    size selection with Nextera XT

    I made a plant organelle DNA library of 5 samples using the Nextera XT kit. A bioanalyser run prior to Normalization shows fragments 200 bp onwards (less than 200 also in some samples) with a peak around 1500 bp( attachment bioanal 1). Since I'll be using a Miseq v2 reagent kit(300cycles PE), I need to size select.
    Q1. Is the size range of 200-500bp OK ?
    Q2. How will fragments of <200 bp affect cluster generation?
    Q3 . What protocol shall I use for size selection?
    a)I tried double SPRI (1x followed by 0.55X + 0.12X), and though it was effective in removing <200 bp I got nothing in the final eluent after 0.12X.
    b)I also tried eluting with E gel size select agarose gel(Invitrogen) in MQ water and collected fractions of 288,363,450,563bp ( attachment bioanal 2- Samples 1,2,3 have been eluted from sample 3 of bioanal 1)
    Q4. Can these be used further? Nowhere have I come across e-gel being used -I wonder why?
    Q5. If these can be used, then
    a) fraction of what size(s) can be taken/ pooled?(Each fraction is eluted in 25ul)
    b) What should be the consideration for the minimum concentration (for eg.the 288bp fragment is 1403pmol/l in sample 1) that can be used for proceeding with the normalization which is the next step. Since normalization of concentartion is done using the beads provided in the kit by Illumina, I don't know how this will work across samples.Particularly since I need to load 20 indexed libraries in one flow cell.
    Thanks
    Attached Files
    Last edited by abkg; 09-26-2014, 11:58 PM. Reason: forgot the attachments!!
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #2
    I don't think the tight size selection was necessary. If you are doing a 2x150 base run, then having your amplicons with >150base inserts is desirable. That means around 300 counting the adapters. So doing a single ampure at 0.7x or so would probably be sufficient. Long amplicons mixed in don't hurt anything, although very few of them actually cluster.

    But since you did it, probably the longest length fraction would give you the best assembly results. Presuming you have enough to cluster. Normally the amount you show on the high sensitivity chip would not be (unless it was diluted prior to running). But if you use the normalization beads it is possible you will have enough for the heat denature method.

    --
    Phillip

    Comment

    • abkg
      Junior Member
      • Sep 2014
      • 2

      #3
      Thank you for your suggestion. I will proceed for normalization after 0.7X ampure selection for the rest of my samples.
      abkg

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM
      • SEQadmin2
        New Genomics Technologies Take Aim at Long-Standing Limits
        by SEQadmin2


        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
        ...
        09-28-2026, 10:25 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 10-22-2026, 11:13 AM
      0 responses
      26 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 10:48 AM
      0 responses
      20 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-15-2026, 10:52 AM
      0 responses
      36 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-14-2026, 05:39 AM
      0 responses
      42 views
      0 reactions
      Last Post SEQadmin2  
      Working...