reads filtered out for? low quality score, unmappable, mapped to mtDNA? Need more information to give you even a guess.
              Originally posted by jamesdinn View Post
              Hi all,
              I would like to ask you about atac-seq sequencing,
              I sequence a library for 100M reads, and after filtering i get much less reads (TotalReadPairs:9 thousands !!) and i don't know whay this huge difference after filtering, Filtering file:
              xxx.trim.filt.srt.nodup.pbc.qc

              best

              Comment

              • bman
                Junior Member
                • Sep 2014
                • 7

                #67
                Originally posted by nucacidhunter View Post
                ATAC-seq primers differ from Nextera by:
                1- Lack of i5 index in Ad1 (equivalent to Nextera index2 read primer) so libraries are single indexed
                2- Addition of 6-7 bases to 3’ end of primers that extends primer binding into transposome recognition sequence common between both adapters

                I do not see any difference beside possible different concentrations of adapters used in PCR noting that Nextera adds PPM to supplement indexed primers as their concentration is lower than needed for Nextera. If you are using Nextera primers you might need to adjust annealing temperature and primer volume to optimise reaction.
                I'm in the same boat here. Have Nextera Index kit and was wondering if I could use it rather than the custom primers from the protocol. Anyone have any experience with this?

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #68
                  Originally posted by bman View Post
                  I'm in the same boat here. Have Nextera Index kit and was wondering if I could use it rather than the custom primers from the protocol. Anyone have any experience with this?
                  The NexteraXT index kit could work--however the standard ATACseq protocol expects the primer concentrations to be ~5X higher than they are supplied in the NexteraXT index kits. To find this out we just measured the concentration of the index oligos (which Illumina refuses to disclose for unknown reasons) using a Qubit.

                  So you could speed vap the Nextera index primers, or add more to the reactions (instead of the water in the protocol) or both. Or just use less primer, I suppose. Might still work.

                  I would recommend just ordering dual index oligos for the PCR step from a reputable oligo synthesis company. If you are planning to run these together in the same lane of an Illumina patterned flowcell instrument (HiSeq 3000/4000/X or NovaSeq -- and now maybe iSeq?) then you would want to use unique dual indexes anyway...

                  --
                  Phillip

                  Comment

                  • bman
                    Junior Member
                    • Sep 2014
                    • 7

                    #69
                    Phillip,
                    Thank you very much for the help. I think that ordering the custom primers for the future is definitely something that we will do, but I have the nextera index kit so was hoping to use it. I have found several examples of protocols that claim to have used the nextera index kit but clearly most people use the custom primers. I will try the nextera kit out and see what kind of results I get.

                    Comment

                    • bman
                      Junior Member
                      • Sep 2014
                      • 7

                      #70
                      I have been working on optimizing an atac-seq protocol on frozen tissue. I have been trying a few different things, and have attached bioA traces from two of my best runs. Has anyone ever gotten results like this? If so, is it sequenceable? I have found it hard to find a consensus "perfect" bioA trace for an atac prep online. I feel like the library might be best with some more input material. I also can't figure out why my 190 bp peak is consistently smaller than the 300 bp peak.





                      Thanks!
                      Last edited by bman; 02-22-2018, 10:36 AM.

                      Comment

                      • pennypeverell
                        Junior Member
                        • Mar 2018
                        • 2

                        #71
                        Atac seq problem - large fragments

                        Hello everyone!

                        I am new to ATAC seq and I'm having problems with my libraries. I'd very very thankful if anyone of you is willing to help.

                        From the bioanalyzer profile I se the 180 bp peak and it should be the mono-nucleosome peak. And that's ok.

                        The problem is I see other larger peaks (possibly the multiple nucleosomes?) but they have higher intensity than the 180 bp peaks!

                        I'm a completely noob, but I've been told this is a problem. That larger nucleosomes peaks should be present, but with less intensity compared to the 180 bp peaks. Do you suggest me to sequence these library or throw them away and trying again? Maybe trying to improve my tagmentation step?

                        I read some similar thread in the forum but nobody is actually having my same bioanalyzer profile (or at least I didn't found it).

                        I'm attaching some pics of the byoanalizer profiles so maybe you could better figure it out what I am talking about.

                        Please, any opinion/advice/suggestion would be a great help!

                        Thanks

                        Penny

                        Image Screen Shot 2018 03 16 at 18 42 40 hosted on ImgBB


                        Image Screen Shot 2018 03 16 at 18 41 53 hosted on ImgBB

                        Comment

                        • Márquez
                          Junior Member
                          • Jun 2021
                          • 1

                          #72
                          ATAC-seq

                          Good afternoon, I have a query. I want to use ATAC-seq to study accessible chromatin regions of a chromosome 9 genome region in cell lines that will be stimulated with pro-inflammatory cytokines.
                          How can I study this specific region of the chromosome 9 genome in cell lines with this method?

                          Would you need specific primers for this region of chromosome 9?
                          In which step of the ATAC-seq protocol would you use them?

                          I hope you can help me

                          Thank you

                          Comment

                          Latest Articles

                          Collapse

                          • SEQadmin2
                            Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                            by SEQadmin2


                            I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                            Here are nine questions we think about, in roughly the order they matter, before...
                            06-18-2026, 07:11 AM
                          • SEQadmin2
                            From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                            by SEQadmin2


                            Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                            The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                            ...
                            06-02-2026, 10:05 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by SEQadmin2, Yesterday, 05:37 AM
                          0 responses
                          5 views
                          0 reactions
                          Last Post SEQadmin2  
                          Started by SEQadmin2, 06-26-2026, 11:10 AM