Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • yxl
    Junior Member
    • Feb 2011
    • 4

    #1

    KAPA, dual-SPRI size selection

    The KAPA Hyper kit has a protocol for using AMPure beads for size selection (dual-SPRI), but it seems to me that the beadNA ratio is off (or perhaps I am not understanding the concept correctly). The protocol is below (also on p10-11 of the attached protocol pdf); to summarize, size selection of ~250-450bp fragments is performed with these steps:

    Part 1:
    - start with adapter-ligated library in 50uL volume
    - add 30uL of AMPure XP beads
    - place on magnet
    - keep the 75uL supernatant and dispose beads (beads contain >450bp fragments)

    Part 2:
    - the to the 75uL supernatant, add 10uL of AMPure XP beads
    - place on magnet
    - discard supernatant (contains fragments <250bp)
    - wash with 80% EtOH twice
    - dry beads
    - elute DNA

    Part 2, as I understand, is the "left side" selection. The addition of 10uL of AMPure beads to 75uL supernatant is a 10/75 = 0.13x bead ratio; would this not lead to a almost-complete loss of material? It seems to me that a more appropriate AMPure bead volume would be 60uL (60/75=0.8) for a 0.8x bead ratio. Perhaps I'm missing something.


  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Part 2, as I understand, is the "left side" selection. The addition of 10uL of AMPure beads to 75uL supernatant is a 10/75 = 0.13x bead ratio; would this not lead to a almost-complete loss of material? It seems to me that a more appropriate AMPure bead volume would be 60uL (60/75=0.8) for a 0.8x bead ratio. Perhaps I'm missing something.
    Supernatant from Part 1 contains 0.6x bead solution, so by addition of bead in Part 2 it will be 0.8x.

    Comment

    • AdrianaGeldart
      Junior Member
      • Feb 2014
      • 6

      #3
      Dear yxl,

      I am a member of the Kapa Scientific Support & Applications Team and would be happy to address your question.

      To perform the first size cut (to exclude fragments corresponding to library inserts larger than ~450 bp), you would add 30 uL of AMPure XP Reagent to 50 uL of DNA. This is the 0.6X cut as you correctly described.

      The second step would be to transfer the supernatant containing DNA fragments smaller than ~450 bp to a new plate and discard the plate with the beads to which DNA fragments larger than ~450 bp are bound. The second size cut is then performed (to retain fragments corresponding to library inserts larger than ~250 bp) by adding 0.2 volumes of Agencourt AMPure XP reagent to the first cut supernatant. This would be adding 10 uL of reagent to 75 uL of DNA. This 0.2 volumes corresponds to the volume of DNA prior to the first cut (i.e. 50 uL). We have found 0.2 volumes to be optimal. If you decrease the difference between the two cuts to <0.2, the peak width decreases, but at the cost of a higher loss of DNA. If you increase the difference between the two cuts to >0.2, recovery is better, but the peak width tends to broaden.

      I hope this helps to clarify the size selection. If you would like further explanation, please feel free to call the Support Line at 1-855-KAPA-BIO or email us at [email protected].

      Best Regards,
      Kapa Biosystems Support

      Comment

      • liaoyunshi
        Member
        • Jul 2017
        • 14

        #4
        Originally posted by nucacidhunter View Post
        Supernatant from Part 1 contains 0.6x bead solution, so by addition of bead in Part 2 it will be 0.8x.
        Hi nucacidhunter,

        So is that means the factor deciding the DNA size capture is the bead solution volume instead of the bead amount itself?

        Thanks.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          Yes. Bead solution to DNA solution volume ratio.

          Comment

          • finswimmer
            Member
            • Oct 2016
            • 60

            #6
            Hello,

            Originally posted by liaoyunshi View Post
            So is that means the factor deciding the DNA size capture is the bead solution volume instead of the bead amount itself?
            yes, because the PEG concentration in the solution is the critical reagence in the size selection process. I've found a nice blog post about it some time ago here: http://core-genomics.blogspot.de/201...eads-work.html

            fin swimmer

            Comment

            • liaoyunshi
              Member
              • Jul 2017
              • 14

              #7
              Originally posted by finswimmer View Post
              Hello,



              yes, because the PEG concentration in the solution is the critical reagence in the size selection process. I've found a nice blog post about it some time ago here: http://core-genomics.blogspot.de/201...eads-work.html

              fin swimmer
              Hi,

              I also found it! Thanks.

              Actually before reading this post, I just wondered why the second selection with less bead amount can capture small DNA. And I finally realize that it is the bead solution makes it capture different size DNA, and the bead solution is more and more concentrated after beads pouring in for the second selection.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              26 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              20 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-15-2026, 10:52 AM
              0 responses
              36 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              42 views
              0 reactions
              Last Post SEQadmin2  
              Working...